I recently was lucky enough to acquire a Fluke 731B, which seems to have been one of the last ones made, close to 1987. I don't know the unit's cal history ( no stickers present ). The unit also has a new Ni-Cd battery installed, which reads the correct voltage
( about 16V ). However, when I did a quick basic test with a Fluke 287 ( recent calibration ), I found a potentially ominous issue. Although the voltage is initially 10.000 Vdc on that output, there is a 1mV positive spike about every 30 seconds. At first I thought this was normal for a unit that had been in storage a very long time, and the unit needed a long time to stabilize, or the unit was really, really out of cal. However, even after running it for several days, the 1mV positive spike still will not go away, nor has it become less frequent with time. Also, the unit has a negative -1mV spike about every hour. These problems exist both when the unit is run on both battery power and the unit is charging from being plugged in. No AC voltage is found on the 731B's output with that test. The spikes are DC voltage only. Also, when measuring the 731B's voltage, the analog bar graph on the 287 is constantly jittering
slightly upwards, it's never stable.
At first I though it was the switch issue, and I found all the gold plating was gone, so I tried measuring the 10.000V output on the Ref Amp board itself, at TP1 & TP2. The same 1mV spike problem exists, both here and on the front panel terminals. When I checked the 731B service section of the manual, I verified that Q1's collector voltage is correct. According to the manual, the only other potential areas of trouble are that either the LM308 ( U1 ) or- really shudder- the SZA263 ( U2 ) ref amp has become defective.
My only other theory is a defective R11- the 10V output adjustment control is either damaged, or has become noisy over time.
The other problem I have is that currently I don't have access to a 6.5 or 8.5 digit DMM to measure the 731B's voltage outputs more accurately. A Fluke 287 has a basic DC accuracy of only 250 ppm, so the 1mV digit can't be really relied upon. However, I did use the recording feature of the meter to do a basic voltage stability plot every 1sec for 2 hours. It clearly shows an almost constant peaking of 1mV positive, and confirms the once a hour -1mV drop.
Has anyone experienced this fault before, and if so, know what it is? Can it be repaired, or was my unit sold because it had developed a serious instability and isn't worth fixing?