Also to help any one in the future here is the list of things I checked for:
Fuse TestsInsert the black lead into the COM socket and the red lead into the V/Ohms socket
1. Switch meter to the diode function
- Use the red lead to probe the A socket. If it beeps the 10A fuse is good.
- Use the red lead to probe the mA socket. If it beeps the 400mA fuse is good.
2. Switch meter to the resistance measurement function
- Use the red lead to probe the upper part of the A socket. The meter should read between 0.0R and 0.5R.
- Use the red lead to probe the upper part of the mA socket. The meter should read between 0.995k and 1.005k.
- If the meter reads out these ranges, it could be that either the fuse is blown or it's out of spec.
3. Switch meter to the continuity function
- Use the red lead to probe the A and mA. If the meter beeps the fuses are good.
LCD TestWith the meter off, hold down autohold button and keep holding it down
- Switch the meter to AC volts measurement
- All segments on the LCD will switch on, allowing you to make sure the LCD is 100%
Button TestSwitch the meter to AC volts measurement
- Press all buttons on the meter and listen for a corresponding beep.
- If you push a button and the meter doesn't beep, it could be faulty.
Calibration CountHold down the min max button and turn the meter on into AC voltage mode
- Press the autohold button
- The number of times the meter has been calibrated will be displayed in the form nxxx
- Where xxx is the number of times the meter has been calibrated.
- Don't push any other buttons, switch the meter off.
- Use this to check claims of recent calibrations for example.
Other Tests1. Check the battery terminals for corrosion
2. Tests the probes by switching the meter to continuity mode. Bend leads around to make sure they have no intermittent faults.
3. Take some common components, resistors, batteries etc and check that the corresponding modes on the meter give real results.
4. If possible, open the meter and visually inspect the fuses. Also look for any signs of a high energy short.
5. Shake the meter and make sure nothing rattles