Author Topic: Ultra Precision Reference LTZ1000  (Read 809604 times)
Online TiN
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: Ultra Precision Reference LTZ1000
« Reply #2450 on: November 16, 2018, 12:07:16 pm »
Jumper is same LTZ1000A chip from eBay 3458 A9.
CH1(2) reference had few days runtime before. CH2-CH8 are all first power on (not counting few minutes long test runs). K2002-6 is my 24/7 10V FX reference (uninterrupted power since Feb'18).
I'll need to check CH6 and CH8 units though.
Here I rearrange data to use last yesterday value as 0ppm reference. It's bit easier to see settling time on difference channels.
« Last Edit: November 16, 2018, 12:08:49 pm by TiN »
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
Offline branadic
« Reply #2451 on: December 04, 2018, 05:07:05 am »
One reference fully assembled and running on 3458A since 400h with 10V boost installed. Another 5 references only buffered are almost ready to go and 4 references are still waiting for their LTZ resistors to arrive. What about the other boards that have been spread over the whole world?
-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
The following users thanked this post: Inverted18650, GregDunn, alex-sh, eplpwr
Offline Sprock
  • man sollte sich selbst nicht so wichtig nehmen.
« Reply #2452 on: December 04, 2018, 08:21:52 pm »
Hello all,
Resitors still in Frankfurt / Germany.
regards Sprock
Online BradC
« Reply #2453 on: December 04, 2018, 08:44:38 pm »
Sitting in a big box with other lonely parts. Resistors have been sitting in a USPS facility in Sanfransico for a week now. They should get here before Christmas, just in time for the festive pause hopefully resulting in some assembly in the new year.
Offline pansku
« Reply #2454 on: December 04, 2018, 09:04:18 pm »
Waiting for parts and the ability to use my right hand for soldering *sigh*
Offline SvanGool
« Reply #2455 on: December 04, 2018, 10:45:27 pm »
@Sprock:
Nice card guides, may I ask where you got those?
I assume you are using a bigger enclosure than the Hammond 1455C1210, not (I like the direct ref option you still maintained)?
# Don't hurry, the past will wait. #
Offline Sprock
« Reply #2456 on: December 05, 2018, 01:00:31 am »
Hi
@ SvanGool,
guides are from scrapp box. The original is from Phönix Type FLP/SKBD.
Yes, you are right the case is bigger. It is the 1455 J 1210  instead of C.
No, all 3 boards will get different boxes. Only one in the black aluminium
with board guide.
regards
Sprock
Offline SvanGool
« Reply #2457 on: December 05, 2018, 01:12:54 am »
Thanks !
Mouser has them: https://www.mouser.de/ProductDetail/Phoenix-Contact/2204121 (but not on stock)
« Last Edit: December 05, 2018, 01:19:28 am by SvanGool »
Offline alex-sh
« Reply #2458 on: December 06, 2018, 07:51:38 am »
-branadic-
Mine is still in the box waiting for resistors (group buy on eevblog).
Nice boxes you have. What are they and where did you get them? Are they on Digi-key?
Offline Haasje93
« Reply #2459 on: December 06, 2018, 08:40:53 am »
Hi!
On page 93 is your awnser. But i have here the link for the cases on Digikey.
https://www.digikey.com/products/en?keywords=1455C1201
Regards,
The following users thanked this post: alex-sh
Offline branadic
« Reply #2460 on: December 06, 2018, 08:59:54 am »
Exactly, Hammond 1455C1201 is what the board is designed for. I had a colleague milling the frontpanels for me. The files for them are available a few posts above.
-branadic-
Offline dkozel
« Reply #2461 on: December 07, 2018, 10:39:14 am »
I've ordered the long lead time items but have to finish selecting the last few more basic bits and get them ordered. I've been distracted with figuring out how I'm going to measure these once they're assembled. I don't want to miss the burn-in and do want to measure the three I'm building simultaneously.
I've started testing and warm-up of a pair of 3458A multimeters that I've borrowed (Thread). Along side that I've been looking at different scanner options and managed to find a nice deal here on the forum for a Keithley 7001 with 7011-S cards. They're spec'd for <500nV offset voltage, so >10x worse than TiN's 7168 cards, but I imagine >10x easier to get too. ;) I look forward to the day they aren't good enough for my needs.
Still need to get some cables together, but the last few pages of this thread as well as the various other low EMF threads have been useful there.
Branadic, quick question for you, do the LTZs have >4.2 mm of vertical clearance when in the Hammond case? I'm considering socketing the LTZs if it will work. Also, do you have any plans for how you're using the thermistor? I'm wondering if it's thermally a terrible idea to drill a small hole in the LTZ plastic covers (thanks for the models) and placing the thermistor in direct contact with the case. I have to imagine the thermal transfer would be minimal, but...  :-//
Online Kleinstein
« Reply #2462 on: December 07, 2018, 07:25:59 pm »
There is no real need to measure temperature of the LTZ1000 itself. The temperature sensor is on the chip and the temperature is very well regulated - down to the mK range.
Temperature is more effecting the resistors and maybe thermal EMF from the contacts LTZ to copper. Another point is the change in heater power that might effect the reference.
So a temperature sensor would be better placed at the case or at the resistors.
Offline branadic
« Reply #2463 on: December 08, 2018, 09:08:07 am »
There is for sure enough space inside the Hammond case above the LTZ, but do me a favor and don't use a socket for your reference. I do have a minimal gap between board and LTZ to avoid any stress transfer between them.
