After watching w2aew #55 I tried to measure a series LC circuit for resonance. I found that the resonance point was where I expected, but I could not get the total signal (across both L+C) to 0V or even close. After some head scratching I realized that it's because the inductor phase was not 90° to the input signal. Am I doing something wrong?
Here are traces of the capacitor and inductor alone, with a 50r current sense resistor (annotated "load") to measure the input signal. The inductor is 7m2 with DCR 83r at the test frequency of 5300Hz. Inductor SRF is 1.4MHz. I guess I should test at the SRF but I haven't come across any information that the frequency affects the phase.
In the cap trace, the phase is -91.x and this is within the phase error of the 2 probes.
For 20 traces, I found that the probes themselves are out of phase an average of 0° but a max of about 2°. That seems pretty damn noisy to me, is my expectation out of wack? I'm measuring across ch. 2 and ch. 3 which are on different "banks" of the scope, would I get better results if I used ch. 1 and ch. 2 instead? Even if that corrected it or I otherwise solved that issue, note that the inductor phase error is 13° by itself, and that in the LC circuit the phase error L and C is 20°. So the probe error is small compared to inductor "error".