I'm tagging onto this thread because my post is related, and I have been exchanging notes recently with the OP.
This is about my recent acquisition of a PRS-50 - and it arrived today. The unit is immaculate inside, and clean outside with just a few scuffs. Operationally, though, "no joy". But there's enough hope to encourage me to move forward with a few fundamental repairs before throwing in the towel. When the unit first arrived, I decided to go for broke and connect 2 independent 48VDC supplies to the rear panel. I had previously set up monitor3 and connected up a working USB-to-RS232 cable to my laptop. Initially, the DC POWER LED on the front panel illuminated - although suspiciously not with what I would call normal brightness - and then after about 10 seconds all of the front panel LEDs (except for the 2 bad fuse indicators) illuminated for another 10 seconds, and then finally the LEDs started to blink on and off in what appeared to be a fairly complicated, but repeating, sequence. Again, all of the LEDs were a bit dim in my opinion when they did illuminate. Meanwhile, there was absolutely no activity on the RS232 ports, no matter what I did. So I suspected a power supply problem and decided to pull the main module bearing the CBT. Removing that module was rather easy, and I gained an appreciation for the layout of the build that apparently had serviceability in mind.
Without any schematic to know what voltages ought to be at which test points, I chose to probe 3 places that were obvious. Two of those places were the inputs to the two HV bricks, for which the input voltages were each marked on the respective brick. The other was an LM340T-12 linear regulator that clearly fed the OCXO. The results confirmed power supply problems:
- Input to HV brick 1: 10.8v (should be 24)
- Input to HV brick 2: 6.65v (should be 10)
- Output of LM340T-12: 5.83v (should be 12), problem due to only 7.37v into the regulator
Further checking how many other test points are wrong was immaterial at this point, since it's obvious that the power supply card has issues. So ... time to take the cover off to the compartment housing the power supply and take a look at the board. After gaining access to the board I, of course, discover all but 2 of the 16 larger electrolytics were badly swollen and about 11 of them were oozing electrolyte. Been down that road plenty! The good news is that electrolyte was simply oozing out of the tops of the caps and absolutely none of it got on the PCB or anywhere else. Once I order some replacements I'll make the repairs and test the voltages at the caps themselves. Somewhere on eevblog I recall that someone posted a "map" of the caps and the voltages on them from his working unit (may not have been a PRS-50 but one of its siblings.). I'll dig that up to guide my testing once I replace the caps.
More after I have a working supply. Perhaps there's some good news here - though it may end up being just be wishful thinking on my part - and that is that the tube actually may still have some life left in it, and that it was the power supply failures that led to the unit's retirement. Perhaps that retirement was fortuitous for the tube.
Jim