Just like the 3458A, I think the "autocal" feature is great, but it's no substitute for ultra high quality components. ..The Fluke 8508A is accurate to full specs +/-5C from Tcal, and this is *without* any "autocal" feature.
I think, i already posted this earlier in this thread..
The AUTOCAL of the 34470a is really nothing compared to the 3458a. It also has to be calibrated on each mode and each range, whereas the 3458a only needs 10V and 10kOhm references.
Simple reason: The 34470a only is linear to 1ppm, but the 3458a to 0.02 ppm, so precise 10:1 Transfers are possible.
The 8508a only lives from its many ultra stable Vishay bulk metal foil resistors.
Its linearity is not as good, as the 3458a, so it needs the usual > 34 Cardinal Point referencess from a calibrator, plus additional ratio standards..
If you would use better resistors for the 3458A, it would outperform the 8508a, of course.
Stable thermal environment is needed in any case for analogue metrology.
To my opinion, Keysight is not capable anymore to design such an improved sophisticated instrument..brain drain.. Otherwise the 34470a would've been better.. worth the 7 1/2 digits..