Just some general food for thought, since RHB seems to be just starting out:
If you run LED lighting, even if you run RAW (directly from mains) be aware there is a small current transient when the diodes switch "on", so pay attention that the power source is quiet DC (and no switch mode power supply allowed near low-ppm test area - batteries or regulated linear + isolation XFMR only). Also pay attention to the LED CRI - If it's not at least 90+ it becomes more difficult to see the difference between some shades of wire or component code colors - say orange and red.
For those reasons we generally only use halogen incandescent from 12V, 24V or 48V battery banks during critical tests.
As far as cabling goes for chasing low ppm - you're generally not ever going to be measuring a Vref with a 'scope probe (not without an LNA). Cat5 or even shielded Cat6-7 with faster twist rates is fine in most cases, and a decent DMM is required. CLEAN Copper to Copper connections only, watch out for oxide Copper or tarnished silver plate on that silver-plated wire (that creates thermal EMFs and noise). Gold plated connectors are easier to keep clean, but add a small amount of noise. Keep the leads dressed SHORT and use DIRECT wiring, keep power cords away from sensitive circuits, make use of shields / guard circuits as required, and understand ground loops.
Remember: Every circuit loop that encloses a non-zero area is a potential magnetic pickup point, and every PN junction is a potential demodulator.
Copper / Aluminum shielding is not the best at low freq - for that application look at steel and or MuMetal. If you're really serious about cable shielding at low freq you start to look at MuMetal braided cable sheathing - depending on your work environment. In our work environment we have to use that since the 3458a's are being used near some servo drives, and we're trying to grab a low ppm DC measure. In that case the servo wiring is on one shield system, and the measuring cables are on a separate shield / guard system.
More Suggested reading - Keithley's Low Noise Measurement handbook, and Art of Electronics, 3rd edition.