Author Topic: EEVblog #470 - Agilent N9344C 20GHz Spectrum Analyser Teardown  (Read 27853 times)
Offline nitro2k01Topic starter
Offline Laidukas
Re: EEVblog #470 - Agilent N9344C 20GHz Spectrum Analyser Teardown
« Reply #1 on: May 15, 2013, 06:37:20 am »
Dave, were you wearing a wrist strap for ESD protection during this tear down? I think Agilent guys hope you did  :)
Offline babi
« Reply #2 on: May 15, 2013, 07:22:10 am »
Hi Dave
Did you get any high-res photos of the RF micro-strips on the PCB?
Offline EEVblog
« Reply #3 on: May 15, 2013, 07:51:31 am »
Some photos on my Flickr site, but just 2MP taken with the video camera.
Offline JoeMuc2013
« Reply #4 on: May 15, 2013, 09:21:48 am »
Hello Dave,
what a beauty once again. Just one thing I didn't like is the odd placement of the L2 inductor on the input board at 39:53 in the video (left side) and also shown in some of the closeups at Flickr. I guess there is no "RF" reason for doing this? Just a small flaw I'm not used to see from Agilent. Well, it appears to be working though, right?
Did you manage to put it back together? :)
Greetz
Joe
Offline icon
« Reply #5 on: May 15, 2013, 12:02:24 pm »
"Looks like a phillips, smells like a phillips" - but isn't. Is therefore, possibly, a Japanese Industrial Standard screw? Found in laptops and other consumer electronics, on the adjustments of Shimano derailleur mechs, on RC helis, on the lids of Nissin motorcycle brake fluid reservoirs, etc. I have a set of Hozan JIS screwdrivers that get a lot more use than you'd think. A phillips screwdriver will rapidly knacker a JIS head, especially as they seem often to be made of some kind of silvery cheese.
John
Online SeanB
« Reply #6 on: May 15, 2013, 12:08:51 pm »
JIS is easy to identify, the head has a small depression on the one side to indicate that it is JIS.
Offline KJDS
    • my website holding page
« Reply #7 on: May 15, 2013, 12:27:58 pm »
Circuit design by Slartibartfast. Just as RF should be.
Offline saturation
  • Doveryai, no proveryai
    • NIST
« Reply #8 on: May 15, 2013, 12:39:12 pm »
GHz RF, always a blast to see layouts, so black magic.  :-+
Best Wishes,
 Saturation
Offline free_electron
    • SiliconValleyGarage
« Reply #9 on: May 15, 2013, 01:31:50 pm »
Could very well be. He always loved the fiddly bits like the fjords of norway...
The reason for having gps onboard this machine is that it most likely is a site analyser. These are used to measure field strength for things like wifi and cell phones. Hook it up to a pc with mapping software and drive around. And it builds you the power map.
Offline EEVblog
« Reply #10 on: May 15, 2013, 01:38:07 pm »
Yes, its inclusion was driven from customer requirements for that.
Offline icon
« Reply #11 on: May 15, 2013, 01:55:13 pm »
JIS is easy to identify, if the head has a small depression on the one side to indicate that it is JIS.
Edited.
I've often encountered 'phillips' (ie not Pozidriv type) screws with no markings, but which no size of phillips screwdriver will properly fit. JIS usually do the job (if I haven't rounded the damn thing in the process of discovery - then it's Dremel-time).
John
Online mikeselectricstuff
    • Mike's Electric Stuff
« Reply #12 on: May 15, 2013, 02:23:23 pm »
Can it use it for a frequency reference I wonder?
This reminds me of a story I heard a while ago - in the early days of analogue cellphones, Vodafone had some vans which used to drive around to evaluate call-site performance - because of the test gear they were using, they had to drive round fairly slowly, and on a couple of occsions got pulled over by cops thinking they were kerb-crawling - apparently they were persuaded otherwise by all the gear in the back!
Day Job: Mostly LEDs
Offline lewis
  • Nullius in verba
« Reply #13 on: May 15, 2013, 02:49:37 pm »
Voodoo black magic shit.
I will not be pushed, filed, stamped, indexed, briefed, debriefed or numbered.
Online SeanB
« Reply #14 on: May 15, 2013, 03:01:28 pm »
Dremel not always going to work, but you can get left hand cutting drill bits, or my favourite item, a Ridgid extractor kit. The left hand drill bits both drill and loosen the screw, great in a deep recess. Just need a reversible drill, and tighten the chuck pretty tight.
Offline senso
  • Country: pt
    • My AVR tutorials
« Reply #15 on: May 15, 2013, 03:29:17 pm »
There are some places riddled with what I would call sloppy soldering, and those botches, what a crap job..
