Author Topic: new voltage reference from MAXIM  (Read 24791 times)
Online BU508ATopic starter
« on: June 11, 2016, 05:59:44 pm »
Hello,
I received today a newsletter from MAXIM. Seems, they have a new voltage reference:
https://www.maximintegrated.com/en/products/analog/voltage-references/MAX6126.html
From their website:
************************snipp***********************************
Key Features
    Ultra-Low 1.3µVP-P Noise (0.1Hz to 10Hz, 2.048V Output)
    Ultra-Low 3ppm/°C (max) Temperature Coefficient
    ±0.02% (max) Initial Accuracy
    Wide (VOUT + 200mV) to 12.6V Supply Voltage Range
    Low 200mV (max) Dropout Voltage
    380µA Quiescent Supply Current
    10mA Sink/Source-Current Capability
    Stable with CLOAD = 0.1µF to 10µF
    Low 20ppm/1000hr Long-Term Stability
    0.025? (max) Load Regulation
    20µV/V (max) Line Regulation
    Force and Sense Outputs for Remote Sensing
Andreas
Offline uncle_bob
Re: new voltage reference from MAXIM
« Reply #1 on: June 11, 2016, 07:39:38 pm »
Hi
Not bad, but not great either. There are now other chips on the market that will beat it noise and stability wise. It was a bit more interesting a while ago.
Bob
Offline uncle_bob
« Reply #2 on: June 11, 2016, 07:51:18 pm »
Here is a paper that is about a study that was done on a small collection of reference chips-- including the MAX6126.  It did fairly well in the tests, but of course as always, it would never beat an LM399 or an LTZ1000(A) for hysteresis, temperature, humidity, or time drift.
ANY unheated reference IC will experience hysteresis with large temperature swings.  The hermetic packages do better with this, but it is not totally eliminated.  The LM399 has almost no hysteresis, and the LTZ1000A has a little more than the LM399-- but both of these are orders of magnitude better than ANY unheated chip [including the LTFLU-1].
This is OK, if the reference fills your need and the various drifts [when added together] are within the specs that you are looking for.
Hi
Paper ? (link appears to be missing :) )
I agree with everything you have said. My only point is the missing link.
Bob
« Last Edit: June 11, 2016, 07:54:34 pm by uncle_bob »
Offline necessaryevil
« Reply #3 on: June 12, 2016, 03:02:50 pm »
Yeah, the MAX6126 is a nice one. Forum member Blackdog  made a voltage reference using four of them in parallel.  Search the forum! You should also check out the LT1021.
Do heated chips really not suffer from hysteresis? If they do, it will be worse than non-heated chips, because the heating will cause them to go through more temperature cycling at turn-on/switch-off.
Offline uncle_bob
« Reply #4 on: June 12, 2016, 03:23:28 pm »
Hi
If the heated chip is powered up during the temperature cycle, the die sees much less "delta T" during the cycle. The chip does indeed change the amount of power into the heater. That change *does* have an impact. It is much less than the impact of a cycle on an un-heated part. Indeed the ideal case would be a heated part inside a heated enclosure. That has it's problems as well ....
Bob
Offline Andreas
« Reply #5 on: June 13, 2016, 06:45:42 pm »
Hello,
if you really want a "new" (old) voltage reference you could try the brand new LT1027 in hermetically LS8 package.
http://www.linear.com/product/LT1027LS8
with best regards
Andreas
The following users thanked this post: EmmanuelFaure, Squantor
Offline branadic
  • Sounds like noise
« Reply #6 on: June 13, 2016, 08:05:35 pm »
Wow, nice to see that they heard your prayer and  LT1027 is now available in LS8 package. I remember a mail conversion from 2012 with you Andreas, when you mentioned such a wish. So I guess you plan to do some comparision between LT1027xCH-5 and the new LT1027LS8?
Parameters are somewhat comparable to LT1027DCH-5, but at least a little bit different. I'm very interested in such results, as hermetic package version was not available for long time. How will they compare to LT1236LS8 and LTC6655LS8?
New input for you, I'm sure :)
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
Offline Andreas
« Reply #7 on: June 13, 2016, 09:13:26 pm »
Hello,
unfortunately I do not have any old TO-5 hermetically LT1027.
And I guess that those you can buy nowadays are only fakes.
