you will probably be ok, but if you go measuring a high frequency from a high impedence source like that expect distortion and inaccuracies
Its not just high impedance, even low impedance sources can have distortions appear when using a BNC + cliplead adapter. Just saw this at work the other day. For some reason, my coworkers like to use coax + clip lead adatpers on scopes (fair enough I guess, they are laying around for on power supplies etc.. ) But with ~ 1m of coax and a 5 MHz RC filtered signal, a bunch of ripple on the rising edge showed up. Why? The source impedance from the high speed buffer was too low (way below 50 ohms) so the reflections off the 1M input impedance of the scope hit the low impedance source, and bounce back up to the scope again. But the same behavior isnt observed with a signal generator. The sig gen has a 50 ohm series termination, so any signals that bounce back get absorbed by that. With a proper 10x probe, or with a 50 ohm tee'd on the original (crap 1074Z's dont have internal 50 ohm?
) Everything looked fine.
50 ohm series resistor for the probe, 50 ohm termination, if you can set your scope up manually to have a 2x probe, then the readings will be correct too. Guessing a power supply wont complain about a 100 ohm load on it