Author Topic: Fluke 5500A Multifunction Calibrator: Short Teardown  (Read 4786 times)

0 Members and 1 Guest are viewing this topic.

Offline Echo88Topic starter

  • Frequent Contributor
  • **
  • Posts: 826
  • Country: de
Fluke 5500A Multifunction Calibrator: Short Teardown
« on: February 21, 2018, 06:04:53 pm »
Today i had the chance to take a few photos of a defective Fluke 5500A which we have at work. I couldnt find the error and will reassemble it tomorrow, so it can be repaired by Fluke. So if there are any needs for more photos of a particular board say it now. Enjoy.  :)

Outer case photos and manuals can be found here: https://doc.xdevs.com/doc/Fluke/5500A/ Also pictures of the Keyboard Assembly, which i didnt take photos of.

Time limited repair attempt failed, but heres the repair-thread: https://www.eevblog.com/forum/repair/fluke-5500a-startup-error/

« Last Edit: February 21, 2018, 06:38:30 pm by Echo88 »
 
The following users thanked this post: Lexy, TiN, Andreas, doktor pyta, l3VGV, TizianoHV

Offline Echo88Topic starter

  • Frequent Contributor
  • **
  • Posts: 826
  • Country: de
Re: Fluke 5500A
« Reply #1 on: February 21, 2018, 06:06:27 pm »
The reference on A6_2 shows the LTFLU-1ACH flying in the breeze, they didnt even bother to squeeze it in a heated enclosure. Of course its sufficient to this mediocre class of calibrators but it kinda hurts to see.
« Last Edit: February 21, 2018, 06:18:15 pm by Echo88 »
 
The following users thanked this post: TiN, doktor pyta

Offline Echo88Topic starter

  • Frequent Contributor
  • **
  • Posts: 826
  • Country: de
Re: Fluke 5500A
« Reply #2 on: February 21, 2018, 06:13:06 pm »
All in all it was interesting to disassemble it, but of course it doesnt come close to a Fluke 57XX Calibrator with all its shielding and TaN-Resistornetworks.  ;D
« Last Edit: February 21, 2018, 06:21:22 pm by Echo88 »
 
The following users thanked this post: TiN, doktor pyta

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Fluke 5500A Multifunction Calibrator: Short Teardown
« Reply #3 on: February 22, 2018, 12:14:43 am »
Thank you for photos.

LTFLU in these is very same REF as used in Fluke 8842A, even resistor is same ;).

« Last Edit: February 22, 2018, 04:19:32 am by TiN »
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 
The following users thanked this post: doktor pyta, mycroft

Offline dl1640

  • Regular Contributor
  • *
  • Posts: 222
  • Country: cn
Re: Fluke 5500A Multifunction Calibrator: Short Teardown
« Reply #4 on: February 23, 2018, 12:48:49 am »
Looks A6 was repaired, residual flux on U45.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo