 SO i recently bought a ballantine 1066s oscilloscope, it powers on all the switches feel nice but all i get is a little dot which i can not get to form into a line at all i tried testing a few frequency sources to no avail. My question is this why only a little dot (focus works and turns it into either a vertical line or a bigger dot) and how do i fix it. Like an idiot i bought it from a long ways off so i either make it work or i have no oscilloscope. please help
« Last Edit: July 23, 2016, 09:32:09 pm by neo »
A hopeless addict (and slave) to TEA and a firm believer that high frequency is little more than modern hoodoo.
 

Offline MosherIV

  • Super Contributor
  • ***
  • Posts: 1530
  • Country: gb
Re: Ballantine 1066s oscilloscope question, need help
« Reply #1 on: July 23, 2016, 09:41:21 pm »
It is either in xy mode, external time base mode or it is broken.

Can you post a picture of the front panel so that we can tell if it is in exteranl time base mode?
 

Offline neoTopic starter

  • Super Contributor
  • ***
  • Posts: 1694
  • Country: us
  • The specialist.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #2 on: July 23, 2016, 09:59:42 pm »
forgive the delay my net is not the best
A hopeless addict (and slave) to TEA and a firm believer that high frequency is little more than modern hoodoo.
 

Offline MosherIV

  • Super Contributor
  • ***
  • Posts: 1530
  • Country: gb
Re: Ballantine 1066s oscilloscope question, need help
« Reply #3 on: July 23, 2016, 10:12:05 pm »
The time base is on 20us per division,
Trigger is on channel 2 , -ve, so they are ok.

Is the beam finder button stuck on? ( can you press it down and does it release?)

Try adjusting the intensity and focus to see if they is a line but it is not looking good.
 

Offline neoTopic starter

  • Super Contributor
  • ***
  • Posts: 1694
  • Country: us
  • The specialist.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #4 on: July 23, 2016, 10:23:40 pm »
Beam finder does work and as i said focus varies between that dot and a 2 division vertical line or 2 division blob (i am more specific on this here than on my original post though) if it is broke then the question is how to fix it because i dont have money for another one
A hopeless addict (and slave) to TEA and a firm believer that high frequency is little more than modern hoodoo.
 

Offline MosherIV

  • Super Contributor
  • ***
  • Posts: 1530
  • Country: gb
Re: Ballantine 1066s oscilloscope question, need help
« Reply #5 on: July 23, 2016, 10:32:23 pm »
Try openning up the scope and look carefully around the inside. Look for any loose wires hanging off.

I got lucky with my scope, similar problems to what you have, openned it up and found 1 wire was hanging off, just plugged it back onto the connector and it was working perfectly.

If you do not find anything like that, I am affraid you will have to find the fault and fix it. Start with searching for the service manual or schematic diagram.
 

Offline neoTopic starter

  • Super Contributor
  • ***
  • Posts: 1694
  • Country: us
  • The specialist.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #6 on: July 23, 2016, 10:36:57 pm »
about that service manual i dont have one and cant find one, give me a couple minutes to look at the guts
A hopeless addict (and slave) to TEA and a firm believer that high frequency is little more than modern hoodoo.
 

Offline neoTopic starter

  • Super Contributor
  • ***
  • Posts: 1694
  • Country: us
  • The specialist.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #7 on: July 23, 2016, 10:44:29 pm »
i see no loose wires though i can provide pictures of the interior if you would like
A hopeless addict (and slave) to TEA and a firm believer that high frequency is little more than modern hoodoo.
 

Offline MosherIV

  • Super Contributor
  • ***
  • Posts: 1530
  • Country: gb
Re: Ballantine 1066s oscilloscope question, need help
« Reply #8 on: July 23, 2016, 11:07:25 pm »
Affraid no good to me, I would need schematics to work from. Thought it might have been an easy thing with the controls. I am not that good at repair myself, I have got a Weller iron I was going to ask for help myself.

Try the repair foruum.

Or if it was an ebay sale, try asking for a refund since it does not work.
 
The following users thanked this post: edy, neo

Online tautech

  • Super Contributor
  • ***
  • Posts: 28381
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #9 on: July 23, 2016, 11:13:52 pm »
@ neo
If you hook a probe to the Cal output do you get a vertical line?
Can you move the dot left, right and up, down?

