Author Topic: Building your own voltage reference - the JVR  (Read 54925 times)
Offline Alex Nikitin
  • Femtoampnut and Tapehead.
    • A.N.T. Audio
« on: October 04, 2015, 09:20:38 pm »
With so many discussions and data on precision voltage references, I've decided to resurrect my old (about 1993) idea of building my own voltage reference from scratch. At the time I had no means to measure it performance accurately, and even now, my two Keithley 2015 are struggling. The circuit I've built yesterday on my bench is deceptively simple, essentially one JFET and one resistor. I've used one half of U440 dual FET with a rather high (about 4V) cut-off voltage, connected a multi-turn W/W pot in place or R1 and adjusted the pot for a near-zero tempco of Vref output by heating the JFET with a soldering iron and looking at the voltage changes as it cools down back to room temperature. Then I've measured the resistance of the pot and replaced it with a couple of metal film resistors (standard type, 100ppm/C, nothing special). The beauty of this circuit is that the resistor changes reflect on the output voltage at about 1/100 of a value, i.e. 100ppm change in the resistance only creates about 1ppm change in the voltage - and the same applies to the supply voltage, about 1/90 ratio I've measured. I've used LH0070 as the supply regulator to avoid a possible supply influence though. The reference voltage for this particular sample of U440 is 3.75308V at 23C.
I did not "tune" the reference further and the circuit was just sitting on the bench overnight. I've recorded the reference voltage for over 12 hours and for about 5C temperature change the voltage changed from 3.75308V down to 3.75301 at the lowest point and came back to 3.75308 again after the room warmed up back to 23C (making the tempco about 4ppm/C, nicely linear, so it can be tuned better). I've tried to see a thermal hysteresis by first freezing the JFET with a freezer spray and then heating it with a soldering iron for about 5 sec, however the voltage recovered nicely back to 3.75308V after the device cooled down.
So here it is. If somebody with a better than mine measuring capabilities can have a go at this simple arrangement, I will be happy to see the results. I can see one serious problem with this circuit - the voltage reference point would be unique for each JFET sample , so it is difficult to use such a circuit in a production equipment. I will eventially build a proper boxed version with a scaling buffer to get, say, exactly 10V out of it. When I'll do it I'll post an update here.
Cheers
Alex
« Last Edit: October 04, 2015, 11:05:05 pm by Alex Nikitin »
The following users thanked this post: evb149
Offline TiN
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: Building your own voltage reference - the JVR
« Reply #1 on: October 04, 2015, 09:52:30 pm »
Voltage on LM399/LTZ's are also unique, but that's zero problems, as it's calibrated out in software/processing.
Unlucky to me, I don't have LH0070 or U440 to test your circuit idea though.  :-//
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
    • A.N.T. Audio
« Reply #2 on: October 04, 2015, 09:58:21 pm »
I've just used what was available in my lab  8) . You don't need specifically LH0070 - "any" voltage regulator with a better than 100ppm/C tempco will do nicely, and U440 I've chosen for the metal case, again, "any" JFET has a "zero tempco" point, it is just reasonable to use ones with a higher cut-off voltage, i.e J112 may work well for example, though it is in a plastic case.
Cheers
Alex
« Last Edit: October 04, 2015, 11:31:56 pm by Alex Nikitin »
    • A.N.T. Audio
« Reply #3 on: October 04, 2015, 10:31:30 pm »
Here is the JVR reference output voltage recorded by Keithley 2015 over ~6.5 hours from 10PM till about 4:30AM, the temperature dropped from 23C to 20C over that time with the heating switched off at midnight. Vertical scale is 2ppm/div.
Cheers
Alex
« Last Edit: October 04, 2015, 10:42:36 pm by Alex Nikitin »
Offline poorchava
  • Troll Cave Electronics!
« Reply #4 on: October 04, 2015, 10:52:55 pm »
This thing begs for an oven :)
Wys?ane z mojego HTC One M8s przy u?yciu Tapatalka
    • A.N.T. Audio
« Reply #5 on: October 04, 2015, 10:56:01 pm »
The tempco is easily adjustable to near-zero at a particular temperature. With a dual JFET like U440 I've used, the second FET's junction can be used as a temperature sensor  ;) .
Cheers
Alex
    • A.N.T. Audio
« Reply #6 on: October 04, 2015, 11:15:51 pm »
Here is a 30 min time span "zoomed-in" at the flat part of the graph, with 1ppm/div vertical scale. This is on the bench in open-air, so the "slow" variations easily can be due to thermal effects. Even a simple enclosure may make it better. On the other hand my Keithley 2015 noise at 10V 10NPLC is about 0.6ppm p-p from 10V, or ~1.5ppm p-p at 3.75V, so a large part of variations could be the 2015  ::) .
Cheers
Alex
P.S. I need HP3458A  :-\
« Last Edit: October 04, 2015, 11:25:27 pm by Alex Nikitin »
Offline Andreas
« Reply #7 on: October 05, 2015, 03:02:42 am »
Hello,
interesting finding.
I already had the idea to use a BF245C as "zero quiescent current 5V" voltage regulator (e.g. for a real time clock IC) but never as voltage reference.
The other "problem" is that you will need a output buffer.
Every draw of current will change tempco + output voltage.
with best regards
Andreas
    • A.N.T. Audio
« Reply #8 on: October 05, 2015, 05:00:14 am »
Hi Andreas,
A buffer is required but it is not a difficult task to make one with a chopper opamp and, say, a foil potentiometer. These are just usual design details. What I like about that reference is that the tempco is adjusted "internally" to zero, and there is only one JFET makes the reference, the influence of other parts on the Vref output is easily manageable. The reference can be bootstrapped to the buffer output like in the circuit below (it is a simplified circuit, as one may need a start-up circuit to get this arrangement to work reliably on a single supply).
Cheers
Alex
« Last Edit: October 05, 2015, 05:02:01 am by Alex Nikitin »
Offline ralphd
    • Nerd Ralph
« Reply #9 on: October 05, 2015, 09:49:03 am »
That sounds interesting too.  Do you have a schematic?
Unthinking respect for authority is the greatest enemy of truth. Einstein
Online Marco
« Reply #10 on: October 05, 2015, 11:04:01 am »
I *do* know that buried Zener references have far less time-drift than band-gap references
XFET voltage references on the other hand have had some outlandish long term stability claims in the past, they are no longer present in the current datasheets ... but I'm not so sure they were too far off the mark.
Look at this complaint from Linear about the claimed performance. Yes, without burn in they measured quite a bit of drift. After 400 hours though it became pretty flat, except for the large amount of 1/f noise which they intentionally didn't remove. Pity there aren't any hermetic XFET references.
Offline Andreas
« Reply #11 on: October 05, 2015, 04:37:37 pm »
Hello, (sorry for off topic)
nothing special, I am using the pinch off voltage of the gate as voltage reference.
Drain is input source is output.
Gate is tied to GND.
As the cirquit is very high impedant you will need 2*100nF for stability.
A diode against reverse polarity.
And a protection resistor against transients; thats all.
Schematic + simulation is for 2N4119.
With a BF245C I typically get around 4.0V idle and 2.8-3V with a 1K resistor as load.
But usually a clock only draws some microamps.
Input voltage is up to 25V. (30V).
With best regards
Andreas
Online Marco
« Reply #12 on: October 06, 2015, 04:39:23 am »
Would some foam and mylar be able to isolate the FETs well enough to use the second FET as the heating element? (If so you could just bias it with a source resistor and put a current limited voltage on the drain, it will have a single temperature equilibrium).
    • A.N.T. Audio
« Reply #13 on: October 06, 2015, 04:47:15 am »
Yes, I've looked at that option, however it is only possible for devices with a high enough drain current - not all of them are good in that respect. At the moment I would like to go further with the main reference circuit, the thermal stabilisation would come next. Today I've found an easy (and reasonably cheap) way to get standard voltages (5V and/or 10V) from this circuit. 10V is preferable because the output can be used to supply the reference  ;) . I'll test this idea soon and will update the thread with the results.
Cheers
Alex
Online Marco
« Reply #14 on: October 06, 2015, 05:05:12 am »
No need to use Vg=0 here, a bit of drift from a resistor divider for the temperature setting isn't that critical, so IDSS isn't a limiting factor.
Offline ralphd
    • Nerd Ralph
« Reply #15 on: October 06, 2015, 05:54:35 am »
Thanks.  So the regulated voltage really depends on the current and varies from part to part?
After looking at the prices for jfets (~20c ea), it seems like more of a curiosity than a practical application since LDOs like the AMS1117 and Torex 662K cost 3-5c ea.
    • A.N.T. Audio
« Reply #16 on: October 06, 2015, 08:07:25 am »
So, to scale the voltage up is easy, just needs one quality resistor in series with the reference JFET, like this:
The FET's current is as stable, as the reference Vgs voltage and it just scales up. R2 should have a low tempco though.
Here is the first 30min run's results (vertical scale is 0.5ppm/div  8) ):
One important change from the first prototype. This circuit is still hard-wired and in open air on my bench, however I've used <10ppm/C precision resistors for R1 and R2. From the start I did not expect to get into single ppm/C digits so I've used cheap resistors. However I've found that if I tune the circuit accurately, the largest drift component comes from the resistors! Now I'll leave the circuit on overnight and record the voltage for 7-8 hours.
Cheers
Alex
P.S. - it also lookls like a lot of LF noise was coming from the resistor. Now the noise on the graph is dominated by the noise of the Keithley 2015 I use to measure the voltage.
« Last Edit: October 06, 2015, 08:15:33 am by Alex Nikitin »
Online Marco
« Reply #17 on: October 06, 2015, 08:56:57 am »
Alex, but now you are back to having the resistor drift having an unattenuated effect on your output voltage again. Goodbye potential long term stability.
    • A.N.T. Audio
« Reply #18 on: October 06, 2015, 09:27:09 am »
I am not planning to design a replacement for the LTZ1000  ::) . And if you need a certain standard voltage you have to use some resistors. At least here it is only one really stable resistor required (and an opamp buffer, but that is not a problem).
Cheers
Alex
    • A.N.T. Audio
« Reply #19 on: October 06, 2015, 06:27:23 pm »
Here is the overnight data, the vertical scale is 2ppm/div. For about the same three degrees C temperature drop the voltage change was ~12ppm, or ~4ppm/C, at 5V with the additional scaling resistor. I may yet try to tune it a bit better.
Cheers
Alex
« Last Edit: October 06, 2015, 06:56:22 pm by Alex Nikitin »
    • A.N.T. Audio
« Reply #20 on: October 07, 2015, 02:09:29 am »
As U440 is not easily available, I've also tried a different (and very cheap) JFET - J112. The sample I've used has a near-zero tempco point at ~3.25mA drain current, ~1.75V Vgs. I've scaled it up to 5V with standard 100ppm/C metal film resistors and it again looks like these resistors are the main source of temperature-related voltage variations. The resulting drift, I would estimate at about 20-30ppm/C. Here is the result for a 30 min run in open air (2ppm/div vertical scale).
