Author Topic: HP8592A Correction Factors issue etc.  (Read 1542 times)

0 Members and 1 Guest are viewing this topic.

Offline vxhTopic starter

  • Newbie
  • Posts: 1
  • Country: it
HP8592A Correction Factors issue etc.
« on: March 01, 2017, 07:34:48 pm »
Hi guys,

I bought an HP8592A with several problems.
I tried to search the service manual and I found all other 85XX versions, nothing about the 8592A.
On power on, the grid with annotations come up with no signal or noise line. I made calibration of Freq, Amptd, YTF, stored all but after restart the problem is the same.
If I switch Corr to Off, noise floor comes up with a 10db step at 12.8GHz and I can measure Cal Output signal.
I suspect the Correction Factors are corrupted and at power on wrong information put the device in that situation.

With no service manual for 8592A I tried the procedures for the others HP85XX in order to enter the service menu. The Preset, Frequency, -37Hz (and -2001Hz), Cal do not work. The passcode is not recognized and after pressing the first More softkey there are not more steps.

Any idea about a way to solve the problem ?
Anyone with a Service Manual available ?
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf