Author Topic: WTB: Attenuator for HP 8648C (part 33322-60011)  (Read 2127 times)
Offline davekeckTopic starter
« on: January 17, 2018, 01:11:12 pm »
Hey all,
The attenuator in my HP 8648C signal generator died so I'm looking to replace it. The part number is 33322-60011. Ebay has a few but they're priced as much as I paid for the whole signal generator, so I thought I'd check here.
Thanks!
Re: WTB: Attenuator for HP 8648C (part 33322-60011)
« Reply #2 on: January 18, 2018, 07:23:39 pm »
Interesting, do you know if the 33322Z attenuator would be compatible with the 8648C? It looks like it needs 5 volts instead of 24 volts, and the connector is a different size. But if the IO signals are the same it seems like that would work if I routed a 5V supply to it.
« Reply #3 on: January 20, 2018, 11:14:15 am »
After more research I found that 33322-60011 is the only attenuator that attenuates up to 130dB using 10, 20, 30, and 60 dB sections. So using any other attenuator would result in discontinuities when sweeping the signal amplitude since the attenuation that the 8648C tries to select wouldn't match what the mismatched attenuator gives.
With that in mind I successfully repaired my 33322-60011 thanks to a series of small miracles. For those Googling, the problems with the attenuator were:
Contact fingers not mating well with the bypass pads / attenuator pads
   Solution: clean pad mating surfaces with a q-tip and rubbing alcohol until a good connection is observed with an ohmmeter. [1]
Plunger o-rings missing / deteriorated
   Solution: ideally buy new o-rings, but being impatient I used a small amount of hot glue to secure the plungers with bad o-rings to their respective metal plates. Totally amateur but it worked. [2]
Solenoid wire was severed
   Solution: re-solder the very fine green wire back to its respective post
[1] Take a look at page 9 of this PDF showing attenuator internals. Note the gold flaps between each bypass surface, and how the ends of the gold flaps make contact with the bypass pad:
   http://ve2azx.net/technical/HP_StepAttenuatorRepair.pdf
[2] This video goes into detail regarding the o-rings: