The attached curve is the deviation from the current measured by the 3458A (which was quite stable). One can see, that the selfheating due to the 1A is already around 100ppm. Therefore, I think the Prema spec is very optimistic.
Well, looking at the datasheet of POWERTRON USR 4-3425
http://www.farnell.com/datasheets/2168435.pdfthey differ between
- TC: max. ±5ppm/K (-55 to 155°C), typ. ±3ppm/K (-55 to 125°C), typ. ±1ppm/K (25 to 60°C)
- load stability, which is given as 0.01%
- shelf live stability: 25ppm / ?R after 1 year, 50ppm / ?R after 3 years
So what you see is a mixture of load stability plus self-heating and thus TC. So the question is, what is really given in the manuel of Prema 6000 that you are comparing to?
Accuracy (1 year, 23°C ±5°C ± (%of reading (%rdg.) + % of full scale (%f.s.)) *)**)
±2mA and ±2A ... 0.01 (%rdg.) + 0.005 (%f.s.)
Temperature Coefficient
(10°C to 18°C and 28°C to 40°C)
2mA and 2A-range (up to 1A) ± (0.002%rdg. + 0.001 %f.s.)/°C
Temperature Coefficient
(0°C to 50°C) ... double the value at 10°C - 40°C
*) Add ± 1 Digit. After autozero.
**) Guard connected to "V/Ohm-LO" input socket, sinusodial signal greater then 5% of full scale.
So nothing about load stability, only TC and stability after 1 year. And please don't miss the extra digit mentioned in *).