Cheers
Alex
Online Marco
« Reply #21 on: October 07, 2015, 02:15:29 am »
Supposedly zero tempco for a JFET is near the pinch off voltage plus 0.63V.
Offline Kalvin
  • Embedded SW/HW.
« Reply #22 on: October 07, 2015, 02:28:20 am »
Would placing a few J112-based voltage references in parallel give better results, what do you think?
    • A.N.T. Audio
« Reply #23 on: October 07, 2015, 02:33:05 am »
I think it would be more sensible just to use good low tempco resistors, and, if you need the best stability, a temperature control. By the way, a single device with the same chip inside as the U440 is J211, still made by Fairchild Semi (and available from DigiKey at 15p each).
Cheers
Alex
« Last Edit: October 07, 2015, 02:53:10 am by Alex Nikitin »
Offline Bud
« Reply #24 on: October 07, 2015, 03:44:25 am »
By a coincidence i just went through "The art of electronics" 3rd edition reading on JFETs. The autors do not recommend using JFETs at high source-drain voltage (which your circuit is) because above a few volts gate current raises dramatically, by several orders of magnitude. I'd imagine this may have effect on this circuit out V drift, and make it dependent on the primary regulator output voltage which is the JFET input. But im not good enough with JFETs yet to say for sure. Anyway, something you may need to have a look at and perhaps to test.
    • A.N.T. Audio
« Reply #25 on: October 07, 2015, 04:01:51 am »
By a coincidence, I knew about that nuance of JFETs probably before the first edition of "The Art of Electronics" was published (was it what, 1980? - yes, I am that ancient  :( ). Also (not by a coincidence this time, but as a normal design practice), I tend to look at datasheets of devices I plan to use before actually using them. And in the datasheet for the U440 you may find the following diagram:
And I also have in my library an old catalogue from Siliconix where one can find the same graph for the J112 device (and the gate current for it  starts to increase substantially only for Vds over 15V .
Cheers
Alex
P.S. for 1ppm effect you would need the gate current at about 1-3nA ( for 1-3mA drain current), by the way. Look at the graphs. Even at 125C and 10V the gate current does not reach that level.
« Last Edit: October 07, 2015, 04:11:47 am by Alex Nikitin »
Offline Kalvin
  • Embedded SW/HW.
« Reply #26 on: October 07, 2015, 04:13:22 am »
How well this simple and very affordable JFet-based voltage reference compares to the LM399? Let's assume that one creates a simple "oven" by gluing an appropriate resistor to the J112 as an heater element and places tho whole thing into a styrofoam or similar insulator? Let's assume the constant current through the heater resistor will be good enough to keep the temperature within 1 degree of Celcius. Any educated guesses?
    • A.N.T. Audio
« Reply #27 on: October 07, 2015, 04:27:42 am »
1) The main question would be the long time stability. It is an unknown. My guess would be that unless the junction is damaged by a static discharge, the stability of the JFET itself (especially in a metal can) should be very good to excellent. After all we are looking at a simple single structure, not a complex one as in the LM399 or any other IC reference.
2) It is reasonably easy to create a temperature control down to 0.1 degree, especially for a small device like that, using something like LM35 for a temperature sensing.
I am not sure that it is sensible to look at this circuit as a replacement for the LM399 before some long-term tests are done on it. My aim was to show a simple way of making a voltage reference in a rather unorthodox way, and frankly I am surprised at the level of performance actually available from this circuit.
Cheers
Alex
« Last Edit: October 07, 2015, 04:37:31 am by Alex Nikitin »
Offline 6581
« Reply #28 on: October 07, 2015, 04:39:13 am »
C
Offline Gyro
« Reply #29 on: October 07, 2015, 05:04:04 am »
I'm beginning to wonder if there is too much emphasis to oven implementations rather than temperature compensation / characterized predictable temperature coefficient in all of these references.
Surely long term drift in the oven temperature controller is as much of an error contribution as drift in the reference itself - particularly in a reference as simple as this one (as opposed to LM399 syndrome as Alex hinted). It just seems a bit 'brute force'.
Wouldn't putting the whole thing in an isothermal metal box, insulated to limit rate of temperature change and avoid thermal stresses work better in a home lab environment? Putting the effort into characterization and compensation instead.
Chris
"Victor Meldrew, the Crimson Avenger!"
Offline Bud
« Reply #30 on: October 07, 2015, 05:19:46 am »
Perhaps the dealbreaker is in repeatability. For mass production you throw an integrated IC with tight tolerances and no worries every single unit of the product will be as good as the prototype. Whereas JFETs characteristics differ largely between manufacturers and even within the same batch. The datasheet for J112 says cutoff voltage can be anywhere from 1 to 5v. Having said that this circuit operates at near cut-off voltage on the gate, chances are each unit will need to be individually adjusted for the specified output voltage.
But if you only need one off for personal use then not a problem.
    • A.N.T. Audio
« Reply #31 on: October 07, 2015, 05:27:29 am »
Exactly. If, however, the long term stability happens to be in the range of few ppm per year (and I have a sneaky suspicion that it might, given a metal case and a good attention to the construction and the resistors used), than it is a different ball game altogether  8) . For that one may even select the most suitable FETs out of a large batch (I've just bought 100 of J211 from DigiKey - it cost me less than one LTZ1000 would, even including UPS delivery from the US). But we would know that only in a year or so. Or at least in a 1kh after I'll get my hands on the HP3458A  ::) .
Cheers
Alex
« Last Edit: October 07, 2015, 05:30:47 am by Alex Nikitin »
Offline Gyro
« Reply #32 on: October 07, 2015, 05:59:15 am »
However neither of the 'superstars' ie. LTZ1000 and LM399 is tight initial tolerance. The have quite a large variation requiring adjustments on an individual basis. Stabilty and long term drift are excellent but initial tolerance is poor.
Chris
Offline branadic
« Reply #33 on: October 07, 2015, 06:01:25 am »
I might be wrong, but I found J112 to be TO92 and SOT23-3 only, no metal can. So humidity sensitivity can cause a big influence, right? A big advantage of your U440.
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
    • A.N.T. Audio
« Reply #34 on: October 07, 2015, 06:08:40 am »
Yes, that is correct. However the humidity sensitivity is usually due to the resistors on the chip, here we don't have any. It is possible that for a JFET even in a plastic case it should not be a problem.
 ::) Now I need an environmental chamber as well!
Cheers
Alex
« Last Edit: October 07, 2015, 06:11:31 am by Alex Nikitin »
Offline Kalvin
  • Embedded SW/HW.
« Reply #35 on: October 07, 2015, 06:23:11 am »
I guess it is still not too late the sell those 100 J211s and forget the whole project, and buy the LTZ1000 :)
    • A.N.T. Audio
« Reply #36 on: October 07, 2015, 06:25:53 am »
Nah, that would be too easy. I am looking for fun, not for gain  8) .
Cheers
Alex
P.S. - on the other hand, the LTZ needs a very good board layout and a lot of expensive supporting electronics. Sort of fun too, but it is already a known road  ;) .
« Last Edit: October 07, 2015, 06:35:18 am by Alex Nikitin »
Offline branadic
« Reply #37 on: October 07, 2015, 07:02:12 am »
however the humidity sensitivity is usually due to the resistors on the chip, here we don't have any. It is possible that for a JFET even in a plastic case it should not be a problem.
Are you sure about that? What about the Source-Drain-channel, not some kind of a resistor though? Not sensitive to mechanical stress?
    • A.N.T. Audio
« Reply #38 on: October 07, 2015, 07:10:21 am »
The only way to be reasonably sure is to test thoroughly. I have no data on a single JFET package sensitivity in that respect. On the other hand, everything is sensitive to everything, the only question is to what degree  ;) .
Cheers
Alex
Online Marco
« Reply #39 on: October 07, 2015, 07:15:21 am »
Does a metal can maintain constant air pressure?
    • A.N.T. Audio
« Reply #40 on: October 07, 2015, 07:19:36 am »
AFAIK, strictly speaking, no, as it is though a metal package is hermetic, but not evacuated, and the internal gas pressure should change with temperature.
Cheers
Alex
Offline codeboy2k
« Reply #41 on: October 07, 2015, 02:23:08 pm »
There is a 1973 US patent (long expired now)  3,760,199  that details a manner for determining a compensating voltage for biasing any JFET into it's 0-tempco region.  The goal of the design in the patent was to determine Vp, the pinchoff voltage, at any temperature, and a compensating voltage Vc, such that Vgsz, the gate-source voltage needed to bias the JFET at its 0-tempco region was Vgsz = Vp + Vc.  Since Vp, the pinchoff voltage is temperature dependent, then he develops Vc = 0.64 + a temperature dependent part opposing the tempco of Vp. With proper placement of some PN junctions and some BJTs and resistors, and a little bit of KVL, the inventor develops a voltage loop involving Vp, Vc and 0.64 and places the second JFET's gate-source across the point where those voltages are summed, yielding Vgs = Vp + Vc + 0.64, without ever knowing what Vp was (so it can vary with the standard process variations).  The second JFET will now have this temperature compensated bias voltage applied to it's gate wrt source and develop a nearly constant drain current that is, for the most part, temperature independent.  Very Cool!
It's an easy patent to read and a good one too. I think it was intended for use at chip scale, so that all the elements are matched and isothermic.  It might not work with discrete parts if someone were to try it out.
Offline ralphd
    • Nerd Ralph
« Reply #42 on: October 07, 2015, 03:26:16 pm »
Immerse it in mineral oil in a tictac container.  No humidity effect and more temperature stability.
Offline jwm_
    • Not A Number
« Reply #43 on: October 07, 2015, 03:44:36 pm »
Now that I have my scavenged cryocooler getting to temperatures it can liquify air ( ) and keep high temp superconductors superconducting indefinitely, I thought a neat trick would be to be the ultimate volt-nut and combine it with my rubidium standard to make a Josephson junction voltage reference. First step is to get a SQUID working, it would at least show the effect is reproducible with my setup. I think it is unlikely I will succeed on my allotted budget, but I am sure it will be a fun ride anyway.
    • A.N.T. Audio
« Reply #44 on: October 09, 2015, 06:09:46 am »
Well, my 100 of J211 arrived from DigiKey and I've measured the cut-off voltage on the lot at 5V Vgd. Actually, a very consistent batch for JFETs, it looks like the processing did improve since 1980s. All hundred measured between 2.75V and 3.35V at ~3uA current (10M of Keithley 2015 on 100V range between the source and gate). My "experimental" U440 measures 4.6V in the same setup (and the zero tempco point is at ~3.74V, so the "theory" about 0.63 or 0.64V is not working  ;) ). One obvious bad point about plastic casing - there is some photo-sensitivity, so the device should not be exposed to light when used as a reference. Not much of a problem though. I am nearly finalised a complete reference circuit, which would take 12-24V supply and provide 10V output with up to 10mA current available and would contain only two FETs, an opamp and few resistors. Not temperature regulated - that would be the next step.
