Author Topic: Can I use 1:1 passive probes with 64pF shunt capacitance with a 1054z?  (Read 368 times)

0 Members and 1 Guest are viewing this topic.

Offline julian1

  • Frequent Contributor
  • **
  • Posts: 269
  • Country: au
  • http://blog.julian1.io

Does 64pF, imply actual compensation adjustment anywhere from 10pF - 64pF ?
 
The probes are these,

http://www.keysight.com/en/pd-1000000977%3Aepsg%3Apro-pn-10439A/11-2m-miniature-passive-probe?cc=US&lc=eng

If not - then can I change the capacitor trimmer?

Offline tggzzz

  • Super Contributor
  • ***
  • Posts: 5255
  • Country: gb
    • Having fun doing more, with less
Re: Can I use 1:1 passive probes with 64pF shunt capacitance with a 1054z?
« Reply #1 on: February 21, 2017, 08:01:08 PM »
Does 64pF, imply actual compensation adjustment anywhere from 10pF - 64pF ?
If not - then can I change the capacitor trimmer?

It means the tip capacitance that loads your circuit is 64pF. *1 probes don't have a frequency compensation trimmer capacitor.

They can be used with a 1054Z scope, with the standard performance limitations.
There are lies, damned lies, statistics - and ADC/DAC specs.
Gliding aphorism: "there is no substitute for span". Retort: "There is a substitute: skill+imagination. But you can buy span".
Having fun doing more, with less
 
The following users thanked this post: julian1

Online danadak

  • Super Contributor
  • ***
  • Posts: 1311
  • Country: us
  • Reactor Operator SSN-583, Retired EE
Re: Can I use 1:1 passive probes with 64pF shunt capacitance with a 1054z?
« Reply #2 on: February 21, 2017, 11:56:23 PM »
Of course you have significant circuit loading, at 1 Mhz Xc ~= 2.5K
ohms, quite low.

Completely out of the question to probe a crystal oscillator circuit
for example, at the crystal.


Regards, Dana.
 