Author Topic: Measuring nanoamps and below like a Ninja  (Read 24820 times)

0 Members and 1 Guest are viewing this topic.

Offline chuckb

  • Regular Contributor
  • *
  • Posts: 213
  • Country: us
Re: Measuring nanoamps and below like a Ninja
« Reply #100 on: November 14, 2018, 09:43:21 am »
With +31 V bias the MEDER HI05-1A66 had 10.5 fa of leakage across the contacts. With -31 V bias it had -12.5 fa of leakage. So the average leakage at 31 V bias is 11.5 fa. This equates to over 2000 T ohms isolation resistance. I have previously measured the contact to coil leakage current being 1/3 of the contact to contact leakage current. This relay has been sitting on my bench for several month with no cleaning before this test. Hope this helps.
 
The following users thanked this post: Echo88, e61_phil

Offline chuckb

  • Regular Contributor
  • *
  • Posts: 213
  • Country: us
Re: Measuring nanoamps and below like a Ninja
« Reply #101 on: November 16, 2018, 03:23:38 pm »
I have also tested the  COTO 9012. This is a very small relay for less than $10 USD. The contact isolation resistance is 4000 T ohms. The coil to contact resistance is 1000 T ohms at 100V.

 

Offline Vtile

  • Frequent Contributor
  • **
  • Posts: 957
  • Country: fi
  • Ingineer
Re: Measuring nanoamps and below like a Ninja
« Reply #102 on: November 30, 2018, 07:52:14 am »
With my Marconi / Meratronic Meratester I'm in like a Flyn. Sensitivity (noticeable relative change) is somewhere in region of 50 fA.  ^-^




 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1243
  • Country: de
Re: Measuring nanoamps and below like a Ninja
« Reply #103 on: December 01, 2018, 01:28:05 am »
We do have one of those brown bricks, Keithley 617. Nice piece of kit. I ordered some Ohmite high value 1% metal oxide resistors (100meg, 1G, 10G) and build a small setup with the resistors mounted on teflon standoff. After all these years without being calibrated the readings on the meter are still consistent.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 3790
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: Measuring nanoamps and below like a Ninja
« Reply #104 on: December 23, 2018, 07:50:05 am »
Accidentally bought myself Xmas gift...  :o



Source is Keithley 6221.

Oh well... I hope buying not broken modern equipment will not become a habit.  :-//
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
 
The following users thanked this post: Inverted18650

Offline Inverted18650

  • Frequent Contributor
  • **
  • Posts: 875
  • Country: us
    • Test Gear Addicts
Re: Measuring nanoamps and below like a Ninja
« Reply #105 on: December 23, 2018, 11:11:02 am »
TiN, I am going to message you about the Keithley 417 PicoAmmeter w/ 4170 input head.
« Last Edit: December 23, 2018, 11:18:14 am by Inverted18650 »
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 3790
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: Measuring nanoamps and below like a Ninja
« Reply #106 on: December 23, 2018, 07:04:30 pm »
Thanks for offer, Inverted18650, but perhaps some other eager amp-nut would like that 417. I already have enough of backlogged projects.  :scared:
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
 
The following users thanked this post: Inverted18650

Online MasterTech

  • Frequent Contributor
  • **
  • Posts: 852
  • Country: 00
Re: Measuring nanoamps and below like a Ninja
« Reply #107 on: December 23, 2018, 07:22:01 pm »
That is a beast, can handle fA like nothing wow, minor teardown at least coming? ::) ::)
 

Offline VintageNut

  • Frequent Contributor
  • **
  • Posts: 517
  • Country: 00
Re: Measuring nanoamps and below like a Ninja
« Reply #108 on: January 07, 2019, 05:17:26 am »
I believe that I have shared these resistors on a different thread. For this thread, these resistors are intended to be used with an electrometer to check calibration points.

One end of each resistor has a BNC jack. The jack is for the voltage source of the electrometer. Any leakage from HI to LO through this jack is not measured. So, this jack can be a junk BNC with no negative effect on it intended use.

The other end of the resistor is a 3-lug triax jack. This jack is for the triax jack of the electrometer.

These resistors can also be used with a sourcemeter. Care has to be taken to account for any leakage (if any) associated with the BNC jack.   

I had these calibrated at a primary lab in 2017. I will be taking them back again for a cal in February of this year.

