The other methods can still use the formulas in NBS TN 430. Favored Cell reduces the number of measurements by only performing the measurements required for one (favored) cell. The Ring System allows a system with more than five cells to use the five cell formulas and methods. (The five cell system is efficient and produces the lowest uncertainties compared to the three and four cell systems, and is balanced, unlike the six cell system.)
In NBS TN 430, an intercomparison of three cells takes six measurements and any one cell gets four measurements. For four cells, there are twelve total measurements and any one cell gets tested six times. For a five cell system, there are ten total measurements and any one cell gets tested four times. I don't like the six cell system, because it is not (completely) balanced.
A favored cell system reduces the amount of measurements by only taking the measurements needed to calculate the pooled value of the favored cell (and the formulas in NBS TN 430 can still be used). For a five cell system, only four measurements are required (which are the measurements that the favored cell is used in). Then the pooled value and pooled uncertainty for the favored cell can be calculated. Tomorrow you could pick another favorite. The other cells could still be used to their (predicted) values and uncertainties. (There hasn't been enough measurements for the other units to have a pooled value. For one, for the five cell system, each one would have only been used on either the A or B line and only have one measurement, meaning the system was not balanced for them. Also, more measurements would reduce short term drift and noise.)
A ring system is for when there are more than five cells in a system, otherwise it is the same as the default system. If the cells were placed in a ring, any one cell is compared to the two on its left and two on its right. Then the five cell formulas could be used.
The most basic form of pooled uncertainty is if the uncertainties of the (predicted) values were all the same, then the (average) uncertainty could be divided by the square root of the number of cells involved to arrive at the pooled uncertainty.