A little help, when executing ACAL 0 does not execute, executes ACAL (simple). I have checked it manually and through GPIB.
HP3458A, HP3245a, Keithley 2000, Fluke 87V, Rigol DP832, TEK TDS5052B, HP33120A
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: 3458A ACAL 0
« Reply #1 on: November 12, 2016, 12:54:08 pm »
In posts like this one error code or exact description of unit behaviour is a must ;).

My Palantir says you have calibration security activated on the box. Read the manual for commands SECURE and CAL :)

Edit: What you trying to do however? There is no specific ACAL 0.
« Last Edit: November 12, 2016, 12:56:01 pm by TiN »
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline klaus11Topic starter

  • Supporter
  • ****
  • Posts: 156
  • Country: 00
Re: 3458A ACAL 0
« Reply #2 on: November 12, 2016, 01:48:06 pm »
I have tested:
3458A: OPT 1; OPT 2
FW: 9.1 fresh update ;  Docs / xDevs
NVRam ALL: Fresh but keep data (may be the problem)
Test> Pass
Error Display: None
Acal: Pass without errors and without asking for security code
ACAL 0: Does not execute correctly, Manual Calibration picks up CAL 0 to correct offset.
Initial precision: It takes away my dream.  :-+

I am going to test security code and delete memories physically (it is meant not delete Cal data)
« Last Edit: November 12, 2016, 01:53:20 pm by klaus11 »
HP3458A, HP3245a, Keithley 2000, Fluke 87V, Rigol DP832, TEK TDS5052B, HP33120A
 

Offline klaus11Topic starter

  • Supporter
  • ****
  • Posts: 156
  • Country: 00
Re: 3458A ACAL 0
« Reply #3 on: November 12, 2016, 03:14:25 pm »
Security code is not the problem, I have touched a bit and blocked and will not let ACAL 0. I will clean NVRam, even though my eyes are heading to the FW.
HP3458A, HP3245a, Keithley 2000, Fluke 87V, Rigol DP832, TEK TDS5052B, HP33120A
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: 3458A ACAL 0
« Reply #4 on: November 12, 2016, 03:16:09 pm »
I still don't get it.

ACAL 0 is same as ACAL ALL, meaning to run autocal for all functions. It's fully equivalent to running just ACAL without any parameters.

Try to rephrase it, What are you trying to get as result? Zero calibration is executed by CAL 0 and 4W low-thermal short at the inputs , not ACAL 0. CAL 0 required to execute separately at front and rear terminals.
This function did not change external behavior in firmwares since Rev5. I tested myself on Rev5, Rev7, Rev8 and Rev9.
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline klaus11Topic starter

  • Supporter
  • ****
  • Posts: 156
  • Country: 00
Re: 3458A ACAL 0
« Reply #5 on: November 12, 2016, 03:25:00 pm »
Yes, I am a stupid CAL 0   :horse:, a video from YouTube confused me

Thank you TiN. Again a great help.