It´s hard for me to understand how you will do that.
I think this is a polite way for you tell me that I need to take more attention to this problem I may overlooked
Connecting the negative sides can be more tricky ... or alternatively averaging resistors as well.
OK let see if I understand you both guys.
PIN 7 of LTZ1000 will be a little bit over the GND and each of the LTZ1000 GND point will be a little different
than the others. If I connect this four pins together I get strange things to happens due to some current flowing
in other paths than "expected" (ground loops).
This means that I need to use an op-amp to cancel out this "offset" and lower everything to a common GND
point (this point than can be used as the LO side of the reference). This 4 current cancellation op-amp must have a
negative rail to push everything down.
Is this correct?
This meas also that special care must be taken to have also the 120 resistor GND side to almost on the same GND level.
I am also not convinced that a PWM-ratio is easier than a resistor ratio. It is only different.
Never told is easy (especially for me). What I was thinking is that the thermal regulation should be a little bit
forgiving so if you can only get it stable to 10ppm (I think) it will be perfectly acceptable. How much is hard to
get it I still have to investigate. Thermal lag may also fix a non perfect pwm filtering if it don't screw the
thermal regulation circuit.
The output of the normal LTZ1000 circuit is already buffered.
Yes but I want to isolate LTZ1000 from capacitative load problems (had a lot of problems in past that I want
to avoid). Than I'd like also to have independent channels for sporadic itercomparison. Don't know if this can
be done with simple resistive averaging.
For the temperature set-point I would be careful with a PWM generated voltage - any failure here could lead to over temperature and there would also be possible noise coming in this way. It might be enough to have the option to measure the actual divider ratio (that is have an output) and allow for some initial calibration measurements with a slightly shifted temperature.
That's nice. Maybe the best is still to use the 13K/1K divider and than add a fine adjustment by PWM.
Measure divider ratio time by time (any 6.5 digits multimeter should be sufficient for this) and digital
compensate for its drift.
Considering that we need just few hundred ppm margin to compensate for divider TC and log term drift
this should be easier to do than a 100% PWM signal. This will also ensure that you will not cook the
reference in case PWM circuit fails (PWM only contribute to a very limited amount on the temperature
set poit). Maybe also a 10/12 bit DAC should be enough stable for the same adjustment purpose.