Anritsu 22NF50 open/short calibration coefficients
« on: May 11, 2017, 02:13:41 am »
Does anyone have the calibration coefficients for the Anritsu 22NF50 Open/Short VNA calibration standard ?

https://configure.us.anritsu.com/catalog/i/22NF50
Facebook-free life and Rigol-free shack.
 

Offline CalMachine

  • Frequent Contributor
  • **
  • Posts: 477
  • Country: us
  • Metrology Nut
Re: Anritsu 22NF50 open/short calibration coefficients
« Reply #1 on: May 11, 2017, 03:30:33 am »
Each individual open/short will have its own unique cal coefficients. 
All your volts are belong to me
 

Offline BudTopic starter

  • Super Contributor
  • ***
  • Posts: 6904
  • Country: ca
Re: Anritsu 22NF50 open/short calibration coefficients
« Reply #2 on: May 11, 2017, 04:09:54 am »
I do not think Anritsu would go into trouble of characterizing individual standards for a commodity grade VNAs. HP did not do that.
Facebook-free life and Rigol-free shack.
 

Offline TheSteve

  • Supporter
  • ****
  • Posts: 3752
  • Country: ca
  • Living the Dream
Re: Anritsu 22NF50 open/short calibration coefficients
« Reply #3 on: May 11, 2017, 06:18:22 am »
Does Anritsu even recommend the use of the 22NF50 for VNA's? I thought that only sold it for scalar use.
VE7FM
 

Offline chris_11

  • Contributor
  • Posts: 48
  • Country: de
Re: Anritsu 22NF50 open/short calibration coefficients
« Reply #4 on: May 11, 2017, 07:13:10 am »
With the outside screw terminals you will need an additional adapter anyway, so I do not see much benefit in accurate delay parameters. You can get those using another cal set with known delays and make a comparison.
 

Offline BudTopic starter

  • Super Contributor
  • ***
  • Posts: 6904
  • Country: ca
Re: Anritsu 22NF50 open/short calibration coefficients
« Reply #5 on: May 11, 2017, 12:55:01 pm »
This is for measuring DUTs with N-female ports, so no external adapters needed if i understood your comment correctly.
Facebook-free life and Rigol-free shack.
 

Offline chris_11

  • Contributor
  • Posts: 48
  • Country: de
Re: Anritsu 22NF50 open/short calibration coefficients
« Reply #6 on: May 11, 2017, 02:53:15 pm »
Here a bit about calibration cal kits. http://www.cel.com/pdf/press/TRL.pdf
Should be possible with your Anritsu too.
 