Hi,
today I wanted to connect my Fluke 5440B to my GPIB "network", but I had some problems. First of all: If I connect the Fluke 5440B to my GPIB chain the VISA stuff will find instruments starting from the Fluke 5440B GPIB adresse. (if it is configured to 20 then 20, 21,..., 30 will appear in NI MAX). All other GPIB devices will work without distortions.
Then I tried to send some commands to the 5440B. However, the 5440B shows "IEEE488 REMOTE ERROR EXPECTING HEADER" and nothing happens. The same commands will work in the company with the Fluke 5440B from the company. The difference between here and there is the GPIB adaptor. I'm using a NI PCI GPIB card and at the company we have some NI USB dongles. But I doubt that is the problem.
I already read the manual twice and split my GPIB bus to reduce cable length. But nothing works. Is there something broken on my 5440B or is there something I overlook?
Thanks
Philipp