In my lab I often repair precision test eqipment. There are cases when repaired DMM needs to receive comprehensive calibration.
Keithley 2000 is a good example. You just can't calibrate one range, You have to follow full AC or DC calibration procedures.
While ago I started looking for a way to expand my lab's capabilities. Best affordable option would be Datron Wavetek 4708.
4708 met my criteria, the parameters were similar to newer Datron 4808.
Panaplex neon displays of 4708 are relatively easy to get by extracting from old Datron 1061...71 DMM (the VFD for Datron 4808 is unobtainium AFAIK).
Keep in mind that these displays may break while shippping.
Back in 2016 there was an event on bidspotter.com called 'HUGE TWO DAY TEST & MEASUREMENT AUCTION ...'. Location: Las Vegas, Nevada. Among many beautiful instruments I felt in love with Datron 4708 equipped in full option.
No further information was available. Until August 2018 when it appeared on ebay. The items description was however inconsistent: 'DATRON 4708 ULTRA PRECISION ACV ACI CALIBRATOR - REV' while photos clearly showed opt. 10, 20, 30, 80. I asked the seller to take a picture of the internals. After few weeks and second email the seller agreed and here is what I received. It looks like we have full option here.
There was one problem: the item is big and heavy (~40kg) so shipping cost to my location in Poland, EU was higher than the price of the item. I started searching for a cheaper option and found a reliable company in U.S. which ships to EU at reasonable prices. Finally I was told that total shipping cost will be approx. 550USD (200USD inside U.S. plus 350USD from U.S to EU). Finally I decided to buy the calibrator.