If all that works I suspect the Sweep generator has failed.

What other equipment do you have for fault finding?
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline neoTopic starter

  • Super Contributor
  • ***
  • Posts: 1694
  • Country: us
  • The specialist.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #10 on: July 24, 2016, 02:26:45 am »
@tautech i have a multimeter, capacitence resistance current and voltage (with a diode ester of course) and thats about it this oscilloscope is my first big piece of test equipment i don't have much in the way of test equipment come to think of it
« Last Edit: July 24, 2016, 02:31:55 am by neo »
A hopeless addict (and slave) to TEA and a firm believer that high frequency is little more than modern hoodoo.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28381
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #11 on: July 24, 2016, 04:30:18 am »
No problem, did you do those tests?
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline neoTopic starter

  • Super Contributor
  • ***
  • Posts: 1694
  • Country: us
  • The specialist.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #12 on: July 24, 2016, 07:38:43 am »
i can move the dot one division by adjusting focus but thats it i hooked a test probe to everything that would accept bnc except for channel 2 out that shocked me as did external trigger not alot but enough to let you know that you've been shocked
A hopeless addict (and slave) to TEA and a firm believer that high frequency is little more than modern hoodoo.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28381
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #13 on: July 24, 2016, 10:21:16 am »
You should have a 1 KHz probe cal output, commonly 3 or 5V p to p that if connected to the inputs as the scope is, should display a vertical line, being as I suspect the sweep generator has failed.
 
The sweep is a ramp waveform, the rising voltage is implied to the horizontal plates that electrostatically deflect the electron beam across the display. Frequency of the sweep is controlled by the timebase setting.

In order to diagnose the problem you must input a known waveform and the probe cal is perfect to this.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28381
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #14 on: July 24, 2016, 10:23:29 am »
You should be able to move the dot around the display with the vertical and horizontal position controls.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: neo

Offline StillTrying

  • Super Contributor
  • ***
  • Posts: 2850
  • Country: se
  • Country: Broken Britain
Re: Ballantine 1066s oscilloscope question, need help
« Reply #15 on: July 24, 2016, 11:37:58 am »
Also switching both CH1 and CH2 on, and switching to 'CHOP', should give you 2 separate dots.

AFAICT the only differences between the A, B, and S models is the bandwidth 15,20,25 Mhz.

Don't forget this thing contains 4 Killer Volts even when switched off.

http://lcweb2.loc.gov/master/mbrs/recording_preservation/manuals/Ballantine%2520Portable%2520Dual-Channel%2520Oscilloscope%2520Model%25201066A.pdf

http://www.ebay.com/itm/BALLANTINE-MODEL-1066B-Dual-Trace-Oscilloscope-Instruction-Manual-schem-17011-/390620703232
.  That took much longer than I thought it would.
 
The following users thanked this post: neo

Offline JPortici

  • Super Contributor
  • ***
  • Posts: 3461
  • Country: it
Re: Ballantine 1066s oscilloscope question, need help
« Reply #16 on: July 24, 2016, 09:32:33 pm »
two things i would do at first:
-check that the BEAM FINDER button is not stuck
-select the auto or free running trigger mode, that way you should *at least* see the dot moving left to right
 

Offline StillTrying

  • Super Contributor
  • ***
  • Posts: 2850
  • Country: se
  • Country: Broken Britain
Re: Ballantine 1066s oscilloscope question, need help
« Reply #17 on: July 24, 2016, 11:26:47 pm »
I would guess this is going to be the plates' HV which is missing, on the last photo of the ebay manual it's looks like it's +180V, which seems low for a 4KV tube to me, but there you go!
.  That took much longer than I thought it would.
 

