I've mounted a large heat sink to the base of the standard, and I"m letting it warm up to see how hot it gets. so far, about 10 minutes, it is warm, but no longer uncomfortable to touch. I don't have an easy way to measure the temp, maybe I"ll try my meat fork. I did, and after about half an hour it reads 43 C
Meanwhile, I've been monitoring the frequency with an HP 5335A counter, and the counter and the standard either drift at the same rate, or both are relatively stable. The reading is the same as it was the last time I powered the standard up a week ago, 9.999 999 28, and doesn't seem to be moving at all.
My question is, since the OCXO in the counter seems to have good short term stability, and perhaps long term as well, would it make sense to adjust it to 10.000 000 00?
If so, I looked at the documentation I have and didn't see a procedure for doing this, has anyone done it? If so, how?
I do recognize that this is a very small change, and adjusting it may be a bit silly.
Thanks all for the info and encouragement.
Brad
Well done, the cooling, and also the measurement of the temperature!
The adjustment of the OCXOs frequency might make sense, as the Rb clock might be closer to 10MHz, as it drifts less.
Myself, having several HP OCXOs (i.e. 10811A-60111 in the 5335A and the 5370B), these have typical stability of around 10^-9, over many months, after all these years.
You can trim these OCXO to a bit better than that level, i.e. a few parts in 10^-10, provided, you always leave the 5335A plugged in, so that the single red LED is illuminated.
Your 7.2*10^-8 deviation is about 2 orders of magnitude bigger than what these OCXOs are capable of, so it makes sense to trim. (The "but" comes later.)
It's easy to adjust the OCXO to the same frequency as the Rb standard.
Simply apply the 10MHz from the Rb to the A - input of the 5335A, choose longest possible Gate Time, about 4 sec, and measure the frequency.
Open the lid of the counter, and adjust the OCXOs frequency, so that the counter reads exactly 10.000 000 00 MHz.
The 5335A resolves about 1..2ns / sec Gate Time, so it has at least 10^-9 resolution @ 4 sec Gate Time.
I made a modification on my counter to have 12.5 sec Gate Time on the front potentiometer, so my 5335A displays 10 digits.
Anyhow, if you have GPIB capability, you may also use Time Lab, a freeware program by ke5fx:
http://www.ke5fx.com/timelab/readme.htmIt allows you, by means of the 5335A itself, to measure the Allan Deviation of your OCXO vs. the Rb, and also to make precise adjustment of the OCXO in the frequency deviation measurement mode.
Here comes the "but": The basic problem, whether your Rb standard is still precise enough, you can solve only by taking the next step, that is acquiring a GPSDO receiver, and adjusting the Rb standard to < 10^-11.
May the Lady Heather be with you!
Frank