From my comparison measurements of DIP8 LT1027 against DIP8 LT1236/MSOP LTC6655
The LT1027 has the lowest hysteresis (typically < 1ppm over 25+/-15 deg C).
So if they did not change the chip layout I expect excellent stability for the LS8 package.
Perhaps even slightly better than my up to now favourate AD586LQ with the advantage that
the LT1027 can be operated at around 10V whereas the AD586 needs minimum 14V for proper stability.
But against the old TO-5 package there will be no mechanical decoupling from the PCB.
So the new parts are worse if not properly decoupled from PCB stress.
Does anyone have a chip foto / bondout layout from a old datasheet of the LT1027?
Or a defective LT1027**H for a photo. (I am shure branadic would do a photo if he gets one of those chips).
With best regards
Andreas
Online lowimpedance
  • Watts in an ohm?
« Reply #8 on: June 13, 2016, 11:50:23 pm »
Had a look in a 1990 Prelim data and 1992 data book and there is no such detail unfortunately.
And seems all other components in the 'vintage paper' data books also do not include pictures of the die/bondouts.
Offline ebclr
« Reply #9 on: June 14, 2016, 01:15:50 am »
What about LT6657
« Reply #10 on: June 15, 2016, 08:38:35 pm »
So, this means that hysteresis is more or less caused by temperature difference in the case and the die? Or, more accurate, caused by difference in expansion/shrinkage of the die and the case?
Offline Andreas
« Reply #11 on: June 15, 2016, 09:38:32 pm »
I blame it on the die attach (usually some kind of (filled) epoxy) which has
a elastic component which "remembers" tensions and stresses the die.
And of course the package if it is a plastic (epoxy) package.
But I also think that with a clever chip layout these effects could be minimized.
k=topic=69565.msg961475#msg961475 date=1465866950]
What about LT6657
[/quote]
No experiences with that part. (not in my focus because of plastic package).
With best regards
Andreas
Online Alex Nikitin
  • Femtoampnut and Tapehead.
    • A.N.T. Audio
« Reply #12 on: June 16, 2016, 07:51:39 am »
Any experience with this LT1021CMH supply?
http://www.ebay.co.uk/itm/191123710347?
I've bought a couple and plan to test these but right now I have very little "free" time to do it. It looks good however and the price is OK for a NOS metal case C grade reference.
Cheers
Alex
Online tszaboo
« Reply #13 on: June 16, 2016, 08:36:48 am »
Hello,
with best regards
Andreas
My only problem with the LT1027 is the tempco. I mean we get DACs with built in 2ppm/K references, the 5ppm/K of the LT1027 is not that impressive. Sure you get less moisture and aging, but the tempco ruins it.
I rather have a good tempco reference like the LTC6665.
Offline d-smes
« Reply #14 on: June 16, 2016, 10:48:50 am »
I think it's the drastic up-curve above 100C that gives the LT1027LS8 the 5ppm/K tempco.  See "Output Voltage Temperature Drift" graph on page 4 of LT1027LS8 data sheet.  Stay within 10-20C of room temperature, and typical tempco curve is fairly flat.
Online tszaboo
« Reply #15 on: June 16, 2016, 10:58:20 am »
Since they only give you the "box method" for some huge, usually -25 to 125 temperature range, the guaranteed drift is going to be more than the actual for a 20 to 30 celsius range. But that is the guaranteed spec. Unless there is a better number, or you can test enough units to build a confidence in it, I go with the datasheet numbers.
Online Alex Nikitin
    • A.N.T. Audio
« Reply #16 on: June 16, 2016, 11:14:18 am »
The tempco is measurable and if required can be compensated for. The ageing, humidity sensitivity and hysteresis are not predictable and not possible to compensate. What the point of a low tempco if your reference may drift considerably on it's own even at the same temperature?
Cheers
Alex
The following users thanked this post: ebclr
Online tszaboo
« Reply #17 on: June 16, 2016, 11:43:24 am »
Cheers
Alex
Have you ever tried compensating tempco? Please just describe how much effort it is to do that.