Cheers
Alex
« Last Edit: October 09, 2015, 06:18:08 am by Alex Nikitin »
Offline BravoV
« Reply #45 on: October 09, 2015, 06:18:31 am »
..... I am nearly finalised a complete reference circuit...
 :popcorn:
    • A.N.T. Audio
« Reply #46 on: October 09, 2015, 06:43:14 am »
:popcorn:
I am not going to publish a complete circuit any time soon, only measurement results  8) . I plan to include a temperature sweep from 0 to 40C and a 1000h stability test against LM399 (I don't have LTZ1000  :( ).
Cheers
Alex
Offline Bud
« Reply #47 on: October 09, 2015, 07:13:14 am »
 :popcorn: ... :popcorn:... :popcorn:
    • A.N.T. Audio
« Reply #48 on: October 12, 2015, 01:37:55 am »
OK, let's not get all this popcorn wasted  8) .
I've decided to try a very simple circuit to get 10V output and to use the J211 transistor as a reference. This is a "budget" solution, as it requires only two good quality resistors (R1 and R2 and I've used 15ppm metal film for that, "padded" to the required value by some standard 100ppm/C resistors, so I should not expect much better performance than this value), cheap J211 JFET and an inexpensive INA133 chip connected as a gain 2 buffer with low tempco (~2ppm/C). Altogether is less than £10 in parts (resistors R3-R6 are internal for the INA133).
Here is the circuit which I've assembled on an old piece of a pcb:
And here is a 30-min run on the bench (vertical scale is 1ppm/div):
My estimate is that the temco around 23C is indeed about 10-15ppm/C. The power supply voltage variations from 12V to 28V create less than 2ppm/V change.  The load regulation is better than 0.5ppm/mA (a conservative estimate, as a load current up to 5mA, sink or source, creates less than 1ppm voltage change). The supply current is about 2mA.
First R1 is selected for the lowest tempco and then R2 for 10V output voltage.
Cheers
Alex
« Last Edit: October 12, 2015, 02:42:55 am by Alex Nikitin »
    • A.N.T. Audio
« Reply #49 on: October 12, 2015, 05:53:15 am »
I've seen this schematic some days ago in the LF355 datasheet, it made me think about you! Your idea is not new.
Thank you, I always thought that it was used before but couldn't find a confirmation!
Cheers
Alex
P.S. - would be interesting to find if that circuit was used somewhere in practice!
P.P.S. - The 2N4118 has a very low drain current (around 100uA for near-zero tempco) - it should make it good for low power but fairly noisy as a reference. A higher current device would make it a very low noise option - much quieter than ~1ppm p-p LF noise of the LM399.
« Last Edit: October 12, 2015, 06:23:46 am by Alex Nikitin »
    • A.N.T. Audio
« Reply #50 on: October 16, 2015, 08:22:04 am »
Well, my two prototypes (5V with U440 and 10V with J211) are running nicely and manage to stay inside +/-10ppm range at a somewhat varying room temperature, without any obvious drift. I am waiting for some high stability wire-wound and bulk foil resistors to be delivered so I can start building a "superstable" version, where all influences except that only of a JFET itself would be below 1ppm/C. The aim is to get a stability better than +/-10ppm in 20-30C temperature range (or better than 2ppm/C) without temperature stabilisation and then check the long-term drift for 1000h with a temperature log.
Cheers
Alex
P.S. - here is a 30min run on the bench of a somewhat "fine tuned" version on J211 device. Vertical scale 1ppm/div, Keithley 2015 works with a 10x averaging filter at NPLC10 to reduce the noise.
« Last Edit: October 16, 2015, 09:30:37 am by Alex Nikitin »
Online Marco
« Reply #51 on: October 16, 2015, 04:38:02 pm »
Might it be possible to get a low chosen voltage (say 2.5V) at near zero tempco by using an opamp based "constant" current sink in parallel with the source resistor?
« Last Edit: October 16, 2015, 05:33:08 pm by Marco »
    • A.N.T. Audio
« Reply #52 on: October 16, 2015, 05:47:49 pm »
Overnight (from 12am till 6am, in open air), 1ppm/div:
Cheers
Alex
« Last Edit: October 16, 2015, 06:49:10 pm by Alex Nikitin »
    • A.N.T. Audio
« Reply #53 on: October 16, 2015, 06:16:31 pm »
It is not difficult to get lower voltages - just by choosing a different JFET. The sample of J211 I'm using right now has the reference voltage around 2.6V and J112 I've tried earlier - even lower, about 1.7V. I suspect however that it could be better to have this voltage as high as possible for a low noise and better stability.
Cheers
Alex
« Last Edit: October 16, 2015, 06:50:50 pm by Alex Nikitin »
Online Marco
« Reply #54 on: October 16, 2015, 07:11:07 pm »
The point wasn't to get a lower voltage, the point was to get a given voltage. It would be nice to have the elegance of the first circuit (where variations from everything but the JFET on output voltage are highly attenuated) but at a specific voltage. If you have 2.5V you can make 5v out of it with a switched capacitor circuit.
    • A.N.T. Audio
« Reply #55 on: October 16, 2015, 09:13:53 pm »
Well, in that case an additional current source should also have a very high stability (means at least some top quality resistors) and would complicate things. An additional resistor gives you the required voltage above the reference, nothing more is required.
Cheers
Alex
Online Marco
« Reply #56 on: October 16, 2015, 10:07:33 pm »
I'd have to the math to be sure, but as far as I can see drift in the current source has an attenuated effect due to transconductance just like variation in the source resistor (but NOT the second resistor you are adding in your current circuits).
PS. doing even a small bit of math in my head showed me I was wrong, I was overestimating transconductance by many orders of magnitude.
« Last Edit: October 16, 2015, 10:31:55 pm by Marco »
    • A.N.T. Audio
« Reply #57 on: October 16, 2015, 10:13:39 pm »
Yes, however there is only one reference voltage point and you can not move it by adding a current, you only can adjust the voltage to that point by an external current if the source resistor has not quite correct value. It is easier IMHO, just to get the resistor right in the first place (or to use a trimmer to get the last 100ppm).
Cheers
Alex
    • A.N.T. Audio
« Reply #58 on: October 17, 2015, 06:22:06 am »
OK, Friday night - a torture night  >:D . As the J211 version runs rather well and I am still waiting for the bits to build a controlled temperature chamber, I've decided to torture the JVR on record, first with a freezer spray for 10sec (about -50C or 70 degrees drop), then with a soldering iron for 20 seconds flat on the plastic case of JFET (I would estimate it as at least 70 degrees rise). Unlike the U440 in a metal case, some hysteresis is clearly visible, about +20ppm after the cold session and about -30ppm after the hot one. Here I have to note that the freezer session was not the first and on the first session I didn't freeze the FET so deeply, however the hysteresis was more - about 50ppm. Clearly for a stable reference a metal case is much better. (A note - the "wave" on the second graph after 400 samples is just me adjusting the voltage back to 10.00000 Volts).
Otherwise it is clear from the graphs that in both cases the voltage drops about 0.2-0.25% or 2000-2500ppm for 70-75 degrees change, so at 23C the reference sits nicely on the very top of the curve and has the lowest possible tempco. With a metal case and the temperature stabilised around, say, 45-50C (and the tempco adjusted for zero at that temperature), the stability should be very good.
Cheers
Alex
Each record is about 6min long:
« Last Edit: October 17, 2015, 06:51:57 am by Alex Nikitin »
    • A.N.T. Audio
« Reply #59 on: October 18, 2015, 01:03:10 am »
A quick estimate - the temperature coefficient for this version (10V with J211 and INA133) is about 5ppm/C with few degrees change near 23C and up to 30-40ppm/C at temperature extremes.
Cheers
Alex
    • A.N.T. Audio
« Reply #60 on: October 18, 2015, 04:21:30 am »
To compare my results with a quality band-gap reference, I've decided to repeat my torture tests on LT1019-2.5 in DIL8 I had in the lab (with a stated 5ppm/C typical and 20ppm/C maximum tempco). This chip did show however a lot of hysteresis - about +600ppm after a freeze and -400ppm after a soldering iron application, leaving a total permanent shift about +200ppm. Ouch!  :o
Cheers
Alex
« Last Edit: October 18, 2015, 05:09:50 am by Alex Nikitin »
Offline Messtechniker
« Reply #61 on: October 18, 2015, 10:16:12 pm »
Since I am new to this game please bear with me.
Simply as a first test, I built your circuit without the amplifier and with the components at hand. FET was BF 246 C and 2N 3819 new old stock resting for some 40 years in the box. Resistors were all standard metal film ¼ W, 1%, 50 ppm. Stability of these FETs was pretty awful +/- 300 to 500 ppm over 1 hour. So I had a look at the power supply. This was a LM 317 set to 24 V (to power the amplifier at some later date). Load was 10 mA approx. (dropping resistor and LED) and fed with 30 V from a benchtop power supply. Surprisingly drift was around +/- 5 ppm at room temperature.
Then to investigate further I changed the output to around 10 V. Feed voltage now being 20 V. What I then got is for you to see.
At the beginning it’s the test rig without me in the room. The droop before the peak is me entering the room whereby the “air blast” seemed to cool the LM 317 a little. The big peak is due to the soldering iron placed at the LM 317 for a short while. Then left the LM 317 to settle with some hysteresis being apparent. 25 ppm approx.
So I’ll probably follow this route and put the thing in a metal box and heat it to some temperature. Since the TO-220 casing is flat, I plan to attach flush one BD 139 (TO 126) as the temperature sensor and two further BD 139s on the outsides as heaters bolted together as a single compact pack. Hope this makes for a fair voltage reference, and this with standard components only.
Yours Messtechniker.
P.S: This is great fun. :)
Agilent 34465A, Hameg HMO1022, R&S HMC 8043, Voltcraft VC 940 and M-Audio Audiophile 192
    • A.N.T. Audio
« Reply #62 on: October 21, 2015, 02:13:08 am »
Oh, yes!
I've just bought the Fluke 731B and a lot of precision wire-wound resistors. I would like at least match the stability on the Fluke (and, as a side point, to be able to calibrate in-house the DC parameters of my two Keithley 2015).
Cheers
Alex
    • A.N.T. Audio
« Reply #63 on: October 22, 2015, 05:20:29 am »
While I am waiting for the Fluke and resistors, here is a 21 hours run of the same 10V version using J211. The Keithley 2015 runs with 40 averaging, vertical scale is 1ppm/div. I've noted the room temperature at some time points. It looks like the TC around 23C is ~2ppm/C  8) .
Cheers
Alex
Offline Andreas
« Reply #64 on: October 22, 2015, 03:42:05 pm »
The TC of the Keithley (according to datasheet) or of the JVR?
