Hi, I recently bought a Tek 2225 scope on eBay. The screen produces a trace and can display both channels at the same time. But when I hooked it up to my function generator, it does not produce a trace whatsoever. What are some troubleshooting steps I can take or should I just return it?
Thanks
 

Offline ataradov

  • Super Contributor
  • ***
  • Posts: 11258
  • Country: us
    • Personal site
Re: Non-functioning oscilloscope?
« Reply #1 on: October 26, 2017, 02:36:58 am »
Check that you have coupling in DC position first. If this is the case, then only disassembly and inspection will tell.
Alex
 

Offline BrianHG

  • Super Contributor
  • ***
  • Posts: 7733
  • Country: ca
Re: Non-functioning oscilloscope?
« Reply #2 on: October 26, 2017, 02:53:02 am »
set your vertical resolution to it's largest volt/div setting and slow down your horizontal time base.
 

Online tggzzz

  • Super Contributor
  • ***
  • Posts: 19497
  • Country: gb
  • Numbers, not adjectives
    • Having fun doing more, with less
Re: Non-functioning oscilloscope?
« Reply #3 on: October 26, 2017, 08:31:46 am »
If this is your first scope, read the operators manual to understand the effects of the controls.

http://w140.com/tek_2225_op_and_opt.pdf
There are lies, damned lies, statistics - and ADC/DAC specs.
Glider pilot's aphorism: "there is no substitute for span". Retort: "There is a substitute: skill+imagination. But you can buy span".
Having fun doing more, with less
 

Offline danadak

  • Super Contributor
  • ***
  • Posts: 1875
  • Country: us
  • Reactor Operator SSN-583, Retired EE
Re: Non-functioning oscilloscope?
« Reply #4 on: October 26, 2017, 10:50:17 am »
Set scope to page 4-1, 4-2 settings, trigger to auto, trigger source to INT (internal), vertical to AC.


http://w140.com/Oscope_Tek2235-op.pdf


Regards, Dana.


Love Cypress PSOC, ATTiny, Bit Slice, OpAmps, Oscilloscopes, and Analog Gurus like Pease, Miller, Widlar, Dobkin, obsessed with being an engineer
 

Online Shock

  • Super Contributor
  • ***
  • Posts: 4216
  • Country: au
Re: Non-functioning oscilloscope?
« Reply #5 on: October 26, 2017, 07:20:00 pm »
Try the inbuilt probe adjustment signal. It's just to the left of the trigger BNC socket. But as people mentioned, an analog oscilloscope requires familiarity with the oscilloscope to manually setup the trigger, coupling, horizontal/vertical etc properly.

Try both channels, you might also want to check your probe as well and swap it if you have two, or test with a multimeter.
Soldering/Rework: Pace ADS200, Pace MBT350
Multimeters: Fluke 189, 87V, 117, 112   >>> WANTED STUFF <<<
Oszilloskopen: Lecroy 9314, Phillips PM3065, Tektronix 2215a, 314
 

Offline james_s

  • Super Contributor
  • ***
  • Posts: 21611
  • Country: us
Re: Non-functioning oscilloscope?
« Reply #6 on: October 27, 2017, 08:52:38 pm »
The first thing to do is make sure you know how to properly use it. I'm not trying to suggest you don't, but more than once people have brought me "broken" scopes that in fact worked just fine, they just didn't know how to operate it.
 