Online Alex Nikitin
    • A.N.T. Audio
« Reply #18 on: June 16, 2016, 12:16:05 pm »
The easiest way is (as you probably aware) is a temperature stabilization. In a small range (say +/-10C) a parametric compensation is relatively simple. If you need the reference to work across the complete range of temperatures, than errors the tempco can introduce are comparable to the drift and hysteresis variation so a low tempco would be a good thing. But in that case we are talking about ~100 ppm errors.
Cheers
Alex
Offline branadic
  • Sounds like noise
« Reply #19 on: June 16, 2016, 05:05:54 pm »
Well, Joe Geller did on his SVR-T:
"The -T option adds a 100 kilo ohm thermistor in series with a fixed temperature compensation gain resistor for a secondary temperature compensation scheme (developed by Lars Walenius, Sweden)."
Offline acbern
« Reply #20 on: June 16, 2016, 06:22:48 pm »
The problem with this is that it is very hard to predict behaviour too. How strong is thermal coupling between thermistor and referenc, what exactly is the compensation required, each reference requires manual adjustment as coefficients are varying...
So, does not replace a tempeature stabilized (oven) solution.
Offline Andreas
« Reply #21 on: June 16, 2016, 09:14:43 pm »
I do this for my 24 Bit battery powered ADCs.
I am using a 3rd order regression curve to compensate for the tempco of the voltage reference.
(The LTC2400 itself has a negligible TC).
The advantage against heating is a very low power consumption. (< 50mW against > 200 mW)
Of course you cannot use such a system together with large environment temperature gradients.
(But any precision instrument is not good at that).
Effort:
On the hardware side: only a cheap NTC + 1% pull up resistor on a 10 Bit (or better 12 Bit) ADC input.
For the adjustment: A automated temperature cycle for measuring the voltage reference (around 10-14 hrs machine time) to calculate a 3rd order regression curve.
The coefficients are stored within a controller which I need anyway to read out the ADC.
With best regards
Andreas
Offline uncle_bob
« Reply #22 on: June 16, 2016, 09:29:47 pm »
Hi
Some fairly basic math:
You have a tempco of (say) 20 ppm.
It is measurable / repeatable / predictable to some percentage of that. Let's say 5%
Your math is perfect (it's a formula in a MCU).
Your temperature range is 100C (say -30 to +70)
So:
If you know the temperature to a repeatable 1 C, it is 5X better than you would need for a straight line correction. For a parabolic, you might want 2 or 3C. If you have 1C, it pretty much drops out.
Result would be a part that is compensated to ~1 ppm over a 100 C range.
For under a dollar, you can get a *bunch* of different solid state temp sensors. They have different accuracy ratings, but in general are all very repeatable and stable long term. Resolutions in the 0.06C range are not uncommon. Averaging will get you a bit better than that. Repeatability to 0.1 C is achievable with some care (don't mount the sensor next to the power regulator ...). They are highly linear and hold resolution over the entire range. A NTC plus resistor has great resolution at some central temperature. Not so much at the extremes.
We dumped NTC's almost a decade ago and switched to the solid state gizmos. From what I have seen, there is no reason at all to go back to the thermistors. Our math is about 10X tighter than what I've shown above. We do indeed average the output a bit.
Bob
Offline Andreas
« Reply #23 on: June 16, 2016, 09:36:35 pm »
Hello,
Just some measurement values of 2 samples of LTC6655B
(Typical TC 1 ppm/K max TC 2 ppm/K according to datasheet measurement method = 3 point box method)
ADC14 is built with a plastic MSOP8 package on a slotted PCB.
ADC20 is built with a hermetic LS8 package mounted dead bug style in a PCB cutout to avoid PCB stress.
in both cases around 180uV (=52ppm) measured output voltage change over 25-30 deg C temperature change giving around 6-7uV/deg C
corresponding nearly 2ppm/K with respect to the input voltage of 3430mV.
The plastic package shows +/-20 = 40uV or 12ppm hysteresis.
The LS8 hermetically package shows +/-14 = 28uV or 8 ppm hysteresis.
Red line: ADC output value in mV at constant input (LM399 reference divided by 2) over temperature.
Green line: 3rd order regression curve.
