Author Topic: Rogol 1054Z actual bandwidth  (Read 2706 times)
Offline rad400
« on: October 31, 2017, 12:01:23 pm »
When a Rigol DS1054Z 50Mhz scope is updated to a 100Mhz, do you really get an improvement in the bandwidth?.   At 100MHZ are you down only 3db or considerably more?
Offline MarkF
Re: Rogol 1054Z actual bandwidth
« Reply #1 on: October 31, 2017, 12:29:32 pm »
The 3dB bandwidth for the DS1104Z 100MHz scope is approx 160Mhz.
The 3dB bandwidth for the DS1074Z 75MHz scope is approx 90Mhz.
Online rstofer
« Reply #2 on: October 31, 2017, 12:41:12 pm »
You need to watch Dave's video on the subject.  There are two filters on the front end of each channel.  One for 50 MHz bandwidth and one for 100 MHz bandwidth.  The choice is made by the firmware.  If it thinks it is a 50 MHz scope, it chooses the 50 MHz filter.  If it thinks is is a 100 MHz scope it chooses the 100 MHz filter.  That's why tweaking the scope settings make a difference.
Set to 100 MHz, the scope has been tested and found to have rather more than 100 MHz of bandwidth to -3 dB.  I don't recall the number but it's up around 150 MHz.
Offline ebclr
« Reply #3 on: October 31, 2017, 03:26:28 pm »
1054Z hacked go far away from 100 Mhz, at least 130 Mhz without attenuation, and +- 200 Mhz with attenuation
Online Fungus
« Reply #4 on: November 02, 2017, 05:43:01 am »
Yes.
The 3dB point is usually around 130MHz (as measured by several people here).
Online Fungus
« Reply #5 on: November 02, 2017, 05:45:16 am »
You can still see 300MHz+ signals on one, although very attenuated, obviously.
Offline rhb
« Reply #6 on: December 20, 2017, 11:49:27 am »
How about a screen shot of a sweep generator feeding a "hacked" scope? Has anyone posted that?
Online xrunner
  • hp>Agilent>Keysight>?
« Reply #7 on: December 20, 2017, 12:20:56 pm »
Fstart ?
Fstop ?
I am a Test Equipment Addict (TEA) - by virtue of this forum signature, I have now faced my addiction
Online rstofer
« Reply #8 on: December 20, 2017, 12:52:58 pm »
Fstart ?
I suppose something less than 50 MHz for those who don't think the original bandwidth has changed.  Say 10 MHz?
Fstop ?
No point in going beyond 200 MHz because it is pretty well establised that the -3dB point is around 130 MHz.
I have no idea if those values are realistic...
Online xrunner
« Reply #9 on: December 20, 2017, 02:10:28 pm »
I don't strictly have an RF sweep generator, but I might be able to cobble together something, if another member doesn't beat me to it.  :)
Offline kahe40
« Reply #10 on: December 21, 2017, 01:40:49 am »
here a  sweep 50M to 500M with 10M step, you see those spikes every 10M.
Rigol is not fast enough, but Micsig and Siglent show them clearly.
Center of Grid is around 225M, so you get an idea of BW and Freq.
Micsig goes down very fast, Siglent is the best of all
and Rigol has a linear degree, like a triangle.
Rigols 5ms are too short for the entire sweep (or user error?),
but I tried to make it comparable.
Sweep was done with RF-Explorer-SigGen,
yes, maybe it is a toy but better than nothing...
Micsig TO1104   https://ibb.co/hTupS6
Rigol 1074Z hack  https://ibb.co/mf2b76
Siglent 1202XE   https://ibb.co/cMZCfR
edit: have not found out, how to post pictures with iPad, sorry
« Last Edit: December 21, 2017, 01:45:58 am by kahe40 »
Offline bingo600
« Reply #11 on: December 21, 2017, 04:18:33 am »
I measured mine to around 134Mhz , for the 3dB mark
DS1054Z (100Mhz Hacked)
/Bingo
Online xrunner
« Reply #12 on: December 21, 2017, 06:03:46 am »
Here's my data. First up is from 10 MHz to 150 MHz. The -3dB point is where the cursors on the right intersect, about 130 MHz.
Second is from 10 MHz to 500 MHz, again where the cursors intersect is -3dB @ 130 MHz. But certainly the scope can easily display signals in excess of 500 MHz.
Note: Using TG output of a Rigol DS815 triggered externally, some amplitude anomalies are expected. The TG output was set to 0 dBm into 50 ohms at the scope.
Offline rhb
« Reply #13 on: December 21, 2017, 12:31:02 pm »
The Rigol has  a very strange response.  It looks like a filter  skirt with the corner frequency below 50 MHz.
Online Fungus
« Reply #14 on: December 21, 2017, 04:38:06 pm »
Rigol is not fast enough, Rigol has a linear degree, like a triangle.
What sort of settings are these?
Rigol has 1GSa/s and 24M pts.
Just a display artifact from the settings, nothing to do with frequency response.
« Last Edit: December 21, 2017, 04:41:04 pm by Fungus »
Online Mechatrommer
  • Country: my
  • reassessing directives...
« Reply #15 on: December 21, 2017, 06:23:05 pm »
no, beyond 500MHz, you will see aliased signal...
if something can select, how cant it be intelligent? if something is intelligent, how cant it exist?
Online xrunner
« Reply #16 on: December 22, 2017, 12:38:34 am »
Yes, if you don't care about amplitude or frequency accuracy. 900 MHz signal shown below, is displayed quite nicely. The amplitude and frequency you would measure with the scope will be wrong, but that's not the point.
« Last Edit: December 22, 2017, 12:44:27 am by xrunner »
Offline scopeman
« Reply #17 on: December 22, 2017, 02:34:10 am »
Hello,
You should re-run this test with a known good 50 ohm feed through termination and start with a 50KHz signal with six divisions of
signal and increase the frequency to the point where you have 4.2 divisions of signal and repeat this for each attenuation setting to get a better idea of what the bandwidth of the scope itself is.
Sam
"scopeman"
Online Mechatrommer
  • Country: my
« Reply #18 on: December 22, 2017, 02:48:07 am »
thats called 100MHz garbage
if the point is try to make out something from garbage, then yes i get your point.
Online xrunner
« Reply #19 on: December 22, 2017, 03:05:02 am »
You aren't understanding the point I'm making. It's not garbage - it's a signal, it's useful information. If you want to trace a signal on a board to see if it exists then it's just fine for that above ~130 MHz. The signal is very easy to view on the DS1054Z hundreds of MHz higher than it's bandwidth limit (which is simply the -3dB point - not where it completely stops working).  :-//
Offline rhb
« Reply #20 on: December 22, 2017, 05:46:56 am »
To paraphrase Harry Calahan, "A man's got to know his scope's limitations."  Unfortunately, many people are held  by economic considerations in a state of ignorance.  The linear loss of amplitude on the Rigol, *is* a serious issue if you're trying to measure the rise time of a pulse.
Thank you to those who have provided screenshots.  I'd like to suggest, though, that you document the signal generator used.  There are two terms in the equation.
It would be very interesting to see a head to head of a hacked DS1054Z vs a DS1104Z using a top tier signal generator.