How to *NOT* replace a fuse in a multimeter
« on: February 23, 2012, 02:37:31 am »
Today I heard a strange sound in a Fluke179 (NOT MINE!!!). I opened it and found this. Someone *changed* the fuse and tried to *make it fit well*. I'm not sure if I should now  :)  or  >:(
 

Offline PedroDiogo

  • Regular Contributor
  • *
  • Posts: 62
  • Country: pt
  • EE Student
Re: How to *NOT* replace a fuse in a multimeter
« Reply #1 on: February 23, 2012, 02:47:25 am »
Still, it's better than soldering a wire between the contacts I guess :)
 

Offline Short Circuit

  • Frequent Contributor
  • **
  • Posts: 439
  • Country: nl
Re: How to *NOT* replace a fuse in a multimeter
« Reply #2 on: February 23, 2012, 03:04:15 am »
Odd way of fitting a standard fuse  :o
Bending the clips is easier and more reliable.

(Just don't use the DMM in high current systems)
 

Offline tamtamTopic starter

  • Newbie
  • Posts: 8
  • Country: ch
Re: How to *NOT* replace a fuse in a multimeter
« Reply #3 on: February 23, 2012, 03:11:51 am »
Still, it's better than soldering a wire between the contacts I guess :)
Hm ... I guess not. The disc can move to the other two terminals (Volt and COM) and make a short while you measure the voltage. In the factory where I found this, they have a lot of big drives (up to 700kW) and some other big consumer (HF welding). So the next fuse is maybe 200A, 400A, 800A or even higher by 400V or more (some drives have higher voltage).

Or ... If you have such friends, you don't need enemies.
 

Offline Short Circuit

  • Frequent Contributor
  • **
  • Posts: 439
  • Country: nl
Re: How to *NOT* replace a fuse in a multimeter
« Reply #4 on: February 23, 2012, 11:30:25 am »
Indeed  :(  Even applies to general EE use. Wouldn't be very happy to kill a circuit because some
dickhead upgraded a DMM with some metal parts inside.  >:(

On the subject of replacing fuses with generic types, I figured I check the ratings of these fuses. And it
appears these low-cost ceramic anti surge fuses can handle quite a bit of overload as well;

Original 400/440mA fuses have a 1kV rating, so they have a surge handling capability of 2.5kA are rated for 10kA.
A ceramic 500mA fuse like http://nl.rs-online.com/web/p/cartridge-fuses/7211597/ is rated for 1500Amps.
Not quite the same, but acceptable for most uses. And at 50 cents vs 10-30 bucks for those 'real' fuses...

But when working on such equipment, I'd rather want to stick with the right fuses. You really
don't want a DMM exploding when there is a chance of probing around live circuits carrying such
voltages and currents.
« Last Edit: February 23, 2012, 12:00:13 pm by Short Circuit »
 

Offline Bored@Work

  • Super Contributor
  • ***
  • Posts: 3932
  • Country: 00
Re: How to *NOT* replace a fuse in a multimeter
« Reply #5 on: February 23, 2012, 11:41:43 am »
Or ... If you have such friends, you don't need enemies.

Time to look for new friends.
I delete PMs unread. If you have something to say, say it in public.
For all else: Profile->[Modify Profile]Buddies/Ignore List->Edit Ignore List
 

alm

  • Guest
Re: How to *NOT* replace a fuse in a multimeter
« Reply #6 on: February 23, 2012, 11:41:55 am »
I actually expected the bolt belonging to these washers to be in there instead of the smaller fuse. That would have fixed that annoying blown fuse problem.

Original 400/440mA fuses have a 1kV rating, so they have a surge handling capability of 2.5kA.
Why would a fuse with a 1 kV rating have a surge capability of 2.5 kA? Fluke specs 10 kA. The ones I have here are the same, so it's not a recent spec change.

A ceramic 500mA fuse like http://nl.rs-online.com/web/p/cartridge-fuses/7211597/ is rated for 1500Amps.
Not quite the same, but acceptable for most uses. And at 50 cents vs 10-30 bucks for those 'real' fuses...
Make sure to label the instrument as 'not for mains / high energy / dangerous voltages, only for SELV'. I don't see the issue with using smaller ceramic fuses for SELV measurements, but I would be hesitant to put them in an instrument labeled 'CAT IV 600V', just because of the confusion it may cause.
 

Offline Short Circuit

  • Frequent Contributor
  • **
  • Posts: 439
  • Country: nl
Re: How to *NOT* replace a fuse in a multimeter
« Reply #7 on: February 23, 2012, 11:57:03 am »
Why would a fuse with a 1 kV rating have a surge capability of 2.5 kA? Fluke specs 10 kA. The ones I have here are the same, so it's not a recent spec change.
You're so right. That figure doesn't make sense at all  :-[
 

Offline Tube_Dude

  • Regular Contributor
  • *
  • Posts: 115
  • Country: pt
Re: How to *NOT* replace a fuse in a multimeter
« Reply #8 on: February 23, 2012, 02:58:35 pm »
Hi Tamtam

Think positive, just bend the clips, and you have earned 4 nice washers in the process...  8)
Jorge
 

Offline G7PSK

  • Super Contributor
  • ***
  • Posts: 3861
  • Country: gb
  • It is hot until proved not.
Re: How to *NOT* replace a fuse in a multimeter
« Reply #9 on: February 23, 2012, 06:16:24 pm »
Why can't these meters have a trip like they used to have on the old AVO's then there would be no problem with fuses.
 

