okay dropped the time base down to the ms range (100 or so) and AM getting a square wave though there the rise time seems too fast to register ANY sign of a vertical.. there seems to be significant overshoot too... but no matter... I think those are really tiny flaws..
so does that mean i just need to RTFM more..?
Yep.
Cal signal will be 3 or 5 V @ ~1KHz
SO for a 3V waveform your V/div setting @
100mV/div will show a 3 Div high waveform. (10x probe)But thats not all
@ timebase setting of
100 us/div 1 period will be 10 divisions long.NEVER connect a scope WITHOUT an understanding of the waveform you expect to measure.
Always remember the probe attenuation used and use simple maths to set the scope for the signal amplitude expected.
Overdriving the vertical amps of a CRO could damage your scope.
Got all that?
Now go play with the Cal signal some more until you get a feel for the attenuation and timebase adjustments and report back.