Hi all,
I have recently started having electronics as a hobby and reading a couple of books and watching a few videos (Dave, my deepest, wholehearted thanks) and purchased a few pieces of gear to equip my home "lab". As one of those, I managed to get a Thandar TS3022S dual bench power supply in an auction from a company. One of units seems to work ok, but the other one exhibits a weird time drifting behavior. It can drift as much as half volt if left enough time on. This doesn't seem to be a reasonable feature on what is supposed to be a precision power supply.
So, a couple of days ago I decided to try to something about that and see if I could solve that issue somehow. I started the unit and waited patiently and after a while I could see how the voltage in the indicator started raising (and this seems to match with the readouts from a DMM). Next, I tried unplug the unit and open the case to take a look inside and see if there was anything that would stand out as an obvious problem. There wasn't.
I have a UNI-T 61D DMM which can be plugged to a PC and record the measured values. So my next step was to try to document the phenomena because I had also seen at some point some fluctuations on the output of a few tens of millivolts in a few seconds period. I started recording (output of 12V with no load) and after a few hours the readout hadn't moved more than 10mV. Since I suspected some problem with overheating in the unit and now the cover was removed, I put the cover back and tried again. Nothing. So at this moment the power supply seems to be working fine. It shows a little offset on the voltage (about 70mV) and I even tried to measure the ripple and noise following recent Dave's video instructions (with the limitation that my scope's probes have a 6Mhz bandwidth limit in 1x) and the values are about 2mVrms and 5mVpp which seem to be a little bit out of spec (from specsheet: "typically < 1mV rms") but that could just be me incorrectly measuring. This values of offset and ripple and noise seem to be common for both units.
In any case, now it seems to be working fine for the time being, but it left me scratching my head. So any ideas on what might have been the reason and whether it will show up again are very much appreciated.
While I was trying to measure the output I observed something weird with my UT61D dmm. I tried to record the output in two ways. One of them was using the regular autorange in the V selection and just readout whatever it came out. But since this is a 6000 count multimeter, at 12V you only get 10mV resolution. So, I tried to select manual 600mV range in the mV selection and try to make a relative measurement zeroing the inital value. This seemed to work fine, but then I observed that the power supply was sourcing 6mA (in the V selection this is 0, at least in the readout of the unit which has a 1mA resolution).
I have checked the specifications of the UT61D and it says that for V range, the input impedance is 10MOhm and that for the mV range it should be >3000MOhms. So, I grabbed another dmm (a fluke 179) and measured resistivity on both modes on the UT61D. On the V range it is showing about 11 MOhms, but the funny thing is that in the mV range it shows about 1.54MOhms, way below specs, and actually matching relatively well the 6mA drawn from the power supply . But the measurement in the UT61D is accurate in the mV range, I tried to measure several values in that range and both the Fluke 179 and the UT61D are a match to the tenth of millivolt. So, the UT61D seem to be working fine and at the same time out of spec.
So, does anyone know have a hint on what is going on with both the power supply and the dmm?. Thanks a lot!