I received the scope today via UPS. US mail tracking shows both sets of probes were delivered on Saturday. However, I didn't get the probes. Either the busy mailman put them in the wrong box, or I was hit by Christmas-time rural mailbox thieves. I suspect the latter, so I ordered new probes and will keep a closer eye on my mailbox this time around.
The scope was packed surrounded by newspaper. For an item this heavy and fragile, I personally would only use large bubble wrap. Sure enough, I removed the scope from the box and found one of the rear feet had been broken off during shipment and was laying in the bottom of the box.
I plugged the scope in and turned it on, and a horrible noise emanated from the fan area. It sounded like either the fan was rubbing or the bearings were going out. Fortunately, it turned out to be the fan rubbing on the grill (probably from the rough trip across the US). So I fixed it and then hooked wires between the BNC connectors and a signal generator in order to test the function of the scope. Nearest I can tell, all buttons, knobs, sweeps, lights, switches, delays, etc. appear to be working fine, and the scope is in very nice condition.
The only thing that appears odd or off about the scope is the test loop on the front of the scope only measures 157mv (it's rated at 300mv) with my Fluke DVM (verified with my cheap DVM). I notice my HP scope also measures a low calibration output voltage. Is this common with older scopes. Is it a sign that the scope is out of calibration?
Thanks for your input!