Isolation xformer for a scope; when do you need one?
« on: December 15, 2012, 06:04:56 pm »
Other than the obvious; measuring direct AC from a outlet, what other situations does on need a isolation xformer?
Testing AC voltages that are not directly off of the main, eg; any AC voltage within a device that had to go through some step up or down circuity, is one needed?

This might sound silly, but does it matter if it is in series with the scope or the DUT?
 

alm

  • Guest
Re: Isolation xformer for a scope; when do you need one?
« Reply #1 on: December 15, 2012, 06:46:53 pm »
Other than the obvious; measuring direct AC from a outlet, what other situations does on need a isolation xformer?
Equipment that is not isolated from the mains supply, eg. UPSs.

Testing AC voltages that are not directly off of the main, eg; any AC voltage within a device that had to go through some step up or down circuity, is one needed?
It's useful when measuring on the primary side of the power supply, eg. when debugging a switching power supply.

This might sound silly, but does it matter if it is in series with the scope or the DUT?
You want to connect it only to the DUT, the scope should always be grounded, even in battery powered mode (most decent manufacturers include statements to that effect in the manual). The exception to this is isolated scopes, eg. Fluke ScopeMeters or the Tektronix TPS2000 series, these don't need either a ground connection or an isolation transformer.

Normal scopes have exposed pieces of metal (eg. BNC connectors, USB connectors) that are directly connected to the probe ground leads, so it's essential to ensure this is at a safe potential. Floating the scope and clipping the ground lead to mains could be lethal. Even apart from the exposed pieces of metal, the knobs and buttons are not rated for mains voltages.

An alternative to an isolation transformer would be differential probes.
 

Offline videobruceTopic starter

  • Frequent Contributor
  • **
  • Posts: 464
  • Country: us
Re: Isolation xformer for a scope; when do you need one?
« Reply #2 on: December 15, 2012, 06:52:22 pm »
Using a isolation xformer always "lifts" the ground?
 

alm

  • Guest
Re: Isolation xformer for a scope; when do you need one?
« Reply #3 on: December 15, 2012, 06:59:38 pm »
Not necessarily, but there's not much point in hooking up a scope to an isolation transformer without lifting the ground, since the inputs are ground referenced.
 

Offline videobruceTopic starter

  • Frequent Contributor
  • **
  • Posts: 464
  • Country: us
Re: Isolation xformer for a scope; when do you need one?
« Reply #4 on: December 15, 2012, 07:54:20 pm »
Quote
but there's not much point in hooking up a scope to an isolation transformer without lifting the ground
I think I'm missing something here. I though the purpose of a IT was to isolate the high side from each other, not the ground.
 

alm

  • Guest
Re: Isolation xformer for a scope; when do you need one?
« Reply #5 on: December 15, 2012, 08:35:02 pm »
The standard mains supply is ground referenced. If you are connected to ground somehow (eg. bare foot on a concrete floor) and you touch the live conductor, you form a path between live (which may be at ~110 V relative to ground) and ground, and will get shocked. If you connect the reference lead of a scope probe to this live conductor, you are shorting this conductor to ground, which may lead to damage to the DUT, the scope or you (see Dave's how not to blow up your scope video).

One way around this is an isolation transformer. An isolation transformer isolates the 'live' and 'neutral' wires from ground, so you can safely touch both ground and one of the outputs from the isolation transformer. Note that there is still a potentially lethal 110 V between the two outputs. You can also connect a grounded scope reference lead to one of the outputs, this will place this output at ground level, and the other at 110 V relative to ground. This means the outputs are no longer floating, and fireworks will ensue if you try to also ground the second output. The same applies to any transformerless DUT connected to the isolations transformer.

If the DUT contains a transformer and is double isolated (no ground connection), then the circuit will look like the top circuit in this schematic:


The motor will be floating regardless of the presence of the isolations transformer. The isolation transformer will only affect the primary side of the DUT power transformer.

A scope is usually grounded, that gives the second circuit. The output is grounded regardless of the presence of the isolation transformer. The same applies to any other grounded circuit. This can be bypassed by cutting the ground connection, but this is strongly recommended against for test equipment like scopes, and should also be done very carefully for DUTs with exposed grounded metal.
 