Humidity is relative easy:
just record %rH and measure your reference output.
Minimum over one %rH cycle (around 9kHrs)
Then you will see some correlation according to the seasonal changes of your local %rH.
Of course the reaction time to %rH is not very fast so you will have to filter the %rH somewhat
(perhaps 5days time constant without the rosin with rosin somewhat more).
By the way: you should work on your resolution. I measured around 0.5ppm/%rH on my 2 LT1027 devices.
(perhaps measure in the 100mV range against another stable reference)
And of course you have to calculate out T.C. and ageing before you can get the humidity coefficient.
with best regards
Andreas