Blue line: resulting error in uV (mainly hysteresis of voltage reference with LTC6655)
With best regards
Andreas
Offline uncle_bob
« Reply #24 on: June 16, 2016, 09:46:22 pm »
Hello,
With best regards
Andreas
Hi
... and the next layer to the problem is that the hysteresis "goose egg" is a function both of rate of change of temperature and of the temperature range. Simply put: You can only just do so well and then it's not worth it to take all the data. No, this does not just apply to compensating voltage references. I can show you an *identical* plot (other than units) on ....errr .... something else. Of course if I did, I'd be fired, it's my employers data.
Bob
Offline Andreas
« Reply #25 on: June 16, 2016, 09:57:07 pm »
Hello,
similar game with some LT1027CCN8-5 (DIP plastic package) specced 3 ppm/K with datasheet method.
Temperature variation of samples #4 to #7 ranges from 60uV to 250uV over a 35 deg C temperature range.
Giving 0.5-2.1 ppm/K average T.C.
Hysteresis is much lower than with LTC6655 (usually below +/- 1ppm).
With some "cold creeping" effects (I blame it to the plastic package).
ADC8 shows the regression curve of sample #3.
The resulting error is around +/- 2uV or +/-0.5 ppm.
So I am curious how the LS8 package will perform.
With best regards
Andreas
Edit: of course the temperature compensation will not help against the humidity sensitivity of the plastic DIP package which is around 0.5ppm/% rH for the LT1027C. This will give around 15 ppm seasonal change (40-70%rH) here in Germany.
« Last Edit: June 16, 2016, 10:06:50 pm by Andreas »
Offline uncle_bob
« Reply #26 on: June 16, 2016, 10:02:55 pm »
Hello,
With best regards
Andreas
Hi
Numbers 5 and 6 may be exhibiting a bit of dependance on the low end temperature. it also could be a number of other things.
Bob
Online Kleinstein
« Reply #27 on: June 17, 2016, 07:21:26 am »
The hysteresis in temperature cycles can also include some humidity effect: heating lowers relative humidity levels and thus drives out water from the plastic of the case and the board. It can take quite long (weeks and more) for humidity to come back at room temperature.
The temperature compensation / numerical correction only helps a little with low TC parts, but it can be very helpful with parts with a higher TC. So with compensation the main parameters to look at are long term drift and hysteresis - the TC is not that critical any more. If measured in the real circuit and with the typical slow varying temperature, the thermal coupling is not critical and also linearity of the temperature sensor is no critical, it's mainly repeatability and long term stability that is important. So a temperature sensor in a plastic case (like many of the digital chips) is not such a good idea,as they can show similar hysteresis.
Online tszaboo
« Reply #28 on: June 17, 2016, 08:51:30 am »
Someone finally got it right!
You can characterize a reference with a heat chamber and it is possible to compensate it's change with a different part. But the characterization is very expensive, time consuming, labor intensive and so on. It is not really an option for a production environment, while it might be an option for a one-off reference.
I did calibration of reference voltages. With a 3458A. It was a major operation, with extra hardware designed to burn in stabilize, and measure the references multiple times to get the confidence level. It was a week to do some few hundred boards calibrated. 30 minutes warm up, 1000 NPLC, several measurement, and paralleling the work. It was just barely possible to do this, for references, which deserved it (the one with the 1000 in its name).Now, include a temperature chamber test, selecting the right compensation component soldering, burn-in again, verification... It is just not possible. And not worth it, because once you need to spend labor to do it, mind as well use a heated reference.
Offline uncle_bob
« Reply #29 on: June 17, 2016, 11:19:27 am »
Hi
Doing 100% temperature test and characterization may not be production compatible in all areas, but it is in some. We run a *lot* of parts through temp comp every month. Some of the stuff sells for < $20 in volume.
Bob
Offline uncle_bob
« Reply #30 on: June 17, 2016, 05:57:54 pm »
I think you could automate this with the right setup.  Perhaps running up to 100 devices per 24-hour run in an environmental chamber.  Building the automated DAS and acquiring the necessary instrumentation and low thermal EMF switches might be rather expensive, so you would only do this if you are running many production runs.  As far as I can tell, there just is not much of a market for this kind of thing.  After [and *IF*] you sell your first hundred units, you will be hard pressed to sell another 100 I think.  To make any money from this, there has to be a "Pot O' Gold at the end of the rainbow"...  It could be that the massive investment in money, equipment and time will end up being a huge loss.  UNLESS you are a charlatan, and are just installing an epoxy packaged reference IC inside a plastic box, and then selling it to unsuspecting dupes on eBay...