With best regards
Andreas
    • A.N.T. Audio
« Reply #65 on: October 22, 2015, 06:31:13 pm »
With best regards
Andreas
Hi Andreas,
Your guess is as good as mine - the Keithley TC, noise and drift all included! The only possible clue is that the "fast" changes are more likely to be that of the JVR itself due to the air movements. As you may see there are variations in the evening when I was moving around the lab and in the morning hour when I was preparing to go to work, while at night and during the day there was very little air movement and the changes are more gradual and uniform. I hope that the Fluke 731B will provide me with a more stable reference than the 2015s (I'm running two in parallel though recording only one - and the difference between two 2015 varies from 4 to 6ppm). I plan to record the Fluke against 2015s after a good warm-up and then record the difference between the Fluke and the JVR. I also plan to build the "ultimate" 10V JVR using U440 and a Vishay Foil resistor divider for the 5 to 10V step up.
Cheers
Alex
« Last Edit: October 22, 2015, 06:41:16 pm by Alex Nikitin »
Offline Andreas
« Reply #66 on: October 23, 2015, 08:26:23 am »
Hello Alex,
air movements?
or changes by parasytic capacitance?
have you checked what happens when your hand moves over the cirquit or touch the power supply and voltage output lines?
My "zero quiescent current voltage regulators" were very sensitive in this regard.
 (several volts output change without blocking capacitors).
I had to use the blocking capacitors to keep the noise away from the gate diode.
With best regards
Andreas
    • A.N.T. Audio
« Reply #67 on: October 23, 2015, 08:40:53 am »
Hello Alex,
air movements?
With best regards
Andreas
Hi Andreas,
No, these "fast" changes are actually quite slow - minutes. And the reference is not sensitive to the external interference or capacitive coupling. I can touch it, move it, shake it and, as long as I don't transfer any heat from my hand to the circuit, it stays rock solid. I am now building a 0.1Hz-10Hz filter-amplifier, with the gain 10000 and self-noise around 100-150nV p-p in the pass-band, so I can accurately observe and measure the LF noise.
Cheers
Alex
    • A.N.T. Audio
« Reply #68 on: October 28, 2015, 07:09:45 am »
Well, the Fluke 731B arrived yesterday (no battery though, so I've disconnected R30 to avoid the indicator overload and added noise). It was last calibrated by Fluke in 2009 and agrees with both of my Keithley 2015s within 10ppm on the 10V output. I've run it overnight and it did show about -1ppm/C TC  - see the graph below, taken with NPLC10 and averaging over 100 samples. On the previous night I did run my 10V JVR with J211 using the same settings on the 2015. Below both graphs using the same scale. Again, I've noted the room temperature at some points. Next step would be to put the JVR in a suitable enclosure as the "faster" variations are most likely due to the air movements in the room.
Cheers
Alex
« Last Edit: October 28, 2015, 07:15:54 am by Alex Nikitin »
« Reply #69 on: October 28, 2015, 07:08:56 pm »
Well, here now the next development step of my poor man’s 10 V reference based on the venerable LM 317 with its internal bandgap reference. The LM 317 is now heated by two BD 139 trannies on each side with a third BD 139 serving as a temperature sensor sandwiched in between making a single compact block bolted together with a M3 bolt. The arrangement is now insensitive to air movements although not yet contained in a metal box.
The next development steps will be:
1.)   Replace the metal M3 bolt and nut of the block with a plastic M3 bolt and nut.
2.)   Put the LM 317 block with the heating circuit on an new and clean hole dot matrix epoxy pcb. (instead of the currently used hard paper hole dot matrix pcb. material and avoiding flux residues) in a small metal box.
3.)   Provide separate power supplies for the LM 317 block (+15 V) and the heater (+15 V, -15 V and - 5 V)
4.)   Do some extended burning-in and subsequent ppm measurements. Watch this space.
By the way, my Keysight 34465A DMM set to the 10 V range and with its input shorted produces an output fluctuating at about 0.2 ppm (as measured over 1 hour). So my measurement limit, i.e the contribution from the 34465A in the enclosed graph will be about 10 times that at 2 ppm re. 10 V.
My aim is to produce a voltage reference circuit specc’d for around +/- 10 ppm (20 ppm total) perhaps, using cheap off-the-shelf components only.
What do you think?
Yours – Messtechniker
P.S. Still having fun with this. :)
Offline BravoV
« Reply #70 on: October 28, 2015, 11:01:11 pm »
Messtechniker, schematic and photos please, tia.
    • A.N.T. Audio
« Reply #71 on: October 28, 2015, 11:05:31 pm »
What do you think?
Where are the voltage setting resistors placed? You need to keep these at a stable temperature too if you are looking at <20ppm stability.
Cheers
Alex
« Reply #72 on: October 29, 2015, 12:25:02 am »
Circuit is still at a very early stage (see remark below). Will provide it as soon as I (hopefully) get improved results.
The voltage setting resistors are currently placed close to the heated LM 317 but horizontally on the pcb. Based on your remark I'll put them upright and against each side of the LM 317 in thermal contact. Will look messy, but will probably be more effective. Thanks for the hint.
Yours - Messtechniker
Offline Andreas
« Reply #73 on: October 29, 2015, 08:12:19 am »
What do you think?
I think you could get better results when using a LM723.
the LM317 has alone a "typical" noise of 30 ppm RMS = 2mVpp.
with some clever design you could use the current limiting transistor
as chip temperature sensing diode for the oven.
with best regards
Andreas
    • A.N.T. Audio
« Reply #74 on: October 29, 2015, 10:25:31 am »
Two more graphs, taken at NPLC1, no filtering, for 30 minutes. The room temperature was about 24.5C, the JVR is still in open air on the bench. I was hoping to compare the noise, however the noise in the graphs is completely dominated by the noise of Keithley 2015  :( .
Cheers
Alex
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #75 on: October 29, 2015, 12:16:36 pm »
Why NPLC1? Run NPLC5 ;).
    • A.N.T. Audio
« Reply #76 on: October 29, 2015, 08:53:06 pm »
OK, I'll try. I am building an amplifier with 0.1-10Hz filter to measure the p-p noise on a scope, that should give a much more accurate result.
Cheers
Alex
Offline Andreas
« Reply #77 on: October 30, 2015, 07:52:37 am »
Which of the cirquits / application notes do you use for this?
With best regards
Andreas
    • A.N.T. Audio
« Reply #78 on: October 30, 2015, 10:00:50 am »
With best regards
Andreas
I took the response graph from the OP27 datasheet and simulated a suitable circuit using the OPA2227 opamp with a couple of 9v batteries for the P/S.
Cheers
Alex
Offline Andreas
« Reply #79 on: October 30, 2015, 07:08:15 pm »
Cheers
Alex
Hello,
we are talking of figure 31 in the AD datasheet?
http://www.analog.com/media/en/technical-documentation/data-sheets/OP27.pdf
With this cirquit the input impedance is 10 Ohms.
The DUT has to be below 0.1 Ohms without having too much errors.
(And the output voltage has to be very close to 0 Volts without saturation).
A standard Zener has around 10 Ohms. Your cirquit will have several 10s of Ohms.
Have a look on AN124
http://cds.linear.com/docs/en/application-note/an124f.pdf
Similar results can be get by a 2 stage design with LT1037 and LT1012 + selected (low leakage current)
standard electrolytic capacitors.
And dont forget the (metal) cookies box (when using LT Parts) or the (metal) paint can (when using TI parts).
With best regards
Andreas
    • A.N.T. Audio
« Reply #80 on: October 30, 2015, 07:42:04 pm »
Cheers
Alex
Hello,
With best regards
Andreas
Hi Andreas,
I only took the graph as a reference, the circuit there is not suitable. I also have an option to adapt one of my FET phono stages circuits for this filter-amplifier, the simulation shows that I can get a gain of 1000 with the right frequency response, however I am not sure how quiet the input FET will be at 0.1-10Hz. I'll probably try it anyway, as I have plenty of nice pcbs with an excellent low signal level behaviour and even some good 24V DC power supplies (including some rechargeable batteries) left from the time I was building and selling these plus these fit perfectly into an aluminium Hammond enclosure. If I can get <100nV p-p self-noise, I'll be quite happy.
Cheers
Alex
    • A.N.T. Audio
« Reply #81 on: November 01, 2015, 08:10:45 am »
I've received some Vishay Bulk Foil 5K+5K resistor dividers - I plan to use these for 5V to 10V amplifier in the "ultimate" version of the JVR. I've decided to check how accurate they are. With 0.05% stated tolerance I was not expecting a much better performance, however the measurements show that the resistor pairs are matched very closely - the worst one out of ten is about 15ppm off, and the best two were good to 1ppm or less, as that was the limit how accurate I could measure the error of the divider in my simple improvised DC bridge, using Keithley 2015 as a null-meter (the inconsistency in the results with plugging/unplugging the resistors and repeated measurements were about 1ppm).
Cheers
Alex
« Last Edit: November 01, 2015, 08:35:13 am by Alex Nikitin »
Offline BravoV
« Reply #82 on: November 13, 2015, 06:48:00 pm »
Will 2N4392 TO-18 (datasheet) work decently ? Our local store carries it.
    • A.N.T. Audio
« Reply #83 on: November 13, 2015, 07:48:50 pm »
Yes, it looks like a very suitable device, with a good Vgs(off) between 2 and 5V., and a metal case. Another possible option available is 2N4857A/2N4856A.
Cheers
Alex
P.S. I've ordered some of 2N4392 and 2N4391 from Mouser, it looks like the cheapest TO-18 metal can JFET available new right now. Thank you for the pointer.
« Last Edit: November 14, 2015, 12:06:51 am by Alex Nikitin »
    • A.N.T. Audio
« Reply #84 on: November 17, 2015, 09:22:27 pm »
1) I've received 2N4391 and 2N4392. The 4391 devices look really interesting, as the zero tempco point is very high (from 6V to 8V) and the current at that point is low (1-1.5mA), that makes for a very low internal dissipation (only 2 to 6mW with 10V supply) and very good stability. The gate is connected to the case and is grounded in the circuit, which is also good. I will build one JVR sample with the 2N4391 now, it looks very promising.
2) I've done a small mod on the Fluke 731B. As I am running it without a battery, the needle meter on the front panel is essentially useless. I've installed a temperature sensor (LM35) and wired the meter so it shows the temperature of the internal screen around the reference board. It did not require any significant changes, only a cut in one of the power supply board tracks and a couple of resistors. Now I have an accurate internal thermometer with the scale roughly from 18C to 27C (the green sector is from 20C to 24C) which gives me a very good idea of the output voltage deviation.
Cheers
Alex
Offline ltz2000
« Reply #85 on: November 18, 2015, 12:24:09 am »
Another mofication worth doing, which actually improves the stability, is replacing the crappy pre-regulator. I used the good (enough) old LM723.