I have no idea about temp dependence of these. That info would be needed for any rigorous cal usage. 
working instruments :Keithley 260,261,2750,7708, 2000 (calibrated), 2015, 236, 237, 238, 147, 220,  Rigol DG1032  PAR Model 128 Lock-In amplifier, Fluke 332A, Gen Res 4107 KVD, 4107D KVD, Fluke 731B X2 (calibrated), Fluke 5450A (calibrated)
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 723
  • Country: de
Re: Measuring nanoamps and below like a Ninja
« Reply #109 on: January 07, 2019, 06:11:44 am »
Could you say some words about the calibration in the lab? How did they cal these resistors?
 

Offline VintageNut

  • Frequent Contributor
  • **
  • Posts: 517
  • Country: 00
Re: Measuring nanoamps and below like a Ninja
« Reply #110 on: January 07, 2019, 09:29:07 am »
Could you say some words about the calibration in the lab? How did they cal these resistors?

I can sketch the methodology, but I do not have total command of the calibration procedure.

It was explained to me that the calibration system is a bridge. The bridge uses a very old collection of Keithley 5155 resistors that have a long calibration history. The bridge also has an old Keithley electrometer. The remainder of the bridge is unknown to me and the exact procedure is also unknown. I will try to find out more details when I visit in February.

Attached are the calibration reports


« Last Edit: January 07, 2019, 09:38:49 am by VintageNut »
working instruments :Keithley 260,261,2750,7708, 2000 (calibrated), 2015, 236, 237, 238, 147, 220,  Rigol DG1032  PAR Model 128 Lock-In amplifier, Fluke 332A, Gen Res 4107 KVD, 4107D KVD, Fluke 731B X2 (calibrated), Fluke 5450A (calibrated)
 
The following users thanked this post: e61_phil

Offline VintageNut

  • Frequent Contributor
  • **
  • Posts: 517
  • Country: 00
Re: Measuring nanoamps and below like a Ninja
« Reply #111 on: January 07, 2019, 09:37:34 am »
I double posted and missed one. Here is the missing one.

working instruments :Keithley 260,261,2750,7708, 2000 (calibrated), 2015, 236, 237, 238, 147, 220,  Rigol DG1032  PAR Model 128 Lock-In amplifier, Fluke 332A, Gen Res 4107 KVD, 4107D KVD, Fluke 731B X2 (calibrated), Fluke 5450A (calibrated)
 

Offline VintageNut

  • Frequent Contributor
  • **
  • Posts: 517
  • Country: 00
Re: Measuring nanoamps and below like a Ninja
« Reply #112 on: January 07, 2019, 10:10:59 am »
Since you asked about resistor calibration, attached is a calibration report for my Fluke 5450A. It was calibrated against two Measurements International bridges. The low ohms are calibrated with a constant current bridge and the high ohms are calibrated with a constant voltage bridge. You can see that the 100M resistor in the 5450A is measured with better than 10X uncertainty compared to the 100M stand-alone resistor that I built.

Every measurement on the MI bridge requires a soak time of about 30 minutes. Very time consuming.
working instruments :Keithley 260,261,2750,7708, 2000 (calibrated), 2015, 236, 237, 238, 147, 220,  Rigol DG1032  PAR Model 128 Lock-In amplifier, Fluke 332A, Gen Res 4107 KVD, 4107D KVD, Fluke 731B X2 (calibrated), Fluke 5450A (calibrated)
 

Offline Inverted18650

  • Frequent Contributor
  • **
  • Posts: 875
  • Country: us
    • Test Gear Addicts
Re: Measuring nanoamps and below like a Ninja
« Reply #113 on: January 07, 2019, 08:34:24 pm »
That is a beast, can handle fA like nothing wow, minor teardown at least coming? ::) ::)

We're you asking me or TiN?  :palm:

Online MasterTech

  • Frequent Contributor
  • **
  • Posts: 852
  • Country: 00
Re: Measuring nanoamps and below like a Ninja
« Reply #114 on: January 07, 2019, 11:26:47 pm »
That is a beast, can handle fA like nothing wow, minor teardown at least coming? ::) ::)

We're you asking me or TiN?  :palm:
Oh I was asking TiN, that shiny new equipment calls for a teardown
 
The following users thanked this post: Inverted18650