Bench supply remote sense
« on: April 12, 2017, 10:43:06 pm »
Why would one need a remote sense on a benchtop power supply?
 

Offline langwadt

  • Super Contributor
  • ***
  • Posts: 4414
  • Country: dk
Re: Bench supply remote sense
« Reply #1 on: April 12, 2017, 11:11:38 pm »
Remote sense compensate for the current dependent voltage drop on the wires.
With thick wires, low currents and no need for super accuracy you don't need it

I'm sure I remember Dave had a video that nicely illustrated it but I can't find it
 

Offline ratioTopic starter

  • Regular Contributor
  • *
  • Posts: 59
  • Country: us
Re: Bench supply remote sense
« Reply #2 on: April 13, 2017, 12:10:59 am »
Remote sense compensate for the current dependent voltage drop on the wires.
With thick wires, low currents and no need for super accuracy you don't need it
The only use I could come up with on my own was to compensate for undersized leads. I'd imagine there must be some other uses, but I just don't know what they might be.
 

Offline sleemanj

  • Super Contributor
  • ***
  • Posts: 3024
  • Country: nz
  • Professional tightwad.
    • The electronics hobby components I sell.
Re: Bench supply remote sense
« Reply #3 on: April 13, 2017, 05:58:25 am »
The only use I could come up with on my own was to compensate for undersized leads. I'd imagine there must be some other uses, but I just don't know what they might be.

When you want to deliver a constant voltage to the device under test (DUT) under various load conditions.

Very common example is testing dc-dc converter efficiency, you want to test their characteristic at different current draws, but in order to have useful data you need to ensure that the input voltage is the same for each series of tests (ie you want to test input yV, output zV at various different currents so you can see how the efficiency or other characteristics vary with current).  If you don't have a remote sense power supply, you have to measure and tweak the voltage each time.
~~~
EEVBlog Members - get yourself 10% discount off all my electronic components for sale just use the Buy Direct links and use Coupon Code "eevblog" during checkout.  Shipping from New Zealand, international orders welcome :-)
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Bench supply remote sense
« Reply #4 on: April 13, 2017, 06:59:53 am »
Very common example is testing dc-dc converter efficiency "accurately" , you want to test their characteristic at different current draws, but in order to have useful data you need to ensure that the input voltage is the same for each series of tests (ie you want to test input yV, output zV at various different currents so you can see how the efficiency or other characteristics vary with current). 

If you don't have a remote sense power supply, you have to measure and tweak "and guess" the "different" voltages each time , "at different load level".

Added for completeness sake.  >:D (highlighted in red)

Offline dimkasta

  • Regular Contributor
  • *
  • Posts: 185
  • Country: gr
Re: Bench supply remote sense
« Reply #5 on: April 13, 2017, 08:05:46 am »
for quick testing and prototyping, It can also eliminate or at least moderate the effects of incomplete or non-finalized local decoupling.
 

Offline Avacee

  • Supporter
  • ****
  • Posts: 299
  • Country: gb
Re: Bench supply remote sense
« Reply #6 on: April 13, 2017, 08:49:22 am »
I'm sure I remember Dave had a video that nicely illustrated it but I can't find it

OP: This video demonstrates how with the voltage drop from the Power Supply to the Device Under Test the actual voltage at the DUT is not what you think it is.
It also shows how using the remote sense terminals shows the actual voltage at the DUT.

If you see the same YouTube image as I do below you can see one Multimeter reads 4.691V yet the Power Supply reads 5.000V.
There's a drop of 0.309V across the wires. The Remote Sense would let the PSU set itself to have 5V at the DUT.
Dave later has to set more than 5V on his PSU to get 5V at the DUT.

« Last Edit: April 13, 2017, 09:03:10 am by Avacee »
 