Author Topic: DMM versus temperature comparison  (Read 2500 times)

0 Members and 1 Guest are viewing this topic.

Offline rigrunnerTopic starter

  • Frequent Contributor
  • **
  • Posts: 261
  • Country: gb
DMM versus temperature comparison
« on: November 25, 2017, 05:21:19 pm »
It's the time of year when my lab temperature gets to wander about at the chillier end of things and i thought it would be interesting to see how the different DMMs compare with lab temperature change.  All i've done is set the EDC to 10V and take a measurement at 30 minute intervals with NPLC at 100 for the HP meters.  Both of the 3457A are soft calibrated when the temperature change is  +/- 1C.

The 3456A is remarkable and the Datron 1071 would make a pretty good thermometer :)

Internet of Things: A solution desperately trying to find its problem
 
The following users thanked this post: e61_phil, doktor pyta, CalMachine

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3238
  • Country: de
Re: DMM versus temperature comparison
« Reply #1 on: November 25, 2017, 05:36:35 pm »
Hello,

as comparison:

My K2000 has around 0.5ppm/K in the 10V range.

https://www.eevblog.com/forum/metrology/project-pimp-a-keithley-2000/msg1106839/#msg1106839

And no significant correlation to temperature in the 100 mV range.

with best regards

Andreas
 

Offline rigrunnerTopic starter

  • Frequent Contributor
  • **
  • Posts: 261
  • Country: gb
Re: DMM versus temperature comparison
« Reply #2 on: November 25, 2017, 06:02:43 pm »
Using the down temperature slope from reading 89 onwards the DMMs here show:

3456A: -0.5ppm/K
3457A-1:  +0.83 ppm/K
3457A-2: -1.16 ppm/K
D1071: -2.0 ppm/K

I'd expect the 3457As would be slightly better in their 3V range. I wasn't going to check any other ranges but now that you've mentioned it i might flip the EDC to 100mV and log a little data there too.


Internet of Things: A solution desperately trying to find its problem
 

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5468
  • Country: de
Re: DMM versus temperature comparison
« Reply #3 on: November 25, 2017, 06:18:12 pm »
How do you know that your 10V reference supply was stable during the change of temperature?
There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline rigrunnerTopic starter

  • Frequent Contributor
  • **
  • Posts: 261
  • Country: gb
Re: DMM versus temperature comparison
« Reply #4 on: November 25, 2017, 06:37:57 pm »
How do you know that your 10V reference supply was stable during the change of temperature?

For this set of measurements i don't know the stability of the 10V. All i was interested in was a comparison of the meters,. Whatever drift the EDC has is included in the measurements. Before i switch my 1271 over to the new build LTZ i can measure the 10V over a temperature range and get an idea of it's drift.

Internet of Things: A solution desperately trying to find its problem
 

Offline maxwell3e10

  • Frequent Contributor
  • **
  • Posts: 869
  • Country: us
Re: DMM versus temperature comparison
« Reply #5 on: November 26, 2017, 06:45:52 am »
I looked at temperature sensitivity of Keithley DMM7510A and 34470A for periodic temperature changes on a timescale of 1000 sec:
https://www.eevblog.com/forum/metrology/dmm-adc-noise-comparison-testing-project/msg1330643/#msg1330643

For 34470A the measured sensitivity for these relatively fast temperature changes was about 0.8 ppm/K, while for DMM7510 it was 0.4 ppm/K. The specs for steady-state temperature change are 2 ppm/K for 34470A and 0.15ppm/K for DMM7510.

Someone must have done this for a 3458A?
 

Offline 3roomlab

  • Frequent Contributor
  • **
  • Posts: 825
  • Country: 00
Re: DMM versus temperature comparison
« Reply #6 on: November 26, 2017, 08:03:37 am »
It's the time of year when my lab temperature gets to wander about

how about sticking the VREF in a wine box?
https://www.eevblog.com/forum/metrology/dmm-adc-noise-comparison-testing-project/msg1136604/#msg113660
 

Online Dr. Frank

  • Super Contributor
  • ***
  • Posts: 2380
  • Country: de
Re: DMM versus temperature comparison
« Reply #7 on: November 26, 2017, 12:13:51 pm »
I looked at temperature sensitivity of Keithley DMM7510A and 34470A for periodic temperature changes on a timescale of 1000 sec:
https://www.eevblog.com/forum/metrology/dmm-adc-noise-comparison-testing-project/msg1330643/#msg1330643

For 34470A the measured sensitivity for these relatively fast temperature changes was about 0.8 ppm/K, while for DMM7510 it was 0.4 ppm/K. The specs for steady-state temperature change are 2 ppm/K for 34470A and 0.15ppm/K for DMM7510.

Someone must have done this for a 3458A?

Well, yes, it's described here: https://www.eevblog.com/forum/metrology/ultra-precision-reference-ltz1000/msg1336575/#msg1336575, how to determine the 3458As T.C., and how to design a reference which is relatively independent from temperature changes, serving as a fix point for such measurements.

I've got a 3458A which has +0.4ppm/K, and Tins both 3458As are less than 0.1ppm/K.

Frank
 

Offline rigrunnerTopic starter

  • Frequent Contributor
  • **
  • Posts: 261
  • Country: gb
Re: DMM versus temperature comparison
« Reply #8 on: November 26, 2017, 05:10:24 pm »
how about sticking the VREF in a wine box?

I've got a wine cooler similar to the thing Dave showed in episode #101 but the EDC is too large to fit inside :(
I will redo the measurements with an LTZ in the cooler when i've put the next one together.

 
Internet of Things: A solution desperately trying to find its problem
 

Offline Mickle T.

  • Frequent Contributor
  • **
  • Posts: 467
  • Country: ru
Re: DMM versus temperature comparison
« Reply #9 on: November 26, 2017, 06:49:43 pm »
R6581T: +1...+1.2 ppm/K (w/o selfcal)
D1271: +0.5 ppm/K (w/o selfcal)
S7081 #1: +0.06 ppm/K (20K span, LTZ mod + Cu mod)
S7081 #2: +0.03 ppm/K (20K span, LTZ mod + Cu mod)
S7081 #3: -0.55 ppm/K
HP 34401A: +0.1 ppm/K
K2100: -1...-1.5 ppm/K
D1071: +0.8 ppm/K
« Last Edit: November 26, 2017, 06:58:34 pm by Mickle T. »
 
The following users thanked this post: quarks, e61_phil, doktor pyta

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 962
  • Country: de
Re: DMM versus temperature comparison
« Reply #10 on: November 28, 2017, 08:33:22 am »
I've attached a measurement from a Solartron 7081 I had some time ago. It was measured against my Fluke 5440B (10V). It gives a bit less than 0.8ppm/K
 
The following users thanked this post: Mickle T.