Offline alsetalokin4017

  • Super Contributor
  • ***
  • Posts: 2055
  • Country: us
Re: Ballantine 1066s oscilloscope question, need help
« Reply #18 on: July 25, 2016, 05:09:02 am »
i can move the dot one division by adjusting focus but thats it i hooked a test probe to everything that would accept bnc except for channel 2 out that shocked me as did external trigger not alot but enough to let you know that you've been shocked

That's bad news. There should be nothing capable of shocking you even a little bit on the outside of the scope.
The easiest person to fool is yourself. -- Richard Feynman
 
The following users thanked this post: neo

Offline neoTopic starter

  • Super Contributor
  • ***
  • Posts: 1694
  • Country: us
  • The specialist.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #19 on: July 25, 2016, 09:39:32 am »
Good news i messaged the seller and after a brief conversation got half my money back and i get to keep the scope, i may bite the bullet but at least its a reasonable and peaceful end  :phew:
A hopeless addict (and slave) to TEA and a firm believer that high frequency is little more than modern hoodoo.
 

Offline neoTopic starter

  • Super Contributor
  • ***
  • Posts: 1694
  • Country: us
  • The specialist.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #20 on: July 25, 2016, 09:46:58 am »
   I'll choose to see this as a lesson, don't buy oscilloscope from over 700 miles away especially not ones that are dirt cheap and over 40 years old with no manual or schematics to speak of, which is obvious in hindsight but baby steps i am relatively new to this.
A hopeless addict (and slave) to TEA and a firm believer that high frequency is little more than modern hoodoo.
 

Offline MosherIV

  • Super Contributor
  • ***
  • Posts: 1530
  • Country: gb
Re: Ballantine 1066s oscilloscope question, need help
« Reply #21 on: July 25, 2016, 10:00:24 am »
Quote
Good news i messaged the seller and after a brief conversation got half my money back and i get to keep the scope
Nice one  :-+

Maybe it should not be too hard to fix.

As one of the others said, sounds like the time base ramp generator. Either
1. the low voltage circuit is not generating the ramp
2. the ramp is not reaching the HV amplifier
3. the HV amplifier is not working.

In my case it was 2. the wire+connector had fallen off the pcb connector. It was near the connector but no longer plugged in. The owner had told me that the scope had been knocked over and then the screen was all blured.

Did the owner say it was working before shipping?
Since your scope was shipped, it may have been knocked. Have another look around the inside (careful not to touch the HV stuff) and see there are any wires that look like they are not plugged into anything.
Are they any components that look like they are broken or blown?
Is there anything on the back of the scope that has been knocked?
 

Offline neoTopic starter

  • Super Contributor
  • ***
  • Posts: 1694
  • Country: us
  • The specialist.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #22 on: July 25, 2016, 09:45:43 pm »
 The owner said working but i don't know whether he meant it turns on or it generates traces working. I have yet to do a thorough poke around, and it does look a little beat up. As for a replacement oscilloscope how does this one look http://www.ebay.com/itm/Hitachi-Oscilloscope-V-212-20-MHz-/221976372042?hash=item33aed2a34a:g:wjIAAOSwp5JWYd5p i have the manual in pdf form complete with schematics i'm also looking at the Tektronix 2465 complete with manual if the auction doesn't go crazy
« Last Edit: July 25, 2016, 09:59:15 pm by neo »
A hopeless addict (and slave) to TEA and a firm believer that high frequency is little more than modern hoodoo.
 

Offline StillTrying

  • Super Contributor
  • ***
  • Posts: 2850
  • Country: se
  • Country: Broken Britain
Re: Ballantine 1066s oscilloscope question, need help
« Reply #23 on: July 26, 2016, 12:59:11 am »
As one of the others said, sounds like the time base ramp generator. Either
1. the low voltage circuit is not generating the ramp
2. the ramp is not reaching the HV amplifier
3. the HV amplifier is not working.

I'd say none of the above, because it effects both X and Y movements.

I wouldn't throw the scope away yet, from the photo the front panel still looks in very good condition for its age.
« Last Edit: July 26, 2016, 01:02:30 am by StillTrying »
.  That took much longer than I thought it would.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28381
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Ballantine 1066s oscilloscope question, need help
« Reply #24 on: July 26, 2016, 01:20:29 am »
As one of the others said, sounds like the time base ramp generator. Either
1. the low voltage circuit is not generating the ramp
2. the ramp is not reaching the HV amplifier
3. the HV amplifier is not working.

I'd say none of the above, because it effects both X and Y movements.

I wouldn't throw the scope away yet, from the photo the front panel still looks in very good condition for its age.
Quite.
Once this one's fixed with another you can use this one to keep the other one fixed.  ;)

Don't ask how I know.  :phew: