Hello there,
I have bought an Rigol DS1052E. I reckon there will be a lot of users of this oscilloscope on the forum, so this must be the best place to ask.
Is there a way to adjust the amount of data collected when the scope is triggered? I feel that if the signal was longer (in regards to time), then that would help me.
I have tried the single sweep mode, and is not very familiar with the features of the scope.
Thank you for my time.
Offline Psi
Re: Rigol DS1052E: Captured signal isn't long enough
« Reply #1 on: September 29, 2011, 09:58:38 am »
Under the Acquire menu there is a memory option, Normal or LongMem.
Normal is 8k/16k  and longmem is 512k/1024k  (depending if you're using one or two channels)
I think the Longmem mode will give you more samples.
« Last Edit: September 29, 2011, 10:02:17 am by Psi »
  • Country: no
« Reply #2 on: September 29, 2011, 09:59:48 am »
Thanks man! :) M