If you are just doing this as a hobby, well then the amount of money you spend on the reference and the amount of "labor" you put into the project is irrelevant, right?
There are some very good references in epoxy packages that are good enough for 4.5 digits of accuracy [100ppm], and there are some monolithic ICs in hermetic packages that might be good enough for 5.5 digits of accuracy [10ppm], but if you want or need accuracy >= 6.5 digits [1ppm], then you are going to have to go with a heated reference, and at this level of accuracy, it's hard to beat the LM399 for lower cost projects or the higher performance LTZ1000(A) for projects where you can afford the best.
All of this talk about getting super high performance [1ppm] out of an unheated monolithic reference IC is becoming tiresome, because it has been explained over and over in multiple threads why this will never happen.  Any effort to this will just be a "fool's errand"-- so, don't be a fool.  If it sounds too good to be true, then there is almost a 100% certainty that it is false.  This fact of life also applies to voltage references, and also to charlatans on eBay claiming to provide you with 2ppm of accuracy for $99...
Hi
The thing you need to do is to scale your test setup to the market demand. If you figure that the total available market is a hundred a year, a single channel system is probably plenty good enough. Maybe two single channel setups if you worry about this sort of thing a lot. The cost of running the chamber might change that a little. You don't need a very big test volume for a single piece, so maybe not.
Bob
Online splin
« Reply #31 on: June 19, 2016, 05:54:04 pm »
Bob
It would be interesting if you could quanitfy long term stability as it is rarely specified for low cost sensors (and often for expensive devices too) and if it is the spread is usually large - eg. LMT70 specifies drift as < 10mK typical, +/= 100mK max, 10k hours @ 90C. Presumably you used calibrated sensors to monitor their stability?
Offline uncle_bob
« Reply #32 on: June 19, 2016, 06:47:57 pm »
Bob
Hi
We plot them against stuff that should show us 10s of mK over the course of months. That testing is done on the entire system rather than an individual component. Unless the sensor *always* happens to drift in the same direction as the rest of the system .... we would see it.
Bob
Online splin
« Reply #33 on: June 19, 2016, 08:10:45 pm »
Hmm, I'm not quite sure what you mean - you're comparing against a number of different sensors/types? If you aren't taking any absolute, calibrated measurments surely you can only conclude that some (perhaps most) drift at similar rates and any that don't stand out beause they are different - but those could be the ones that aren't drifting significantly?
Isn't this the problem of having lots of clocks or voltage references and never being sure which are right?
Offline uncle_bob
« Reply #34 on: June 19, 2016, 09:02:39 pm »
Hi
Not really:
You have a voltage reference, it is out in the ambient environment
You have a voltage controlled gizmo that you are compensating, it has a known sensitivity to the control voltage
You have a temperature sensor that gives you the data for the compensation process
You compensate the whole thing to a few 10's of ppt using the temperature data from the sensor
You validate the compensation process and put the beast on long term burn in
You come back in 3 months / 6 months / 12 months and see how everything is holding up
If the compensation holds to a few ppt or so of what it was, you can back into the stability of the parts.
The compensation of the device is a temp sensor driven process so the end result is the sum of all the errors along the way. You can put upper bounds on things, but not lower bounds. We have a few points in the system we can monitor and take out a couple things, it's still an "upper bound only" estimate.
Yes, there are a bunch of grubby details about sensitivity and how tightly you are compensating things vs how bad they were at the start. Since it's my employers "secret" :) process,I can't get into quite everything. I suspect I've put enough in there that you can pretty well guess what we are playing with.
Bob
Online Kleinstein
« Reply #35 on: June 20, 2016, 05:04:19 pm »
Having a heated reference does more than just reduce the TC. It also reduces hysteresis quite a lot, as the thermal history will be similar after each turn on. Also the relative humidity of the heated part will be low - so the humidity effect is also reduced. However there are two down sides: the part gets hotter and thus ages more, and the power consumption is higher, usually by something like a factor of 3 at least.
Offline lars
« Reply #36 on: June 27, 2016, 05:53:42 pm »
Cheers
Alex
I bought two in March last year.
