Hi Splin,
Hi splin,
There is plenty of evidence that old 34401As can have very stable 10V ranges due to the stability of aged LM399s.
what is "very stable" in quantitative terms?
Regards
try
Sorry I can't give you any proper statistical data, but there is anecdotal evidence. See reply #9 for example:
https://www.eevblog.com/forum/testgear/34401a-long-term-accuracy-question/
After 13 years it's still within the 12 month spec on all ranges, and 5x better on most DC voltage, current and resistance ranges.
I am a simple person. I expect ppm terms!
Being within 12 month specs is not that difficult and "5x better on most...", well, "most" does break my calculus.
And from http://www.candlepowerforums.com/vb/showthread.php?418230-Checking-my-DMM-DC-Voltage-Accuracy&styleid=32
What I was extremely lucky on was the Agilent 34401a. Its spec is 15 PPM for one day and 35 PPM for one year. My unit is 15 years old and appears to have the original factory calibration. It passes all self tests and has no reported errors. It had been certified several time but apparently never re-calibrated. I have never observed any error greater than 10 PPM even at turn on. With 15 minutes of warmup the error is usually less than 4 PPM.
Against his Asia-Volt?
In the text you referenced they mention having two references with the same calibration figures. That would make me suspicious.
FWIW, my own 34401A, which hsn't been calibrated for at least 8 years exactly matched, to the uV, that of a cheap Chinese AD584 calibrator which supposedly was measured by the vendor with a calibrated 34401A. I thought that was remarkable but it doesn't prove anything of course.
Also two posters here have LM399 based references which they state have annual drift rates of less than 1ppm.
You are comparing the 10V range of a 34401A against a LM399 based reference (7V or 10V or another output figure?).
This appears like comparing apples and pears to me.
Another post here stated that the LM399 has essentially zero drift when unpowered but I don't know of any tests to confim or deny this. This leaves the 7 to 10V scaling resistor network as the likely main source of reference drift in the 34401A.
What is "essentially zero"? That could have taken from a Vishay datasheet. "0.1" can be considered as "essentially zero" as well.
The best data would come from someone who calibrates these meters for a living but I don't know if we have any posters here who does that and is prepared to provide it.
Anyway, you seem to be the right candidate to join the
maker faire 2018 here in Hannover, Germany on the 15. and 16.9.2018. Going there is my top priority.
It is very easy to get there, something fellow eevblog member Tom from the United Kingdom who came here last year could confirm you.
You can calibrate your device yourself at the PTB booth with the only disadvantage being faced with higer uncertainties due to the uncontrolled environment.
For reports in German language describing the last two events check the following link out and look for "r2d3".
As for r2d3's LM399-based 10V reference you can see on the pictures that it moved 3 ppm from 2016 to 2017 in 16 months. In 2017 it was 10V spot on but that does not say alot given the uncertainty of the 3458A and the conditions involved on the fair in 2017.
During this 16 months r2d3's 34401A moved a lot more (any quantitative terms omitted on purpose to fit the forum style
) than his LM399-based reference.
Both instruments ran 365/24 between the two events. You might consider power-on time as an influencing factor as well.
https://www.mikrocontroller.net/topic/398160Have fun!
Regards
try