    • A.N.T. Audio
« Reply #86 on: November 21, 2015, 08:03:25 am »
I've built the first 10V prototype with the 2N4391. The reference voltage was around 7.228V, so I've used OPA227 and some wire-wound resistors to scale it up to 10V, the same type wire wound resistors were used for the reference source as well. Without much tuning I've got around 2ppm/C and very low noise level. This is the first time I've measured the JVR against the Fluke 731B, using Keithley 2015 on 100mV range. The variations are around 1ppm, the noise is well below 1ppm p-p.
This is a step towards a properly built reference - I've used a good aluminium case, good resistors and Tellurium Copper terminals, though the circuit is built on a prototyping board.
Cheers
Alex
10V JVR on 2N4391 measured for 30 min against the Fluke 731B by Keithley 2015, 100mV range, NPLC1, no filtering, 1ppm per division:
    • A.N.T. Audio
« Reply #87 on: November 22, 2015, 09:45:10 am »
Here is another graph of the 2N4391 JVR v Fluke 731B. This time I've used a better cable (a screened pure copper twisted pair with copper plugs which I've just received), a foam cup on the 2N4391 FET and NPLC10 on the Keithley 2015. 30min run, 1ppm per division.
Cheers
Alex
« Last Edit: November 22, 2015, 09:48:20 am by Alex Nikitin »
Offline Andreas
« Reply #88 on: November 22, 2015, 10:11:12 am »
Hello,
ist the lesser noise mainly from NPLC10 or also from the additional shielding?
with best regards
Andreas
    • A.N.T. Audio
« Reply #89 on: November 22, 2015, 10:19:15 am »
Hello,
with best regards
Andreas
The noise is less because of NPLC10, however the slow voltage variations are less due most likely to smaller thermal effects from pure copper cabling and connectors. I've used just standard cables previously and at least some of the drift was due to thermoelectric effects.
Cheers
Alex
    • A.N.T. Audio
« Reply #90 on: November 22, 2015, 09:53:49 pm »
Here is another graph, this time for an overnight measurement, where the lab temperature varied from 22.5C in the evening down to 19C early in the morning before the heating was switched on. Vertical scale is 1ppm/div, so the tempco is about 1ppm/C, on par with the Fluke 731B. I may try to tweak it a bit more to get even lower figures. The circuit allows for a two pot trimming, one for the tempco and another for the voltage, without affecting the tempco, however as there are drifts in the other parts of the circuit, some iterations are required to get closer to zero in a 5-6 degrees C range. I may try to do it later today.
Cheers
Alex
« Last Edit: November 23, 2015, 07:02:42 am by Alex Nikitin »
    • A.N.T. Audio
« Reply #91 on: November 25, 2015, 07:54:38 am »
It looks like I've got the tempco under 1ppm/C, though it is still positive and combined with a bit over 1ppm/C negative from the Fluke 731B the tempco of the differential voltage between two units is about 2ppm/C, on the first graph. The combined noise is, however, very low, a small fraction of a ppm, as visible on the first 24h graph (2ppm/division). On the second 24h graph the JVR is measured by Keithley 2015 and the noise of the meter dominates, however the tempco is just under 1ppm/C. The 2N4391 costs about $2 from mouser.com and probably the cheapest new metal can JFET you can buy - and quite suitable to use in the JVR. A very simple and cheap reference with potentially a good long-term stability, plus low and adjustable tempco can be built with this JFET, a quality opamp (i.e. OP27) and few good resistors (actually, you only need three resistors with the right values, but as JFETs parameters have a large spread it is necessary to build up the right values from several standard ones).
Cheers
Alex
« Last Edit: November 25, 2015, 10:04:34 am by Alex Nikitin »
Offline Macbeth
« Reply #92 on: November 25, 2015, 09:22:25 am »
731B? meh, we need a 734A at least when measuring a cheap FET standard, providing a JJA is not immediately at hand  :palm: (ignore me, just bookmarking!  ;) )
Offline uski
« Reply #93 on: November 25, 2015, 11:14:47 am »
:popcorn: keep going, I've been looking for a cheap reference good enough to roughly calibrate my 5 1/2 digits multimeter
    • A.N.T. Audio
« Reply #94 on: November 28, 2015, 09:48:02 am »
The weather was warmer today and I've pushed the room temperature up to 24-25C. Interestingly, it looks like my current version of the JVR with 2N4391 has a near-zero tempco between 22C and 24C - the voltage changes are less than 1ppm in that range, and at 25C the voltage starts to drop, so the tempco changes to negative. The Fluke 731B voltage changes about 4ppm in the same 3 degrees range, still negative.
Cheers
Alex
P.S. - this version runs untouched for about 150 hours and there is no voltage change at the same room temperature I can reliably detect (~below 2ppm).
« Last Edit: November 28, 2015, 10:03:07 am by Alex Nikitin »
    • A.N.T. Audio
« Reply #95 on: January 13, 2016, 09:11:59 am »
Here is a graph of the JVR voltage v room temperature over one month (from 12th of December till today). As I've recorded the voltage mostly from the display of Kethley 2015, there is at least 1ppm uncertainty of that readout. When the display did flicker between two points, I've recorded a half-ppm point. It looks like the JVR + Keithley combination drifted not more than 2-3ppm over that month. Either both are quite stable or there is a remarkable similarity in the drift character of two units  :-// .
Cheers
Alex
« Last Edit: January 13, 2016, 09:39:41 am by Alex Nikitin »
    • A.N.T. Audio
« Reply #96 on: February 20, 2016, 03:53:45 am »
I've measured the same 10V JVR unit with HP3458A over three and half hours today. The temperature in the lab was between 21-22C all that time. Vertical scale is 0.5ppm/div .
Cheers
Alex
    • A.N.T. Audio
« Reply #97 on: February 25, 2016, 03:34:11 am »
More measurements with HP3458A - this time for 29 hours. At night time temperature in the lab drops steadily to about 18C in the morning, and during the day the temperature is mostly reasonably stable around 21-23C. The day to day stability looks very good, even with the unit travelling with me and so experiencing a much colder conditions (around 3-7C). 1ppm/div vertical scale.
Cheers
Alex
Offline Kleinstein
« Reply #98 on: February 25, 2016, 04:24:40 am »
For such a low current reference the performance looks really good.
With the HP3458 measurements there might also be some contribution from it's internal drift. With changing temperature you are supposed to use ACAL routine from time to time.
    • A.N.T. Audio
« Reply #99 on: February 25, 2016, 05:09:22 am »
Thank you. I've just left the setup running unattended. The HP3458 has option 002 though, so I hope the internal drift (especially on 10V range) should be minimal (ACAL gave me a difference below 0.5ppm when I've run it at different temperatures, i.e 18C and 23C.
Cheers
Alex
Online Vgkid
« Reply #100 on: February 25, 2016, 05:22:20 am »
Thanks for the updates, especially the drift performance.
« Reply #101 on: February 28, 2016, 02:27:59 am »
Odd bit of synchronicity here. This thread came back to life a few days ago and today I was hunting around for an old microphone preamp design of mine. In the process I tripped across the circuit for the AKG C451E microphone's internal preamp. This is phantom powered and lurking inside it is a JFET acting as a voltage reference for the on-board voltage regulator. It's T5 in the diagram below. Note that R15 is marked as select-on-test.
    • A.N.T. Audio
« Reply #102 on: February 29, 2016, 09:36:30 pm »
That is one of the best 12h runs I've had with this reference. It was taken on Saturday with the room closed and no heating, so the temperature was about 18C all day from 7:30 till 19:30. Vertical scale is 0.5ppm/div.
Cheers
Alex
« Last Edit: February 29, 2016, 09:38:40 pm by Alex Nikitin »
Offline zlymex
  • Country: cn
« Reply #103 on: March 05, 2016, 02:48:05 pm »
Anyone heard of the related story? Here is goes.
1. For a very long time, only LT's LTZ1000 claims to have 2uV drift @1kHr(@65C, roughly 0.3ppm @1kHr)
2. In late 90's, Analog annouced its XFET based ADR293 5V Vref claimed to have drift of only 0.2ppm @1kHr @25C (better than LTZ1000!)
http://pdf1.alldatasheet.com/datasheet-pdf/view/48847/AD/ADR293.html
3. There are people, without extended test, written an article in 1999 introducing this 'remarkable' device.
http://www.ti.com/lit/an/slyt183/slyt183.pdf
4. LT were getting mad at this and written a design note in 2000 to criticize this using phrase like 'THIS IS A DELIBERATE LIE!'
http://cds.linear.com/docs/en/design-note/dn229f.pdf
5. Analog made modifications to the datasheet from 0.2ppm to 50ppm, test conditions also modified
http://www.analog.com/media/en/technical-documentation/data-sheets/ADR293.pdf
6. Although this ADR293 is still in production, its not recommended for new design. Rather, Analog gives an alternative ADR3450, which is CMOS based and with much larger noise.
http://www.analog.com/media/en/technical-documentation/data-sheets/ADR3412_ADR3420_ADR3425_ADR3430_ADR3433_ADR3440_ADR3450.pdf
The following users thanked this post: kony, mmagin
Offline Macbeth
« Reply #104 on: March 05, 2016, 03:59:14 pm »
@zlymex - Thanks for that.
So the XFET based ADR293 has a 1000 hour long term stability of just 0.2ppm @ 25C - but the latest datasheet has it at 50ppm @ 125C !!? ... and they recommend this for battery powered devices? Did they really burn this in at 125C? why? :palm:
ETA: Design Note 229 explains it!  :rant:
« Last Edit: March 05, 2016, 04:07:31 pm by Macbeth »
Offline zlymex
  • Country: cn
« Reply #105 on: March 05, 2016, 07:05:56 pm »
That's right. Analog must have tested some devices at 125 deg C for 1000hrs that they typically show 50 ppm drift. And they derived at first that the drift at 25 deg C to be 0.2ppm @1000hrs.
    • A.N.T. Audio
« Reply #106 on: March 07, 2016, 11:19:52 pm »
Well, back to the topic  ;) . Here are the measurements on HP3458A for my 10V JVR unit for two weeks combined, 1ppm/division and with some room temperature markers. There are some gaps in the timeline.
Cheers
Alex
« Last Edit: March 07, 2016, 11:23:21 pm by Alex Nikitin »
Offline Squantor
« Reply #107 on: March 08, 2016, 06:51:37 pm »
Just to share my experience.
When I was matching BF862's on VGS using the setup of gate connected to source and measuring the voltage drop at a certain current. The Vgs usually climbs as the device is heating up, this was prevalent on the higher side of the IDSS current range (10 to 25mA). But I noticed that devices around 15mA did not have this effect at all, and below this current it actually was the opposite.
When I have some time I will perform some tests with the BF862. Problem is that these are SMT devices and interchanging is not as easy. I do have a socket from my old matching setup that I can reuse, but dont know about the thermal EMF quality of this socket
    • A.N.T. Audio
« Reply #108 on: March 15, 2016, 03:51:47 am »
I've checked the load regulation on the 10V unit today. 10K (1mA) load directly on the output connectors drops the voltage by 2uV or 0.2ppm, so the output impedance is about 2mOhm. There is no visible additional voltage shift for over 30min under that load.