In December I put them in boxes with NTC temperature compensation ala SVR-T. I did the same with 2pcs of REF102CM and 6pcs AD587 in ceramic packages I had in my junk box. So totally ten boxes. The boxes are 35x55x20mm plastic and the binding posts are gold plated brass from China. The 15v power is a 5.5/2.1mm connector and I also have two small 4mm banana jacks connected to a 10k NTC. The PCB is just a prototype board. So a lot of not optimal solutions.
For the LT1021CMH I used two 470k NTC (SMD0805) in parallell as I had that in my junk box. I added a resistor about the same value in series to linearize the NTC as I didn´t bother to compensate the second order component. To trim the temperature coefficient I used a trimpot from the  junk box between the output and ground. Used another trimpot and resistor to trim the voltage.
The LT1021CMH´s had +5 resp. +7ppm/C before compensation (well within spec of 20ppm/C). They were about 200ppm low.
Enclose a chart for the first 6 months of test with continuous power. The refs probably was powered down a couple of weeks in end of March as I had an intermittent connection on the +15V supply. On April, 1 I had them powered 4 hours before measuring. All values are relative to a SVR-T board. As null meter I have used a homemade x100 amplifier + 3/2 digit DMM to get 0.1ppm resolution.
The result for the LT1021 are better than I hoped. 5 years ago I bought 6pcs LT1031DMH from Distrilec/ELFA and was very disappointed. See attached graphs. During two periods I have turned of power between the measurements and only turned on it for 1hour for the 2pcs in room temperature and one period for the 4pcs in a box about 13C above room temperature. As can be seen it is a terrible difference with power on continuous instead of off for several weeks between power on. I will do this test with the LT1021 later. as it seems so different from the LT1031.
Lars
The following users thanked this post: BravoV, Andreas, EmmanuelFaure, Alex Nikitin
Offline lars
« Reply #37 on: June 27, 2016, 07:38:54 pm »
Some comments about temperature compensation or stabilization (not inside the IC) of voltage references:
In my opinion it takes time both to design and test the design for both approaches. Design is probably easier with an NTC compensation ala SVR-T. For professional work I prefer to go with stabilization as production test time is expensive. Trimming a compensated hobbyist reference takes time but not necessarily advanced equipment. More than 15 years ago I trimmed the compensation on my first REF102CM´s with NTC´s by just warming them with a lamp (at that time it wasn´t a LED lamp :). Three out of four has about 0.0ppm/C at 23C and the fourth I must have made a mistake as it is 0.4ppm/C :(
For compensation I have used NTC´s for designs with compensations if no uC is involved. NTC´s are non-linear and that is perfect for both first and second order compensation over a narrow temperature range (as in the SVR-T). In my professional work I very seldom design in NTC´s but instead silicon sensors. I really have no idea how the silicon sensors drift but I guess well below 0.1C per year. Assume you want to compensate a volt ref with 5ppm/C. 0.1C per year drift will give 0.5ppm/year. Not easy for most of us to measure.
Stabilization might not be better than compensation as you still need to consider placement of the sensor and temperature gradients and how they change with surrounding temperatures. A more complex design might also have internal changing gradients due to varying power.
I also would like to say that my hobbyist goals are quite different from my professional. My goal as a hobbyist has always been simple and cheap designs. In my professional work I get enough of complicated and expensive equipment (most >1MUSD each). Being a volt nut for just 20+ years but a curious hobbyist for at least the double time I would really encourage to experiment with voltage references if you like measurements and calculations as I do.
Another comment: As a professional I almost 100% go by the data sheet spec but as a hobbyist it is nice to characterize the voltage ref IC´s myself and see if I can get better performance.
Lars
Offline Gyro
« Reply #38 on: June 27, 2016, 08:04:02 pm »
Hi Lars,
Very interesting data, thank you. Can I ask, were your AD587 samples already aged? They seem very stable if not.
It looks as if I can have good confidence in the long term drift of my SVR-T.  :)
Chris
« Last Edit: June 27, 2016, 08:07:24 pm by Gyro »
Best Regards, Chris
Offline lars
« Reply #39 on: June 27, 2016, 10:05:10 pm »
Yes, the AD587 and REF102 was salvaged from old boards.