And there is lots of flux residues, I was watching and screaming, like how can you call that top notch quality? |O
Offline BravoV
  • +++ ATH1
« Reply #16 on: May 15, 2013, 03:38:20 pm »
Probably that is a working unit but with the calibration way out of whack big times beyond repair, and also might be as a target practice for noob repair/cal staffs to play with ? :-//
Offline senso
  • Country: pt
    • My AVR tutorials
« Reply #17 on: May 15, 2013, 04:04:39 pm »
Some examples, there is more top notch quality showing all over those boards, it seems really smart having expensive controlled dieletric pcb's and then just apply some dabs of random flux all over the impedance controlled traces, must be the secret sauce  ::)
Online SeanB
« Reply #18 on: May 15, 2013, 04:12:03 pm »
Those little white chips are Mini circuits amplifiers, some are capable of a flat response from close to DC to close to light. Give away is the power feed, a feedthrough ceramic cap, a chip inductor, a decoupling cap to ground then a feed resistor to the output of the chip. The numbers are an indication of the maximum signal level they can linearly handle. On the input PCB the funny thing with 2 lines next to each other and a resistor on the far end is a sampling loop, probably going to a simple rectifier and a signal strength indicator, though it probably has a connection to the gain stages so as to give a constant level to the ADC stages as it is scanned through the passband.
Offline Corporate666
  • Remember, you are unique, just like everybody else
« Reply #19 on: May 15, 2013, 07:50:16 pm »
Always interesting to see PCB design at a level light years beyond what I do (or ever will do).
Obviously these guys get special training to be able to do that - they aren't just first year EE grads doing those layouts.  Is high-end RF design something that is taught at the masters level, or (as I would guess) based more on in-house knowledge that is passed from generation to generation within a company like Agilent?
Anyone else notice the Chinglish on the stickers?  Left Borad and Right Borad?  :-DD
It's not always the most popular person who gets the job done.
Offline Watermelon
« Reply #20 on: May 15, 2013, 08:13:14 pm »
Does anyone has a datasheet available for those odd 4 leaded RF amplifiers? Cannot find any at first glance  :-//
Offline Hydrawerk
« Reply #21 on: May 15, 2013, 08:15:56 pm »
Did you reassemble this insturment? Does it work again?
« Last Edit: May 15, 2013, 11:23:18 pm by Hydrawerk »
Offline KJDS
« Reply #22 on: May 15, 2013, 08:29:38 pm »
Decent RF design is mostly learnt on the job. My bachelors and masters degrees both had RF projects which provided some basic understanding, but I didn't really know what I was doing until I'd spent ten years or so learning on the job.
I remember one of my undergraduate courses had 20 hours of lectures on the Smith chart and I barely understood how to do something basic with it, whereas now it's second nature. That only comes with spending a lot of time on it and that's just a tiny part of RF. A lot of RF engineers are specialists, so will only play seriously with say LNAs or high power or filters or mixers rather than being bale to readily pick up things from a broad spectrum of the building blocks used in RF. There are certainly very few that are genuinely good at say, both high power RF amplifier design and cavity filter design, certainly if they had to do something from first principal rather than relying on synthesis software.
Offline krenzo
« Reply #23 on: May 15, 2013, 08:34:44 pm »
Look for "4-Micro-X" as the Package / Case.  There are a few on DigiKey that do 0-12 GHz amplification from RFMD: http://www.digikey.com/product-detail/en/NBB-310-T1/689-1001-1-ND/1302337
This one has "N4" on the package like in the picture above.
« Last Edit: May 15, 2013, 08:39:51 pm by krenzo »
Offline robin_
« Reply #24 on: May 15, 2013, 10:28:08 pm »
I’m wondering what these bodge wires are for. What kind of mistakes in the board layout remain unnoticed up until the board is already in production? And in which kind of situations is it even possible to still being able fix it with hacking in small wires afterwards?
« Reply #25 on: May 15, 2013, 11:02:05 pm »
It isn't always mistakes - for example it can be to adapt to use different parts if the original isn't available.
Offline EEVblog
« Reply #26 on: May 16, 2013, 01:47:50 am »
Yes, seems to work fine.
Offline Dongulus
« Reply #27 on: May 16, 2013, 03:21:36 am »
Top-notch session of PCB and systems engineering porn.
I was confused by Dave's comment about how the 10 Mhz reference oscillator would not work after being turned over. I've never worked TCXOs before so I had no idea that they were sensitive to orientation. How can a robustly designed instrument intended for potentially rough field work be expected to operate containing such a component?
Offline c4757p
  • adieu
« Reply #28 on: May 16, 2013, 03:31:52 am »
I suspect it was a joke, similar to his usual "the electrons will fall out!"  :)
Offline EEVblog
« Reply #29 on: May 16, 2013, 04:04:52 am »
 ;D
Offline edpalmer42
« Reply #30 on: May 16, 2013, 04:14:25 am »
Not to mention the 'right tongue angle'.