Cheers
Alex
« Last Edit: March 15, 2016, 03:54:56 am by Alex Nikitin »
Offline Andreas
« Reply #109 on: March 15, 2016, 09:05:36 am »
Hello Alex,
that is unbelievable for me.
I would expect a impedance at least 3 orders of magnitude higer (> 2-20 Ohms).
Are you shure that the cirquit is not oscillating?
(I had some problems with my FET voltage regulator until I put some bypass capacitors on input+output).
With best regards
Andreas
    • A.N.T. Audio
« Reply #110 on: March 15, 2016, 09:42:44 am »
Hello Alex,
With best regards
Andreas
With correctly designed output stage it is not difficult to get very low output impedance (at DC) and good stability. I have 220nF capacitor connected across the output and the circuit is completely stable. Here is the "live" recording of the loading experiment (unfortunately, at the usual NPLC100). You can see the points where the load connected and disconnected by the transient "spikes".
Cheers
Alex
« Last Edit: March 15, 2016, 09:53:32 am by Alex Nikitin »
Offline Andreas
« Reply #111 on: March 15, 2016, 04:22:38 pm »
Hello,
so its the (modified?) cirquit variant with the INA133 buffer from page 2
measured at the output of the INA and not directly at the FET?
with best regards
Andreas
    • A.N.T. Audio
« Reply #112 on: March 15, 2016, 10:07:30 pm »
Hello,
with best regards
Andreas
Hi Andreas,
No, all my latest data are for the version with 2N4391 JFET and Ultrohm resistors, from the end of November 2015. I've used OPA227 (actually, a dual OPA2227) for the 7V to 10V amplifier and output buffer. So the output impedance is for the buffer stage.
Cheers
Alex
Offline The Soulman
  • The sky is the limit!
« Reply #113 on: March 22, 2016, 09:26:17 am »
Hello,
with best regards
Andreas
Hi Andreas,
Cheers
Alex
Hi Alex, did you or maybe someone else did measurements on the relationship between vgs(off) and the current (and therefore resistor) required for optimum (as close to zero) tempco?
Could do it the hard (on the device..) way as in the first post using temperature extremes and a variable pot but would prefer a more scientific approach, at least to get in the ballpark.  :-+
    • A.N.T. Audio
« Reply #114 on: March 22, 2016, 10:52:01 pm »
There is no single formulae - the position of the zero tempco point depends on the JFET geometry, junction temperature and the Vds, so it is possible to make a rough prediction where that point should be only for a particular type/make JFET . For the lowest tempco at a certain temperature you need to adjust it at that temperature +/- few degrees. I will collect some data for 2N4391 but don't have any real statistics yet.
One really attractive point of the JVR is that the tempco can be adjusted directly for the main reference device and so tempcos of the rest of the circuit can be largely compensated for, around the temperature of interest - or even used to "flatten" the curve in 5-10 degrees range.
Cheers
Alex
Offline amspire
« Reply #115 on: March 23, 2016, 05:44:54 pm »
The old Natsemi Discrete semiconductor book is pretty useful - it has diagrams of the chip layout and characteristic data for each family of device.
http://www.introni.it/pdf/NatSemi%20-%20Discrete%20Databook%201978.pdf
The 2N4391 is a type 51 process.
This layout changed between the 1974 version of this data book and the 1978 version, so there may be different versions of the 2N4391 that behave sightly differently.
I wonder if a type 58 process FET would be lower noise - it has a much bigger area and lower resistance.
The 2N5432 (Type 58) has a pinch-off between 4 and 10V and a channel resistance 6 times lower then then the 2N4391. The package is a TO52 that is similar to the TO5 metal can.
A bit harder to get and more expensive. More likely to find them at a smaller distributor or on Aliexpress then the major distributors. I can get them for about US$6 here in Australia.
Richard
« Last Edit: March 23, 2016, 06:00:54 pm by amspire »
Offline zlymex
  • Country: cn
« Reply #116 on: April 07, 2016, 05:35:34 pm »
I have an old 491-page Chinese book(in pdf) called "High Stability Power Supply"where on page 121 the author derived a formula:
Vg0=Vp-0.66V
which means the zero tempco Vg is the Vp(Vgs off) minus 0.66V. And the corresponding current is
Id0 = Idss * (0.66/Vp)^2
I applied this to my DIY 100V voltage reference and it works fine.
« Last Edit: April 07, 2016, 05:45:38 pm by zlymex »
Offline Alex Nikitin
  • Femtoampnut and Tapehead.
    • A.N.T. Audio
« Reply #117 on: April 12, 2016, 07:13:31 am »
Vg0=Vp-0.66V
It is not a very accurate formula, even on the same make and type of JFETs the voltage difference between the cut-off voltage and "zero tempco" voltage varies according to my measurements (at the same temperature) and on top of that the cut-off voltage does vary with temperature, so this difference is temperature dependent as well!
Cheers
Alex
    • A.N.T. Audio
« Reply #118 on: April 12, 2016, 07:33:49 am »
I am struggling to measure the drift on this sample of the JVR. In three months it is less than I can reliably measure with the HP3458A (opt 002) and Fluke 731B as the second reference. The difference between the Fluke and JVR stays about the same for the same temperature (at 24C it is about 90-110 uV) and it is the same absolute value (+/- 1ppm) when measured with the HP after auto-cal over these three months (at the same room temperature), despite several moves between my home lab and my work lab. Here are today's measurements at home of the difference between the JVR and the Fluke, done by the Keithley 2015 with the upgraded input opamp used as a null meter at 100mV range and NPLC 10 with 10 averages.
Cheers
Alex
« Last Edit: April 12, 2016, 08:00:49 am by Alex Nikitin »
Offline The Soulman
« Reply #119 on: April 12, 2016, 08:04:27 am »
Vg0=Vp-0.66V
Cheers
Alex
In the Vishay journal AN103 (http://www.vishay.com/docs/70596/70596.pdf) they use nearly the same number 0.65volt (VGS(0TC)  VGS(off) – 0.65 V) that would probably get one in the ballpark,
however what Alex is saying does make sense and testing and trimming is the way to go for optimum thermal stability.
I need to get some parts ordered and start building one!
Offline quarks
Re: Building your own voltage reference - the JVR
« Reply #120 on: April 12, 2016, 02:52:41 pm »
Very interesting.
Now bookmarked.
    • A.N.T. Audio
« Reply #121 on: April 19, 2016, 08:28:54 pm »
Another update, this time ~200 hours run in my home lab, with the top temperature about 24C and some night time drops. The Keithley 2015 measures the difference between 10V JVR and Fluke 731B, 100mV range, 10NPLC + 10 averages, so the 100NPLC equivalent. 1ppm/div vertical scale. Again, this JVR build is quite stable in time.
Cheers
Alex
Offline amspire
« Reply #122 on: April 26, 2016, 06:22:19 pm »
I finally got around to a few tests myself. I had some MPF102 (plastic) so I decided to test them. The zero temp coefficient point was only 1.2V at 1mA, but this can be useful if you need a reference that can run from 5V.
The voltage sensitivity with a 5V supply was 0.003V/Vsupply. At 15V it reduced to 0.0013V/Vsupply. I am pretty happy with both numbers. As long as you maintain the voltage across the reference fet to 100mV, the voltage sensitivity even with a 5V supply is negligible.
Resistance sensivity is worse. For both 5V supply and 15V supply, the output varied by about 1/3 the resistor variation. This does not include the fact that as the resistor drifts, the FET current drifts and so it gets a temperature coefficient. The TO92 package is not good for temperature tests as you can get a positive or negative change depending on where the heat is applied to the package.
There are a number of 3ppm to 5ppm reference ICs for around $3. To equal this, I will need 10-15ppm resistors.
So on this score, it may not be worth the effort.
The other issue is the long term stability and temperature hysteresis. For reference ICs, it is not unusual to see a 50ppm drift spec  for 1st 1000 hour drift and another 50ppm drift if the chip is temperature cycled from 25degC to the min and max temperatures and back to 25 deg C. It would be interesting to know what kind of drifts and temp hysteresis you get with the JVR reference. It is certainly cheaper and easier getting a hermetically sealed JFET then getting a hermetically sealed voltage reference IC, but of course, the resistor needed for the JVR is probably not going to be hermetically sealed.
If it turns out the JVR is low drift/low hysteresis, it is interesting for custom jobs, especially if the JVR is put in an oven.
Anyone have some suggestions for economical and stable resistors in the 5ppm to 15ppm range?
« Reply #123 on: April 27, 2016, 01:43:13 am »
For surface mount, the Panasonic ERA series has 10 and 15ppm resistors at relatively low cost. They' range from 20p to a pound (GBP) each in small quantities [RS pricing, packs of 5 or packs of 100, 10k ohms 0.05% 10ppm to 0.1% 15ppm, 0603 and 0805 sizes].
Online Vgkid
« Reply #124 on: April 27, 2016, 02:51:55 am »
For through hole, TE Connectivity Neohm- YR1 series offer 15ppm tc, and are cheap.
Offline bingo600
« Reply #125 on: April 27, 2016, 03:34:43 am »
Ask Edwin for a quote - http://www.eevblog.com/forum/profile/?u=96921
/Bingo
    • A.N.T. Audio
« Reply #126 on: April 27, 2016, 05:24:46 am »
/Bingo
That is the best suggestion  :-+ . With his resistors and using a JFET in a metal case with high Vgs (like the 2N4391 in my current prototype) it should be possible to make a reference with less than 1ppm/C tempco in 5-6 degrees range and a very good long term stability (my sample has drifted less than 5ppm for last 3 months - at least that is the limit I can measure it against HP3458A Opt002 and Fluke 731B) .
Cheers
Alex
Offline amspire
« Reply #127 on: April 27, 2016, 09:53:14 am »
Alex, have you done any thermal hysteresis tests? I am really interested in that one. It seems to be the test that affects the IC's signifigantly. Say
room temp ->70degC -> 0 deg C (whatever your refrigerator freezer does) -> room temp.
If the JRV does not have hysteresis, it becomes very interesting.
Offline amspire
« Reply #128 on: April 27, 2016, 10:18:28 am »
Mouser stock these - around about 25c a resistor. Not bad. The specs do not mention 1000hr drift, so that needs testing.
http://www.mouser.com/ds/2/418/NG_DS_1773265_A-722435.pdf
Not sure about the higher voltage zero-TC FETS, but the 1.2V MPF102 I tested (it was the one I had) would get 5ppm/C with a 15ppm/C resistor. I guess you could find the zero TC point of the resistor and FET combined and get a lower TC over a limited range. Get a combined tempco of 1ppm, and then put it in an oven with 0.1 degC stability and it should be a nice reference.