But still I think you can have very good confidence in the SVR-T. I have four SVR-T and four SVR boards converted to SVR-T. They are of different ages but still within a couple of ppm from each other and I have compared them to one of my REF102CM´s that has been powered for more than 15 years and the drift is less than 2ppm compared over about 5 years for my oldest SVR.
I also have two AD587LQ datecode 0045 that I got from Joe five years ago. They have been powered and in a box with about 13C above room temperature (30-43C). One has drifted +3ppm rel the REF102CM and the other -3ppm. During three months 2014 I turned off the box they are in and only powered up them for three hours about every third week. During that time I measured after 1 and 3hours that can be seen in the attach graph. Still the graph is almost a straight line. Compare to the LT1031´s! in the same box.
These two AD587 are not temperature compensated but instead four LM35´s are in the box and data post processed. They have a tempco of about +2ppm/C
My best guess for the SVR-T is about 5ppm expanded uncertainty as a transfer std (3 weeks) and 10ppm over 5 years.
Lars
Offline Gyro
« Reply #40 on: June 28, 2016, 08:54:29 am »
Many thanks Lars, more data than I could possibly have hoped for.  :-+
I think I remember Joe only having acquired one big batch of AD587LQs so they are probably all 'immediate family', mine is also 0045 (batch C67337).
That's quite a reference farm you're running there!
Chris
P.S. If those REF102s were also salvaged then it potentially also says something about their relative response to physical disturbance.... or maybe even CERDIP vs TO-99?
« Last Edit: June 28, 2016, 09:17:59 am by Gyro »
Best Regards, Chris
Offline HighVoltage
« Reply #41 on: June 28, 2016, 08:33:09 pm »
I just came across this one:
Analog Devices voltage references P/N: AD2710LN
Ebay listing for US $218.00
http://www.ebay.com/itm/201420682682
What is so special, that it is so expensive?
Does anyone here had experience with this reference.
Offline lars
« Reply #42 on: June 28, 2016, 08:40:18 pm »
Hi Lars :)
Anything special about this amplifier? In this particular application, the offset and offset drift are a bigger issue than gain error and gain drift. Anything special about offset error compensation?
No it isn´t anything special with this amplifier. Had to check my schematic from 1992.
I use a TLC2654 zero drift op-amp and 10k+1Mohm 1% metalfilm. As you say gain error and drift isn´t a big problem, but I may have selected the resistors.
I have no offset compensation. A note from 1992 says -4uV offset and now it is -7uV. As voltage references have low impedance outputs I read the amplifier shorted before measurements.
Lars
Online tszaboo
« Reply #43 on: June 29, 2016, 08:30:22 am »
So last week TI introduced the REF60xx references. They have a high speed buffer built in, otherwise its like the REF50xx family. It is actually really nice. I always had to buffer the REF5025, which was going to those high speed 18 bit SAR ADCs.
BTW if you havent, then get one of linear technology's 20 bit SAR ADCs to play with. Its can measure one PPM, and do it 100.000+ times a second. Truly amazing.
The following users thanked this post: evb149
Offline SoundTech-LG
« Reply #44 on: August 16, 2016, 01:39:21 pm »
REF60xx references...
Nice, I had just ordered, and received the Maxim, but these appear to add some value with the buffer.
Online tszaboo
« Reply #45 on: August 18, 2016, 08:46:25 am »
No, this 24 bit part has some digital filtering in it. I'm talking about the LTC2376-20 - 20-Bit, 250ksps, Low Power SAR ADC with 0.5ppm INL part. I had the chance to send it againtst a Agilent 3458A, and auto calibrate it with a buried zener reference voltage. That was something truely amazing analog frontend.
Offline uncle_bob
« Reply #46 on: September 01, 2016, 03:29:37 pm »
Hi
I believe you will find that a lot of the 24 bit ADC's have a nasty 1/F noise corner on them. That makes the whole "average to take out the noise" approach a very long and slow process. If you look at the data on the LTC2376-20 there certainly is a low end roll up on all the plots they show. The solution is often to go to a chopper based front end. When they do that, it is up in bold print at the top of the data sheet. They charge you a bit more for it ....
Bob
Online Kleinstein
« Reply #47 on: September 01, 2016, 04:21:06 pm »
These LTC237... ADCs use a sampling input stage, so I would not expect much 1/f noise. The input is much like an AZ OP: the very low zero drift (7 ppb/K  which would be something like 30-70 nV/K) indicates this. The main source of 1/f noise if likely the reference voltage source.
