So the topic boils down to the question on how to derive low uncertainty decade values starting from 10VDC and 10KOhm (up and down).
I do this using a Flue 752A for Volts, respectively a 3458A for the 10V to 1V transfer (based on the very good A/D linearity).
For Ohms I have in fact started with the ESI SR1010 but I have now moved to the Guildline 9975, at least for the lower values up to 100k.
The issue is how the uncertainties of all steps in a transfer pile up and result in lowest uncertainties achievable without having to use someting calibrated besides the two standards (e.g., the 752A is self-calibrating, needs to be done before use). So yes, SR1010 for 1/100 or 1/10 transfer, or 9975 for 1/10 transfer, or 1281 for 1:1 transfer. Keep in mind, for the SR1010s, you still need a 1:1 transfer meter.