Yepp, I also followed that intersting thread. That 400k$ is far, far out of my scope, even I also own a 3458A and a Datron 1281.
My thought during reading that thread was:
If I don't need it for cal business (with return of investment), there would be a cheaper way fitting my needs.
F. e. buy a quite good voltage reference (f. e. Fluke 732B, say cost is 5k$) and let calibrate it regular, let say 10 times a year.
Assuming a good qual cal lab ask 1k$, it could be done >300 times for the same amount of money.
if cal is done 10 times a year, that means 30 years cal history data - a good statistical basis.
Regarding the "another method" - had same thougth, that's why I have noted questions to answer:
? does it really need high lineary (my guess not, but good short term stability)
? need good (low) long term drift
I fully agree, the drift (time, temp. humidity) are in any case the main quality criterias for any reference.
As anything is drifting, this "reference" would drift also. Just how much is the question.
If a drift below 2ppm/yr is realistic it would worth to to try.