« Last Edit: April 27, 2016, 12:09:55 pm by amspire »
Offline Kleinstein
« Reply #129 on: April 27, 2016, 11:35:56 pm »
The TC of the resistor might not be such a big problem, as one could find the point for combined zero TC. The trouble is that you need a different resistor value for each FET. So individual matching / adjustment is needed. So when a pot / digi-pot is used for adjustment two resistors are needed. Otherwise you may need odd individual values - e.g. specially made for the FET or combinations of 2 or more.
The more important point than TC  is stability of the resistor and possibly the FET also. If stable, even a high TC might be acceptable - though I would not go for a PT1000.
Offline Squantor
« Reply #130 on: May 17, 2016, 07:51:38 pm »
Hello All, I am a member of the tekscopes group and I read this interesting bit of information on finding the zero tempco point of a JFET with a curve tracer:
The 577 D1 also allows you to find the ideal bias point (zero temperature
coefficient) of a FET. You can do this with a 7CT1N and a storage 7000 scope
as well. Just display a set of curves and apply heat and/or cold to the case
of the FET. The curves above the ideal bias point will drift up (and get
stored as a smear). The curves below the ideal bias point will move down
(and also get stored as a smear). The ideal bias point curve will not move
so there will be no smear. It pops right out at you.
Original post here: https://groups.yahoo.com/neo/groups/TekScopes/conversations/messages/129217
It has a few nice curvetracer tricks.
    • A.N.T. Audio
« Reply #131 on: August 02, 2016, 07:22:38 pm »
Just a quick update, I've re-measured the JVR after 4 month with the same HP3458A opt 002 (ACAL DCV done in all cases just before measurements) and the result is about -4...-5ppm over that time at the same temperature (24C). As it is in the ballpark of the meter stability, I've checked also the JVR against my Fluke 731B at the same temperature and it shows about -3ppm drift and the Fluke's drift over the same period of time, measured on the HP3458A is about -1ppm at the same 24C. So it all sort of adds up nicely (whatever the real drift is in all three  ::) ). I might now do a hysteresis test to, say, +60C unpowered (we have an oven that runs at that temperature) and back to 24C.
Cheers
Alex
« Last Edit: August 02, 2016, 09:22:33 pm by Alex Nikitin »
Offline CJay
  • M0UAW
« Reply #132 on: August 02, 2016, 09:36:52 pm »
Cheers
Alex
Light sensitive?
now there's interesting, I wonder if it would be possible to build a 'photonic oven' to compensate for temperature drift by exposing it to varying intensity light....
Or I may just be nuts...
M0UAW
    • A.N.T. Audio
« Reply #133 on: August 02, 2016, 09:55:03 pm »
First results on the hysteresis (from 24C up to 60C for an hour, then back to 25C - in that hour the temperature in the lab gone up 1 degree) - about +1.5ppm and even that is likely to be  mostly a temperature change effect. That result is for the complete unit, including the 7 to 10V amplifier and buffers. The shift at ~54C was about -2mV or -200ppm, indicating the tempco of about -7ppm/C box. Near 25C the tempco is about 1ppm/C .
Cheers
Alex
P.S. - the voltage measurements graph is added - 1ppm per vertical division scale. While the JVR was in the oven, the Fluke 731B was connected and slowly drifted down with the lab temperature increase.
« Last Edit: August 02, 2016, 10:49:37 pm by Alex Nikitin »
The following users thanked this post: Muxr
    • A.N.T. Audio
« Reply #134 on: August 03, 2016, 03:46:17 am »
Here is the graph of the hysteresis after 60 min in a fridge (temperature of the unit went down to about 12C) . There was quite a bit of condensation on the unit what I've got it back to the lab. I left for home while the voltage was still rising so the complete data will be available only tomorrow. Here is what I have so far (the "shelf" at 9.999974 is again the Fluke connected in place of the JVR while it was in the fridge):
It looks like the voltage should be back within 2ppm at least.
Cheers
Alex
    • A.N.T. Audio
« Reply #135 on: August 03, 2016, 07:20:26 pm »
Here is the complete graph. It looks like between 12C and 60C I can not reliably detect the hysteresis - if it is there, it is below 2ppm.
Cheers
Alex
Online Marco
« Reply #136 on: August 03, 2016, 08:51:57 pm »
Susumu RG for less than 15 ppm and Stackpole RNCS for 15 ppm? Both passivated with an inorganic layer instead of the usual resin.
Offline montemcguire
« Reply #137 on: August 04, 2016, 04:27:54 am »
Agreed - the Susumu resistors are really nice, far better than typical NiCr resistors. They're also trimmed by laser through the glass layer, after it has been deposited, so there's a lot less mechanical strain left on the part to cause drift. The only issue is that they're tiny SMD chips, but they can be attached to larger thermal masses easily.
Edit: The RG series are nice, but the new URG series are even higher precision, along with a higher price.
« Last Edit: August 04, 2016, 04:30:24 am by montemcguire »
« Reply #138 on: August 08, 2016, 03:00:13 am »
The drawback to foil is when you're using the resistors with low frequency AC signals. The low TC of a foil resistor is done by bonding the foil to an engineered ceramic that has a complementary expansion coefficient. For DC, the current induced foil heating and the substrate will stabilize thermally, and the resulting TC is low. For very high frequencies, there's also no problem, since the waveform changes so quickly that the foil heating is constant over all parts of the waveform.
However, with low frequencies, the foil and the substrate will not heat uniformly over the entire waveform, and the foil can expand and contract differently at different parts of the waveform, causing the resistance to modulate with the signal voltage. What's worse is that the thermal time constants can cause the substrate to expand and contract out of phase with the foil, actually worsening the thermal effects of the signal current foil heating. This creates third harmonic distortion, and at just the wrong frequency, the effect is worse than simply using a 'normal' higher TC resistor that doesn't try to cancel its TC with an engineered ceramic substrate.
So, for audio or other LF AC uses where distortion is important, foil resistors can perform worse than traditional NiCr films like the Susumus, and can generate more distortion than the amplifiers around them. As far as I know, the Susumus have a low TC by tweaking the alloy and the processing, and not by using a trick substrate ceramic, so they'll have better LF distortion performance than a foil.
The following users thanked this post: chickenHeadKnob, EmmanuelFaure, bitseeker
Offline guenthert
« Reply #139 on: August 15, 2016, 04:44:50 pm »
Cheers
Alex
Not a practical application, but the idea (as far as I understand it) is also discussed here: http://www.analog.com/media/en/training-seminars/tutorials/MT-087.pdf
(p. 9). Which is not to say that it wouldn't be worthwhile to explore this further.
    • A.N.T. Audio
« Reply #140 on: August 16, 2016, 02:14:36 am »
That marketing document is OK, but it contains some inaccuracies that are most likely due the fact that ADI wants to sell you on the idea of an XFET reference.  Reader beware, and check your facts.  For example in figure 10, they show that an XFET reference is somehow "better" than a buried Zener reference for long term stability.  Sorry ADI, but this is just NOT TRUE.  I think that their XFET reference might be OK if you need low power, but I think they should be more honest.  This is not the first time ADI has made false assertions about it's voltage references.
It is an old paper (and not really relevant to the topic, as these XFET references use a "band-gap" type 2-FET structure). A single JFET reference would not have that good tempco over a wide temperature range, but can be adjusted for a near-zero tempco at a particular temperature and appears (providing a metal can device is used) very stable in time.
Cheers
Alex
    • A.N.T. Audio
« Reply #141 on: November 15, 2016, 10:37:00 am »
Another measurements before I take the JVR to the newly calibrated HP3458A. Taken with two Keithley 2015 and Keithley 617 as a thermometer (measuring the output of LM35D, 200mV = 20C, 10mV/degree, right scale). So far both JVR and Fluke 731B drifted less than 5ppm over the last year (even with all my attempts to measure a hysteresis). Interesting that the JVR has a low tempco between 22 and 26C
Cheers
Alex
« Last Edit: November 15, 2016, 10:40:40 am by Alex Nikitin »
Offline TiN
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
« Reply #142 on: November 15, 2016, 01:39:30 pm »
It is tempco of 2015, -0.9ppm/K? It's very high out of spec for 731B alone. :)
EDIT: I was thinking about 732B, sorry.
« Last Edit: November 15, 2016, 01:42:29 pm by TiN »
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
Offline David Hess
  • DavidH
« Reply #143 on: November 15, 2016, 07:36:02 pm »
Analog Devices has neatly solved this problem by buying Linear Technology.
    • A.N.T. Audio
« Reply #144 on: November 15, 2016, 10:45:43 pm »
Well, the result from the JVR measured by HP3458A at 25C lab temperature (after 1.5 hours in my backpack and 3 hours warm-up in the lab) is essentially the same as three and half months ago (see my posts from the 2nd/3rd of August) - 10.000055V with about 0.5 ppm variations. The unit was almost continously powered up all this time (except when it was transported from my home lab to my work lab and back).
Cheers
Alex
« Last Edit: November 15, 2016, 10:48:18 pm by Alex Nikitin »
    • A.N.T. Audio
« Reply #145 on: June 08, 2017, 09:37:32 pm »
As the HP3458A just came back from a repair and calibration at Keysight UK, I've brought the 10V JVR unit back to the lab and measured it after a good warm-up. Here is the result for 1 hour, at NPLC100 , combined with the similar one hour measurements at the same 25C temperature in August and December last year. I need to build few more of these and get some statistics ;) .
Cheers
Alex
« Last Edit: June 08, 2017, 09:50:59 pm by Alex Nikitin »
The following users thanked this post: evb149
Offline TiN
  • Country: tw
    • xDevs.com
« Reply #146 on: June 08, 2017, 11:11:35 pm »
I wish I'd have backpack big enough for the 3458A :D
    • A.N.T. Audio
« Reply #147 on: June 08, 2017, 11:15:23 pm »
I wish I'd have my own HP3458A. I would find a backpack large enough to put it in, no problem   ::) .
Cheers
Alex
Offline guenthert
« Reply #148 on: June 20, 2017, 08:13:59 pm »
That looks very promising.  Build many more and sell them and thereby distribute the work of collecting statistics!
Offline Ash
« Reply #149 on: June 20, 2017, 08:36:16 pm »
Just wanted to mention that I'm planning on building a few of these as well. I have some of the parts, still waiting on others. I'm also in the middle of setting up my meters and fixing some scanners so I can automate measuring of everything.  :-DMM
Looking forward to it having some fun!  :popcorn:
Ash.
Offline Vtile
  • Ingineer
« Reply #150 on: June 21, 2017, 05:00:33 am »
Yes, I've looked at that option, however it is only possible for devices with a high enough drain current
No need to use Vg=0 here, a bit of drift from a resistor divider for the temperature setting isn't that critical, so IDSS isn't a limiting factor.
Have anyone used the temp.co difference in the divider resistors to cancel out the reference tempco?? In theory (as two brain cycles without math) it should work.
PS. How is the long term spability now or is the prototybe trashed along the years.
