Solved!
« on: November 15, 2017, 06:57:08 pm »
Photo 1 - two channels grounded but traces not parallel to grid (and not horisontal anyway)

Photo 2 - chanel 2 connected with 36" RG58 directly (without any dividers)  to build-in calibrator 2V 1kHz but fronts are ugly  :palm:

How  can I fix it?

THANX! FIXED!
« Last Edit: November 16, 2017, 01:31:10 pm by 001 »
 

Offline w2aew

  • Super Contributor
  • ***
  • Posts: 1780
  • Country: us
  • I usTa cuDnt speL enjinere, noW I aR wuN
    • My YouTube Channel
Re: Epic scope problem (photos inside)
« Reply #1 on: November 15, 2017, 07:18:56 pm »
The tilt is often caused by an external magnetic field, sometimes even the earth's magnetic field.  Most scopes include a "trace rotation" adjustment to correct for this.  Usually, it's a screwdriver accessible trimmer accessible on the front panel.  For some older scopes, there were wingnuts on the back of the scope that would allow you to physically rotate the CRT to correct for the tilt.

The calibrator signal looks as expected.  Remember, the sole purpose of the calibrator signal (which is more accurately called the "probe compensation signal") is to adjust the low-mid frequency compensation adjustment on 10x probes.  It isn't really designed to drive coax directly.
YouTube channel: https://www.youtube.com/w2aew
FAE for Tektronix
Technical Coordinator for the ARRL Northern NJ Section
 
The following users thanked this post: 001

Offline 001Topic starter

  • Super Contributor
  • ***
  • Posts: 1170
  • Country: aq
Re: Epic scope problem (photos inside)
« Reply #2 on: November 15, 2017, 07:35:30 pm »
Thanx a lot!

I trim my scope now and what I see:

photo 1 - calibrator and 1:10 probes
photo 2 - calibrator and 1:1 probes

Is it fine?

Signal front looks "reversed" isn`t it?
 

Offline w2aew

  • Super Contributor
  • ***
  • Posts: 1780
  • Country: us
  • I usTa cuDnt speL enjinere, noW I aR wuN
    • My YouTube Channel
Re: Epic scope problem (photos inside)
« Reply #3 on: November 15, 2017, 08:04:02 pm »
Thanx a lot!

I trim my scope now and what I see:

photo 1 - calibrator and 1:10 probes
photo 2 - calibrator and 1:1 probes

Is it fine?

Signal front looks "reversed" isn`t it?

Your images look fine/normal.  The calibrator is *only* intended to be used on 10x probes so that you can adjust their compensation adjustment.  The rounded edges on the 1:1 probe (and your earlier direct coax image) is due to the higher capacitance that each of these connections have.
YouTube channel: https://www.youtube.com/w2aew
FAE for Tektronix
Technical Coordinator for the ARRL Northern NJ Section
 
The following users thanked this post: 001

Offline Gyro

  • Super Contributor
  • ***
  • Posts: 9480
  • Country: gb
« Last Edit: November 15, 2017, 08:52:15 pm by Gyro »
Best Regards, Chris
 

Offline james_s

  • Super Contributor
  • ***
  • Posts: 21611
  • Country: us
Re: Epic scope problem (photos inside)
« Reply #5 on: November 15, 2017, 08:25:14 pm »
Epic? This looks like a very minor issue that is part simple adjustment and part user error.

From the title I was expecting smoke damage and a crater in the PCB.
 