edit: and some hi speed opamp will oscillate if driving capacitive load such as coax cable, usually 50ohm impedance (if coax is 50ohm type) is placed on the opamp's output before the coax to avoid this. better if you can afford another 50ohm ground termination, but you'll get half the amplitude, but max power transfer. YMMV
edit2: and i dont think changing to intrument amp will make any difference. the output is all that matters (albeit low noise level), not its intrumentation inputs.
Thanks Mechatrommer
You are right, I don't really need an instrumentation op amp on the output, just on the input. I thought it might be simpler than using two separate op amps (second one being a voltage follower), which is what my current design uses. I can change that one to a dual op amp chip, with a voltage follower, and leave the other AD8221 for the input.
With coax, its better to impedance match the op amp output to the receiver input, even at these low frequencies.
If you're using 50 ohm coax, then put a series resistor Zs and a terminating resistor Zl as 50 ohms as in the above photo. Now adjust the output gain of Vs, which I presume is your sender op amp, to the voltage you require assuming a 50 ohm divider.
Is the receiver an inverting op amp too? If so, the gain resistors will be affected by the input Zl interacting with the gain resistors causing the output voltage to be not what you calculated. It'll be easier to use the non-inverting mode or a general fix is put another voltage follower at the receiver input after Zl to buffer it before the 100k ohm instrument.
Thanks Saturation.
I am trying to drive a scientific instrument which accepts -2 to 10V on BNC. It says it has a 100K impedance. I can't change that end, nor tell what it is using internally. I have other commercial devices that generate DA that drive that device fine over the same coax length.
From your above image, I can only manipulate Zs, not Zl.
It sounds like what I should do is what I am currently doing now, which is use a voltage follower (noninverting op amp) after a normal op amp (which does the voltage math).
But, should a resistor be used in series like Zs from your above image (which is what Dave did
http://www.alternatezone.com/electronics/ucurrent/uCurrent%20Schematic.png ) , or should I use a low ohm resistor in the feed back loop of the voltage following op amp, which is what I am doing now, which won't change the voltage output?
follow up
should I use any capacitors on the output as well?
page 49:
http://www.analog.com/static/imported-files/data_sheets/AD8221.pdfI am using capacitors on the input of the other AD8221 to reduce EMI/RF.
Cheers