Thank you guys
@Damianos , It worked thanks a lot. It would be great if you could explain to me why did I choose those triggering settings.
Some explanation:
- Vertical sensitivity: 2V/div.
- Vertical position: 3 divisions lower from the center.
These two are to place the signal on the screen in a way to have a big vertical image. Play also with the "fine adjustment" of the vertical sensitivity.
- Signal and trigger coupling: DC.
... because the signal is slow.
- Trigger channel: the same that is used for the measurement.
Using this it is not needed a separate connection for the trigger...
- Trigger mode: Single.
By selecting this, there is only one scan, so at the end of it you have a stable image of the signal to make your measurements. Note that the signal is not repetitive.
- Trigger type: edge.
- Trigger direction: up.
- Trigger level: "a few millivolts above zero".
We want the signal capture to start when it begins to increase from zero.
- Horizontal scale: 1sec/div.
- Horizontal position: near the left side of the screen.
As already mentioned, the T=RC is about 2.2 seconds and the phenomenon ends after a few times this amount. So by this adjustment we try to have the entire progress on the screen. To have more details of parts of the signal, try to use the "horizontal zoom" functionality of the oscilloscope.
- While the scope says that is waiting, close the switch
I don't think that this needs an explanation!
Regards,
Damianos