HP 3487A current measurements
« on: August 07, 2013, 11:11:56 pm »
I just did some simple current test measurements with HP3478A multimeter compared to my fairly new Fluke 87 V. I used low noise lab power supply loaded with 3ohm 1% 20W resistor as a load. The resistor was spot on, measured 3.0041ohm with my BK precision LCR meter.

HP measurements: power supply output: 5.495V
                            resistor voltage: 4.999V
                            current: 1.658A

Fluke 87V measurements:  power supply output: 5.4906V
                                        resistor voltage: 5.187
                                        current: 1.723A

Here is my question: Given the numbers I can get about 0.176ohm shunt with Fluke 87V. Is the internal shunt resistor 0.1 ohm actually? (+ the resistance of the leads connected would be the 0.176ohm)

HP 3487A shows approx. 0.3ohm shunt with leads connected. The manufacturer specification say 0.1 ohm? I have no idea why the shunt is so high. I was playing with the front/rear terminals switch, it's kinda dodgy but I don't believe that could be the problem. The current difference between these two multimeters is 65mA, which is very high number. Any idea why?
         
                             
 

Offline ve7xen

  • Super Contributor
  • ***
  • Posts: 1193
  • Country: ca
    • VE7XEN Blog
Re: HP 3487A current measurements
« Reply #1 on: August 08, 2013, 12:49:42 am »
Lower total resistance = higher current, so the smaller shunt makes a smaller total resistance, allowing a larger current with the same input voltage.

Not sure where you're getting the 0.1 ohm specification, the spec says 0.3 ohm maximum, not including leads of course.
73 de VE7XEN
He/Him
 

Offline IvoSTopic starter

  • Frequent Contributor
  • **
  • Posts: 310
  • Country: us
Re: HP 3487A current measurements
« Reply #2 on: August 08, 2013, 01:31:15 am »
I have service manual for HP 3468 and it says 0.1 ohm shunt. Aren't those two units basically the same?
 

Offline ve7xen

  • Super Contributor
  • ***
  • Posts: 1193
  • Country: ca
    • VE7XEN Blog
Re: HP 3487A current measurements
« Reply #3 on: August 08, 2013, 03:08:18 am »
Apparently they have different current shunts  :-DD

Agilent usually posts full service and operation manuals for their older equipment on their own site, and this case is no different. You can get full manuals for your instrument here.
73 de VE7XEN
He/Him
 

Online Dr. Frank

  • Super Contributor
  • ***
  • Posts: 2384
  • Country: de
Re: HP 3487A current measurements
« Reply #4 on: August 08, 2013, 05:00:02 am »
I have service manual for HP 3468 and it says 0.1 ohm shunt. Aren't those two units basically the same?

Hi,

Page 149 of the 3478A service manual shows the 3A shunt, R107 = 0.1 Ohm, but in series there's a 3A fuse which will increase the resistance of the current circuit. This will make the difference. And also, a maximum burden of 1V, equivalent to 0.3 Ohm,  is specified.

Small changes in the overall serial resistance in current mode will lead to big differences in that kind of circuitry you use.

Frank
« Last Edit: August 08, 2013, 08:38:01 am by Dr. Frank »
 

Offline Dave

  • Super Contributor
  • ***
  • Posts: 1352
  • Country: si
  • I like to measure things.
Re: HP 3487A current measurements
« Reply #5 on: August 08, 2013, 01:49:48 pm »
Forget about the shunt resistances and burden voltages. Connect both meter in series and put a current through them. The same current will be flowing through them, no matter what shunt resistance.

KISS ;D
<fellbuendel> it's arduino, you're not supposed to know anything about what you're doing
<fellbuendel> if you knew, you wouldn't be using it
 

Offline IvoSTopic starter

  • Frequent Contributor
  • **
  • Posts: 310
  • Country: us
Re: HP 3487A current measurements
« Reply #6 on: August 08, 2013, 07:59:46 pm »
I thought about the same thing yesterday night so I had to try it before bed time  :D .And yes, both units measured within 1mA difference if all connected in series, reading some 1.6xxA. But the HP3478A is absolutelly unusable for higher current precise measurements due to higher burden voltage. Thank all of you guys for an effort of clarifying this thing.
 

alm

  • Guest
Re: HP 3487A current measurements
« Reply #7 on: August 08, 2013, 09:43:52 pm »
The effect of burden voltage depends on the circuit voltage. For a 1.8 V system it's unusable, but for a 48 V system it doesn't really matter.
 

Offline baoshi

  • Regular Contributor
  • *
  • Posts: 167
  • Country: sg
    • Digital Me
HP 3487A current measurements
« Reply #8 on: August 09, 2013, 08:22:22 am »
Think this was mentioned in EEVblog #2. The 3478A burden voltage "sucks". The same reason I bought a uCurrent immediately after I bought the 3478A. But uCurrent is not able to measure too high current. Heard uCurrent 2 will have amp range.
 