EX300 Micro Amps
« on: March 25, 2012, 04:41:25 am »
I'm getting weird readings when I try to measure uA with my EX-330. When measuring with the mA mode I get 3.9, and when measuring across a shunt resistor I also get 3.9mA. But when the meter is in uA mode, I get around 1000uA. What's going on??? I recall Dave saying the closer you get to the limits of the meter, the more you're affected by burden voltage -- and the circuit also ran on 3.3v. But the mA mode was pretty close, so what's really happening here?
 

alm

  • Guest
Re: EX300 Micro Amps
« Reply #1 on: March 25, 2012, 10:01:17 am »
I assume that the uA range goes up to 4mA. The uA mode probably has a much higher shunt resistance than the mA mode (see DMN manual). The higher burden voltage is screwing up your circuit, which you didn't mention, somehow.
 

Offline jyaanTopic starter

  • Contributor
  • Posts: 18
Re: EX300 Micro Amps
« Reply #2 on: March 28, 2012, 08:01:36 pm »
The "circuit" is just an LED and resistor. I originally supplied 3.3v from a USB development board, and also tried two AAs and a 9v, but they all had similar results. Oh well, maybe I should consider getting a uCurrent.
 