« Last Edit: June 21, 2017, 05:05:25 am by Vtile »
    • A.N.T. Audio
« Reply #151 on: June 21, 2017, 05:23:58 am »
Look on the previous page. Essentially the voltage is quite stable, using the 3458A meter the difference is few ppm over about a year, certainly less than 20ppm drift for a complete reference including the 7 to 10 volts amplifier stage.
Cheers
Alex
Offline Vtile
  • Ingineer
« Reply #152 on: June 21, 2017, 05:25:29 am »
Sounds nice.. Silly me I'm still on page 4, it seems also that the resistor? zero tempco point is discussed there.  :-[  :palm:
Offline zhtoor
  • Banned!
  • Country: pk
« Reply #153 on: June 21, 2017, 05:27:20 am »
Cheers
Alex
very good results. have you tried cascading the fet's like in the 3-fet design paper?
regards and keep up the good work.
Online IconicPCB
« Reply #154 on: June 22, 2017, 08:09:13 am »
3 FET design?
Offline zhtoor
  • Banned!
  • Country: pk
« Reply #155 on: June 22, 2017, 08:16:25 am »
    • A.N.T. Audio
« Reply #156 on: June 22, 2017, 05:44:30 pm »
Thank you. For a voltage reference it is not necessary, as long as the supply voltage is steady enough (in my unit it is bootstrapped from the reference voltage, so no problems there). Introducing the second and the third FETs helps to increase the current source output impedance but in turn, adds new temperature dependencies of the second and the third order, so the FETs should be thermally coupled and ideally temperature stabilized as in the paper you've found.
Cheers
Alex
The following users thanked this post: vindoline
Offline Ash
« Reply #157 on: June 29, 2017, 10:07:35 pm »
Hi Everyone,
I just did a run to characterise the 2N4391 JFETs I have. They are all from Central Semiconductor with a date code of 1620.
I lashed up a circuit with a resistance decade box as the resistor between S and G / GND, and supplied 10V from an Aglient E3640A power supply. I measured the voltage at the source pin on my 6.5 digit Keithley 2015 (uncalibrated). Note that the resistance box is not hugely accurate, but it is sufficient to get me in the ballpark and I'll trim the operating point of each circuit from these values.
Basic process was to put the device in to a machine pin socket at room temperature (about 23.5C), and adjust the resistance box so that grabbing the metal can with my fingers, increasing the temperature didn't move the last digit of the voltage much (at most 1 or 2 counts).
FETResistanceVoltage
A87708.17287
B76007.71515
C86508.10104
D78707.83454
E83007.97806
F77007.77651
G69007.37061
H83708.06369
I65507.20610
J104008.56925
So the I(ds) at the 0 TC point for room temperature is floating around the 1mA point.
I'm also planning on testing these same FETs at an elevated temperature (say 50C) as I want to build ovenised references to combat the temperature swings in my office. In Brisbane it can get to 40C on bad days in summer... I'll add a precision OpAmp as a buffer, also in the oven. I'm then considering my options for 10V generation. I'm leaning towards PWM + another OpAmp with a 2x gain and a filter... Then it is programmable..
Ash.
The following users thanked this post: Alex Nikitin, 2N3055, zhtoor
Online Marco
« Reply #158 on: June 30, 2017, 12:32:37 am »
If you go the PWM route then given the negligible hysteresis wouldn't it make more sense to digitally compensate for temperature than ovenise?
Offline MK
Offline Ash
« Reply #160 on: June 30, 2017, 11:16:37 am »
Hi Marco,
The thought did cross my mind and I've done similar software based temperature compensation for things before, but I had access to a temperature chamber then and I don't now :)
Maybe in the future, but for the initial version, I'll probably ovenise the critical sections, possibly the whole lot. I'll then perturb the oven temperature (say +/-5 deg C) and use that to determine and set the overall system 0 TC point as I should be able to (at least to first order) use the JFET to offset any other system TC.
I haven't yet put much thought into the 10V generation, and it may prove too much trouble, but I'd like to be able to generate an arbitrary voltage in a range (0-12V?), basically like a calibrator. Will have to see how that plays out also with the overall temperature compensation as well. However first I have to get some confidence and understanding of the reference itself. I'll also have to think about the noise implications as well...
I'm also going for something that can be built and calibrated with minimum of equipment (basically a good DMM and/or null against a +10v reference). I'm not making any quantity so I'm happy to spend the time tweaking the TC of each unit. I should be able to do this by setting up PWM for approximately 10V, then monitoring the output voltage as I slew the oven temperature. Once the TC is tamed, then I would allow the over the settle mid-range for a while and then calibrate the 10V PWM against a suitable reference.
http://shop.kuhne-electronic.de/kuhne/en/shop/amateur-radio/accessoires/crystal-heater/Precision+crystal+heater+40%C2%B0+QH40A/?card=724
That might help stabilise the JVR, as it is quite small.
Hi MK, thanks for the link. Interesting. I'll probably end up with a bit larger of an oven than that is for. But saved for future reference - thanks!
Ash.
Offline 0.01C
  • Country: hk
« Reply #161 on: June 30, 2017, 11:52:35 am »
[/quote]
Ash.
[/quote]
A Chinese forum  bbs.38hot.net   ,some people selling their diy oven and reference board , just CNY 200+
Guildline Datron Fluke L&N ESI HP
0.01C
Online Marco
« Reply #162 on: July 01, 2017, 12:59:13 am »
Even without temperature control you still want all the critical components to be in a predictable environment which quickly gets to steady state, which is to say inside a high thermal conductivity housing. The same as with an oven, it just doesn't need insulation. Put some high wattage resistors on it and make your own temperature chamber.
« Last Edit: July 01, 2017, 01:01:54 am by Marco »
Offline alm
« Reply #163 on: July 06, 2017, 01:42:05 am »
Would a heater without any insulation not result in a large temperature gradients, something even worse than temperature fluctuations?
Offline 1001
« Reply #164 on: August 31, 2017, 02:16:46 am »
Cheers
Alex
Using  the same jfet in plastic (to-92) case how worse could be the TC and long-term stability?
Could be a good idea to consider power jfet with lower on resistance then small jfet?
Offline Kleinstein
« Reply #165 on: August 31, 2017, 02:31:24 am »
A plastic case JFET can have much higher long time drift due the humidity in the plastics. As this reference also needs a rather stable (and thus maybe more expensive than the FET) resistor, I see little sense in saving a little one the FET. The 2N4391-3 are still available in TO18 case.
A higher power FET could result in lower noise, just like using several in parallel.
The following users thanked this post: cellularmitosis, bitseeker, 1001
    • A.N.T. Audio
« Reply #166 on: October 03, 2017, 09:28:01 pm »
Tomorrow will be two years since I've started this thread. The 10V JVR unit is back in the lab and connected to the HP3458A. The reference was not powered for two weeks in September when I was away  for a holiday in Canadian Rockies and Vancouver. It was not affected much (if at all) over this time and came back to the exactly the same voltage as it was before the power interruption. The measured voltage today after 2h warm-up is 10.00008 V at 24C (previous measurements were taken at 25C). Overall, the drift of a complete reference (including the 7V to 10V amplifier part) appears to be less than 5ppm per year (+ whatever the meters uncertainty and drift is, so in total probably better than 10ppm/year ), measured by HP3458A Opt 002 and also compared with the Fluke 731B which looks very stable.
Cheers
Alex
« Last Edit: October 03, 2017, 10:28:15 pm by Alex Nikitin »
The following users thanked this post: 3roomlab, 1001, BNElecEng
Offline BNElecEng
« Reply #167 on: November 03, 2017, 04:20:51 am »
Hi Alex,
I've read through the entire thread of this topic. You've got an excellent low-cost voltage reference design by the looks of it. I'm hunting down the parts to make my own now.
I was curious if you had had the time to make multiple copies of the same design to use as references against each other, or are you only using your fluke 731B for this purpose?
Thanks in advance
    • A.N.T. Audio
« Reply #168 on: November 03, 2017, 04:30:00 am »
Hi Alex,
Thanks in advance
I'm using 731B, HP3458A and an LTZ1000 based reference for cross-checks. The HP3458A Opt 002 (4ppm/year) was calibrated in May this year.
I plan to make a proper pcb which would accept either a JFET or LM399 or LTZ1000, and will slide in a standard Hammond case.
Cheers
Alex
The following users thanked this post: RandallMcRee
Online RandallMcRee
« Reply #169 on: November 03, 2017, 04:44:22 am »
Sounds like a winning pcb. I've been following this too, and have some jfets ready to go. Can you make the pcb available on (for example) osh?
Per the ltz1000 I think you should stick to the datasheet circuit as opposed to the more elaborate KX one (my vote--obviously do what you like).
Thanks!
    • A.N.T. Audio
« Reply #170 on: November 03, 2017, 05:11:42 am »
I have my own ideas  ;) .
Cheers
Alex
Offline iainwhite
  • Measure twice...
« Reply #171 on: November 07, 2017, 06:40:15 am »
Is the schematic for the latest reference the same as the one in Reply #49 on Page 2, except with a 2N4391 JFET and an OPA227 OpAmp?
I would like to try building it.
Thanks for an interesting thread.
    • A.N.T. Audio
« Reply #172 on: November 07, 2017, 09:17:20 am »
Here is the latest schematics (for the unit running for 2 years now). R1 is selected for the zero tempco point, R5 and R6 for 10V output.
Cheers
Alex
« Last Edit: November 07, 2017, 09:19:02 am by Alex Nikitin »
The following users thanked this post: BravoV, iainwhite, 2N3055, 1001, Ash, HoracioDos, zhtoor, denimdragon, razvan784
Offline BNElecEng
« Reply #173 on: November 07, 2017, 09:26:22 am »
Hi Alex,
Is J2 used to start up the circuit? I'm assuming it pulls up the input of the op amp when power is applied, in order to generate a voltage to allow J1 to begin regulating.
Thanks in advance
    • A.N.T. Audio
« Reply #174 on: November 07, 2017, 09:32:08 am »
Hi Alex,
Thanks in advance
Yes.
Cheers
Alex
The following users thanked this post: BNElecEng
« Reply #175 on: November 07, 2017, 09:46:35 am »
Hi Alex,
Thanks in advance
Ha! You beat me to it. I was just about to type out the very same question.
Offline BNElecEng
« Reply #176 on: November 07, 2017, 09:48:14 am »
  :) you know what they say about great minds
Offline iainwhite
  • Measure twice...
« Reply #177 on: November 08, 2017, 01:44:53 am »
I appreciate you making this available - Thank you.
« Reply #178 on: November 08, 2017, 02:02:09 am »
Yeah, and also, sadly, about fools.  :)
Online vindoline
« Reply #179 on: March 10, 2018, 01:33:27 am »
Cheers
Alex
Hi Alex, thanks for posting your schematic. I'm thinking of building one up and I have a question. R3 on the output is marked "k12" I'm assuming this is a typo and it should be 1k2? Thanks!
    • A.N.T. Audio
« Reply #180 on: March 10, 2018, 01:38:24 am »
R3 is 120 Ohm .
Cheers
Alex