Wait, I mentioned it, didn't I?   |O
Ed
Online SeanB
« Reply #31 on: May 16, 2013, 04:59:34 am »
I suspect the bodge wires connecting ground planes had more to do with a faulty via only detected during final test as being either open or intermittent, probably going open during the IR reflow of the board. Would have been better to have used a thicker wire, but hopefully the voltage regulator is happy with the added ground inductance.
Offline ftransform
« Reply #32 on: May 16, 2013, 08:08:57 am »
unsolder that gold shielding you coward, I would be proud of you then, like I am proud of war journalists.
put it on a hot plate... agilent said to take it apart......... common I know you want to.... :P
« Last Edit: May 16, 2013, 08:10:36 am by ftransform »
Online SeanB
« Reply #33 on: May 16, 2013, 08:18:14 am »
The copper tape over the hole is a vent for reflow soldering , so the case does not collapse from vacuum after soldering, or blow out the solder during reflow. The copper adhesive is most likely a conductive adhesive.
Offline bitwelder
« Reply #34 on: May 16, 2013, 05:13:58 pm »
What's the purpose of the arrow on top of the gold shielding? Reflow direction?
Offline Watermelon
« Reply #35 on: May 16, 2013, 07:34:49 pm »
Interesting read, thanks! Now i understand why they don't need an external supply, it's just 2 transistors and some resistors... :palm:
Offline pomonabill221
« Reply #36 on: May 16, 2013, 08:30:45 pm »
Or the electrons will all roll to one side of the can when the board is rotated... :-DD :-DD
Offline lewis
  • Nullius in verba
« Reply #37 on: May 16, 2013, 08:32:58 pm »
So you know which way to orientate the units so the electrons don't fa..... bugger it.
Online SeanB
« Reply #38 on: May 16, 2013, 08:48:21 pm »
Some have a lot more than the 2 transistors inside, you can make a pretty complex amplifier if it needs to be level from near Dc to light, and has only 2 pins available. Those square ones must be a lot more, even if they have the same power arrangement.
Offline KJDS
« Reply #39 on: May 16, 2013, 09:05:27 pm »
Having once worked at a test gear manufacturer I can give some insight into why there'll be a few bodges on the boards.
The early demonstrator models are part of the initial pre-production run. These are built and tested using the normal production facilities, but with design support. Some of these will then be used for the qualification testing, full temperature tests, both operational and storage, vibration, EMC and a whole load more detailed testing that would be carried out in production. Sometimes these tests bring up issues that require minor mods to a board and this will often be replicated across most of the batch. It's not unusual for the first couple of units to have components changed by designers who are not expert solderers just to prove that the mod works as intended. Once everyone is happy then some of these units become demo models, others will be used for accelerated and highly accelerated life tests and production will be allowed to ramp up, sometimes carrying on with minor board mods and sometimes waiting for a board respin and any associated requalification.
Offline ftransform
« Reply #40 on: May 17, 2013, 02:00:42 am »
I figured out how to remove the can nicely. Wrap it with resistance wire and heat it. Probably more safe for the board then a hot plate.
Aint anyone else dying to know what the fuck is under those cans?
dave should take some risks... he has it made.....
Offline larry42
« Reply #41 on: May 17, 2013, 02:44:28 am »
Under those cans? Local oscillators and the front end attenuator.
If you have an animated GIF in your avatar or signature then I reserve the right to think you're a dolt.
Offline ejeffrey
« Reply #42 on: May 17, 2013, 02:55:25 am »
Flux on the PCB has essentially zero effect on the impedance.  For a microstrip geometry the capacitance is almost entirely due to fields inside the dielectric not above it -- which is the whole point.  A small fraction of the capacitance does come from the field above the board, but those field lines are relatively long, so a thin layer of dielectric crap won't really affect it.
Offline jancumps
  • New Low
« Reply #43 on: May 17, 2013, 04:15:09 pm »
If it would affect it, it wouldn't be there.
Offline kfitch42
« Reply #44 on: May 18, 2013, 03:56:16 am »
But, how many parts were left over? :)  IIRC you said there were 2 million parts ... so if you had two 'extra' screws at the end that would be 1 PPM, right? That sounds pretty good when you put it that way :)
Offline thewyliestcoyote
« Reply #45 on: October 26, 2014, 12:10:22 am »
I love the tear down! Really shows all the RF design magic that is hard to come by.
I have a couple of 4 port +50GHz PNA's, older VNA's, and about 2 dozen field foxes at work and would look to see what they are in the insides and not get fired for destroying a instrument that costs more then I make in a couple of years.
Dave maybe you could do a teardown of a PNA, field fox, or the PXI unit for a 26.5 GHz VNA?
Offline rfbroadband
« Reply #46 on: October 26, 2014, 03:34:21 am »
at 20:25 its is a 1:1 125 MHz RF balun not a relay...
very nice tear town.