10V with 1LSD error, produce 3LSD on 30V scale; i had 5 such devices (now 2) , and none of them drift more then 1LSD in 3 years on 30V voltage range. they drift most on a resistance range.
black HP label - usually on the latest devices; earliest units has a dual color
you may need replace battery, there sequence in a forum somewhere ; i suggest to replace capacitors on a power filter. shorten on 30 mv range should produce noise within 0.3-0.7 micro-volts; if it over 1.5 -3 micro volts - 100% caps. replace it ; and double it capacity . a huge cap on bottom , it GPIB filter. - i didn't touch it ...
you may find a range resistors better TC like 5PMM to constrict 30ohm 300ohm etc ... and send someone for the favor they can measure it.
then construct 10V ; then 30V; then build a voltage dividers with low TC, then 5.5 became not enough and you buy HP3456A (use this service: www.crossborderpickups.ca)
and finally you done ....
Quick sanity test, after the unit has been powered up for a bit more than an hour. Temperature in the room has been a steady 27.65C, an average from two side-by-side 1-wire probes in 12-bit mode. Relative humidity is about 45%.
Ohms tests:
30K 1% through hole resistor was the closest I grabbed. DMM reads 29.9646k.
1k 5% through hole resistor next measured 0.97591k
60.4ohm 1% through hole measures 61.339 on 2-wire, 60.503 on 4-wire
30.2ohm (60.4 1% || 60.4 1%) measures 31.853 on 2-wire, 29.6518 on 4-wire
22ohm 1% 0805 measures 23.2127 on 2-wire. Can't get on to it with 4 probes.
DC tests, using my DP832:
PSU 10mV reads as 13.0579mV
PSU 100mV reads 103.039mV
1V reads 1.00247V
10V reads 10.0024V
30V reads 30.0035V
DC 30mV range shorted produces 0.00000V on the display
DCA tests using 100ohm 5% 1/4W + DP832:
5V = 48.975mA on the 3478a
1V = 9.757mA
ACV 28V reads 27.9020VAC. BK Precision 2709B reads 28.09.
I've read about shorted power supplies/blown fuses on these due to failed caps. It's on my list.
3456... Don't know that I could sneak that by the wife