Offline Zero999

  • Super Contributor
  • ***
  • Posts: 19527
  • Country: gb
  • 0999
Re: How to *NOT* replace a fuse in a multimeter
« Reply #10 on: February 23, 2012, 08:37:49 pm »
Why can't these meters have a trip like they used to have on the old AVO's then there would be no problem with fuses.
Because I doubt the breaking capacity would be anywhere near as high as the type of fuse used in a Fluke.
 

Offline Pentium100

  • Frequent Contributor
  • **
  • Posts: 258
  • Country: lt
Re: How to *NOT* replace a fuse in a multimeter
« Reply #11 on: February 24, 2012, 07:43:19 am »
This is why you should always try to use a resettable fuse/circuit breaker instead of the regular one. Sooner or later, the fuse is going to blow, the owner, not having a spare, but needing the device working "this instant" will replace it with a piece of wire.

Some time ago where I live we didn't have circuit breakers on the mains - just regular fuses - I bet at least 50% of them were  replaced by wires. Now with resettable circuit breaker sthere is no need.
 

Offline wkb

  • Frequent Contributor
  • **
  • Posts: 910
  • Country: nl
Re: How to *NOT* replace a fuse in a multimeter
« Reply #12 on: February 24, 2012, 09:01:51 am »
This is why you should always try to use a resettable fuse/circuit breaker instead of the regular one. Sooner or later, the fuse is going to blow, the owner, not having a spare, but needing the device working "this instant" will replace it with a piece of wire.

Some time ago where I live we didn't have circuit breakers on the mains - just regular fuses - I bet at least 50% of them were  replaced by wires. Now with resettable circuit breaker sthere is no need.

A resettable breaker with the 'break' capabilities of the fuse in the Fluke would probably be the same size as the whole Fluke itself..
 

Offline Zero999

  • Super Contributor
  • ***
  • Posts: 19527
  • Country: gb
  • 0999
Re: How to *NOT* replace a fuse in a multimeter
« Reply #13 on: February 24, 2012, 05:55:08 pm »
Maybe they could use both?

A resettable breaker with a lower tripping current than the fuse. In the case of most low current overloads the breaker will just trip and can be easily reset but when a large current causes the breaker to arc over, the fuse blows before any damage is done to either the breaker or the meter.
 

Offline paul

  • Contributor
  • Posts: 37
Re: How to *NOT* replace a fuse in a multimeter
« Reply #14 on: February 24, 2012, 08:44:56 pm »
I found this fuse `repair` (fuses fused together?) inside a Fluke 189 at work when I replaced the batteries  :(  no one owned up  :) although it must of taken them a while to solder up.
However I am glad, make that very glad, I spotted it before working on some mains equipment or a three phase supply. 
« Last Edit: February 24, 2012, 08:59:02 pm by paul »
 

Offline SeanB

  • Super Contributor
  • ***
  • Posts: 16284
  • Country: za
Re: How to *NOT* replace a fuse in a multimeter
« Reply #15 on: February 25, 2012, 09:20:04 am »
I had some machinery that used Diazed fuses, which are rated for 10kA breaking. Replaced with  minature circuit breakers of similar rating, as I got annoyed with fuses popping randomly on switch on as they aged, and with replacements costing more than the mains rated breaker. As a bonus I got 3 phase tripping, which made things easier as a blown phase fuse could give strange faults. I later upgraded the supply side to add a 3 phase ELCB, and had to replace industrial heater elements that had developed non tripping leakage to the housings. Extra protection is always good.

I did see a Fluke 77 that was used by Eishkom worker to measure a 11kV line. The fuse survived, along with the battery and the display. The inside of the case had a lovely copper coating, and a PCB that had absolutely no copper traces, except for the solder pads, in the lower section. That came in as a warranty claim.


The lift at work still has Diazed fuses, and I used the spare fuses I had to replace those that had been "modified" by assorted lift mechanics if the blew, so as not to have fuses with external wire elements.
 

Offline wkb

  • Frequent Contributor
  • **
  • Posts: 910
  • Country: nl
Re: How to *NOT* replace a fuse in a multimeter
« Reply #16 on: February 25, 2012, 10:20:33 am »
I found this fuse `repair` (fuses fused together?) inside a Fluke 189 at work when I replaced the batteries  :(  no one owned up  :) although it must of taken them a while to solder up.
However I am glad, make that very glad, I spotted it before working on some mains equipment or a three phase supply.

Not too surprising given that these bloody fuses are typically not on stock and the absolutely ridiculous prices charged for them..

I would not care working on household mains with this Fluke, there are no kA in sight there (more like 16A in systems in NL).
 

Offline paul

  • Contributor
  • Posts: 37
Re: How to *NOT* replace a fuse in a multimeter
« Reply #17 on: February 25, 2012, 07:53:47 pm »
After watching Dave`s video of cheap meters (with glass fuses for protection) going bang I want the right fuse in my meter.
I would not want that fuse protecting the meter whilst I am trying to measure something in a tight space with the meter right next to my face.
This meter has been used to check three phase supplies and switch gear also logging charging voltages on 200Ah lead acid backup batteries, so if I do something dumb with the meter I hope that someone else hasn't done something dumber by `repairing` the fuse and putting it back in the cupboard for me to use.

However I can understand it having to be bodged if you are desperate and dont have a replacement fuse.

And you are right, Fluke fuses are expensive :(
« Last Edit: February 25, 2012, 07:56:51 pm by paul »
 

Offline Randall W. Lott

  • Regular Contributor
  • *
  • Posts: 180
  • Country: us
Re: How to *NOT* replace a fuse in a multimeter
« Reply #18 on: February 28, 2012, 02:53:12 am »
At that rate, you might as well replace the fuse while measuring mains voltage.  Haha, I'm just joking.