Does 64pF, imply actual compensation adjustment anywhere from 10pF - 64pF ?
The probes are these,
http://www.keysight.com/en/pd-1000000977%3Aepsg%3Apro-pn-10439A/11-2m-miniature-passive-probe?cc=US&lc=eng (http://www.keysight.com/en/pd-1000000977%3Aepsg%3Apro-pn-10439A/11-2m-miniature-passive-probe?cc=US&lc=eng)
If not - then can I change the capacitor trimmer?
-
Does 64pF, imply actual compensation adjustment anywhere from 10pF - 64pF ?
If not - then can I change the capacitor trimmer?
It means the tip capacitance that loads your circuit is 64pF. *1 probes don't have a frequency compensation trimmer capacitor.
They can be used with a 1054Z scope, with the standard performance limitations.
-
Of course you have significant circuit loading, at 1 Mhz Xc ~= 2.5K
ohms, quite low.
Completely out of the question to probe a crystal oscillator circuit
for example, at the crystal.
Regards, Dana.