Now let's work on using the scope better.
1. Assure that both the Probes (switch on the probe body) and the Scope Channel input attenuation ("Probe" submenu) are set to the 10x position, so you get realistic voltage values on your readout. I really don't think you are providing 100 volt pulses to your mosfet Gate. Adjust the Channel Vertical Scale settings appropriately.
99 percent of scope measurements will properly be made with the probe and channel set to the 10x position. Using 1x is appropriate only in certain situations like very low signal levels, relatively low frequencies, direct cable connection to a signal source without using the probe, &etc. Almost always we want the probe body switch and the Channel "Probe" settings to match. In the present case using 10x on both the probe and the channel is appropriate.
2. Adjust Trigger level so that you are getting a stable set of traces rather than the jittery things in your photos. This may involve setting the trigger to the very top of the CH1 trace. Alternatively you can try a Single Shot capture, where the scope will automatically stop after a single horizontal scan.
3. Use the built-in screen-saving process to make your screenshots rather than photographing the screen. I see that you have a USB storage device inserted in the front-panel USB port already, so simply press the "printer" button underneath the Help key at the top right corner of the scope's panel to save the screenshot to the USB storage device. The scope defaults to the .png format but you can select different formats in the Storage menu should you desire, but .png is fine. To speed up saving screenshots to USB storage you may want to Stop the scope with the Run/Stop button before saving. Then you can simply post the image file from the USB stick without having to go through the photography process, and you'll get a much better image. With perhaps 1/8th or 1/10th the file size in bytes... and no dust!