Fluke released 732C/734C without any fuss.The internal wire-wound-based resistors have been replaced by hermetic thin-film resistive networks, which are less prone to time- and temperatureinduced drift.
Not sure how I feel about this
They also give a glimpse on the guts. Good old Linear Technology LTFLU-1ACH on ceramic substrate with AD706 and OP07(?).
LTFLU chip manufactured in 2014 from date code, so I guess we can estimate when Fluke started testing for this project.
797933 hermetic film resistor network likely used to generate additional 0.1V range. It is not actively driven, as we can see from max current spec for this channel (20pA).
DC stability spec for 732C is same as 732B, as expected, but now 1V output have specified stability for 90 days / 1 year.
New 0.1V output specified for 1.2ppm/30d or 9.8ppm/1y and TC +/-0.2 uV/V. Hysteresis specified at 0.25ppm, same as 732B.
Now calibration procedure also recommends Keithley 2182A instead of old Fluke 845AB.
And as with modern trend
, no details on service information or schematics in manuals. I bet 734C box needed lot of development work to get that logo badge changed to new fancy yellow
Anybody getting DEMO unit?