The relatively high speed could be an advantage for a chopper type input stage, as there is not much time lost - it could be more like the auto zero mode of a DMM, doing the demodulation in digital and not analog. I don't think this would be needed because of the ADC's LF noise, but more due to noise / drift of the input amplifier itself. For a single reading the noise might not look that low, but it already a challenge finding a suitable reference, especially if you need the LF range too. After averaging down to a lower sampling rate, this gets very competitive compared to DMMs - they too may be limited by the reference.
Offline uncle_bob
« Reply #48 on: September 01, 2016, 04:25:27 pm »
These LTC237... ADCs use a sampling input stage, so I would not expect much 1/f noise. .....
Hi
That's what I though before I actually had to make it all work. Turns out that there are still very real noise sources in the signal path.  People like TI make parts with choppers in front of them.  Yes, I did go back and yack at Linear about it. They ultimately admitted that there is a 1/F corner on the parts.
Bob
Online tszaboo
« Reply #49 on: September 02, 2016, 09:53:20 am »
Hi
Bob
Just look at the "DC histogram" figure. So that is the amount of noise you will get. Getting rid of the pink noise is real pain. Ultimately you need to background calibrate the signal chain (zero measurement on any DMM) and use averaging/longer sampling. ie a 8.5 digit DMM will only give you true 8.5 digit result, if you use very long NPLC settings (1000NPLC on a 3458A). So turn up those integration times. Also mains frequency synchronization might be really necessary.
And yes, choppers help, and they dont help. They increase the wideband noise above the chopping frequency. So you need to filter it. For example with an RC, increasing the impedance driving the ADC. So you buffer it... with a chopper?
Online Kleinstein
« Reply #50 on: September 02, 2016, 01:52:50 pm »
For very high resolution you will need some kind of chopping or auto-zero for the input stage anyway. A normal high impedance input stage will have way more 1/f noise than the ADC. As these ADCs are really fast, they are more suitable to use just a chopping stage at the input and do the demodulation in software. So no classical chopper OP, but more like the auto zero of DMMs. Measuring zero and the signal one after the other (or changing polarity). This way 1/f noise of the ADC is not that relevant anyway.
Even with chopper OP there will be some extra low frequency noise from thermal effects - not classical 1/f noise, but similar. The other part is the 1/f noise of the reference - at a certain level even a LTZ1000 will add to the noise.
With such a fast ADC, there is not that much analog filtering needed before the ADC and thus no high impedance needed.
Online tszaboo
« Reply #51 on: September 02, 2016, 03:06:18 pm »
Yes, that is one way of doing it. Chopper opamp on the input, analog switches filter after it, and a buffer for the ADC. The buffer is a traditional opamp, the analog switch (2x 1:3, signal, ground, vref ) is switched with a few KHz. After that you need to iron out settling time issues for the buffer, timing and digital magic. I had an analog fronend operating with this principle, the end result was excellent. It was compared to the 3458A at work, they were tracking each other very closely for weeks. I think the result would be a lot better, if it would be a purpose designed data acquisition card. Actually I thought it several time to make my own, probably after my NDA with my ex company expires. The only issue is that the BOM cost is so high, that even a functional prototype is expensive, and I dont have access to a volt nutter multimeter anymore.
I've been thinking about making it a kickstarter, but I actually dont think there are too much volt nutter out there that would buy it. And the ones that do, know (at least parts of it) lot better, so they would rather dispute than buy.
Offline David Hess
  • DavidH
« Reply #52 on: September 05, 2016, 04:52:17 am »
I have done this sort of thing before by using a chopper stabilized operational amplifier in parallel with a low noise operational amplifier through the offset null capability.  The time constant of the chopper sets the breakpoint between the low 1/f noise of the chopper and the low wideband noise of the conventional operational amplifier.  Calibration is actually pretty easy; configure a x1000 amplifier, measure the output with a DC voltmeter, calculate the standard deviation to get the low frequency RMS noise, and trim the time constant for minimum noise.  The last time I did this, the breakpoint frequency was very close to where the separate noise curves met.
The one thing I have not tried is synchronizing the integration period with the actual power line cycle which may not be exactly 50 or 60 Hz.