The temperature sensor be it a thermistor or a platinum sensor can be used to measure temperature dependance of the whole circuit to verify everthings works a treat. And as Kleinstein already stated, it's not the temperature of the LTZ you want to monitor, but all the other components such as the resistors.
-branadic-
Offline VK5RC
« Reply #2464 on: December 09, 2018, 05:13:02 pm »
Received some of Mr Pettis' resistors during the week so finished stage 2 of a x6 DrFrank's PCB based Ltz Reference Bank
It is located above TiNs KZ PCB reference bank (plus one old 3458 LTZ reference board alone)
(The 'code' next to each output refers to
   A or C  - LTZ1000 A or C,
   1 or 2  - op amp LT1013=1 or 2057=2
   Resistors V =vishay. , U= TE's UPW series, P=Pettis Engineering
All have 2 stage linear PSUs
     1st common rail LM38xx,
      2nd stage each LTZ has its own TPS7A4901 ultra low noise in a separate box
All connected to mains UPS
Pomona 3770 posts for Ref 4-15
My Thanks to EEVBlog VoltNuts esp TiN, DrFrank, Edwin Pettis just to name a few.
Robert
« Last Edit: December 09, 2018, 05:15:38 pm by VK5RC »
The following users thanked this post: TiN, Echo88, Andreas, Inverted18650
Offline Andreas
« Reply #2465 on: December 09, 2018, 07:15:40 pm »
Hello,
Looks like a serious ageing test.  :clap:
with best regards
Andreas
Offline VK5RC
« Reply #2466 on: December 09, 2018, 08:16:02 pm »
Thanks, references 1 to 9 and `3458` have been almost always switched on since Aug 2017. Reference 8 up until Sept was doing brilliantly but then dropped about 8uV. But now is pretty consistent again - within 1-2uV
My results also may be effected by the 800km road trip my old 3458 went on, for calibration in September - all in spec - so no adjustment. The results before and after cal don't appear to have a consistent trend.
One I get a few months of data from the new references , I read every few weeks I will publish here.
Robert
Offline ha110ween
  • Country: cn
« Reply #2467 on: January 11, 2019, 02:38:43 pm »
Hi!
I bulid my LTZ1000A Reference based on Dr.Frank's PCB,and the LT1763 Supply designed by Cellularmitosis.
The Resistors ratio is 12K4:1K.
Keep the long leads of LTZ1000A and protected from airflow by a foamed plastic block.
The plastic box is temporary, I'll replace it with metal shell later.
I put a layer of aluminium foil on the outside of the plastic box to reduce EMI interference.
I'll do some tests later.
Thanks to Dr.Frank,Andreas,branadic,Cellularmitosis,TiN and others.
The following users thanked this post: hwj-d
Offline hwj-d
  • The economy of capitalism has NO social mission!
« Reply #2468 on: January 11, 2019, 03:57:52 pm »
@ha110ween
Nice Job. Thank you for sharing, and welcome to the forum.  :-+
Some information about 10V 'divider' konfiguration and what type of binding posts you use would be nice, after some days of settle down maybe also some logged data.
I like it.  :)
The following users thanked this post: Inverted18650, ha110ween
Offline ha110ween
  • Country: cn
« Reply #2469 on: January 11, 2019, 05:10:56 pm »
@ha110ween
I like it.  :)
Hi hwj-d,
Thank you :D
The output of the Zener is 7.08840v,so the resistors are (18+250+10K):25K.
The posts were removed from my old instruments.
I have a Keysight 34465A and a Fluke 2638A Data Acquisition Unit,I'll use that to record drift data.
The following users thanked this post: hwj-d, Inverted18650
Offline Inverted18650
    • Test Gear Addicts
« Reply #2470 on: January 12, 2019, 06:23:02 am »
Would building this into a custom copper box enclosure be better or worse than building into an aluminum box? Would a solid copper box aid at all in reducing outside emf from the circuit? Please and thank you in advance.
Offline kj7e
  • Damon Stewart
« Reply #2471 on: January 12, 2019, 07:40:00 am »
A steel box would be best for EMI, Aluminum and Copper would provide good RFI shielding but not much for EMI.
Offline eplpwr
  • Wannabe VoltNut
« Reply #2472 on: January 12, 2019, 09:49:51 pm »
As many others: waiting for resistors ... I'm in the last batch for Edwins resistors. The other day I got a mail from Texas Components that they had misplaced my special order from half a year ago, and had produced exactly none of the ordered resistors (some strange story about not beeing able to mfg resistors with 0.125" spacing).
Regarding the Vishays, my alternative is to buy Z201 and test them for tempco, which will cause further delays.
Precision takes time.  :=\
Online beanflying
« Reply #2473 on: January 13, 2019, 11:25:42 am »
Not at all happy with Texas Components and custom Vishay delivery times lost orders and email excuses :horse:
Coffee, Food, R/C and electronics nerd in no particular order :)
Offline branadic
« Reply #2474 on: January 16, 2019, 12:56:58 am »
Here is one unit based on the previously shown board. It uses Rhopoint 8G16D resistors with 1k:12k for temperature setting at the LTZ1000CH (manufactured in 2018) and 4k (2x 2k) + 10k UPW50 resistors plus a 20ohm 64W pot with 22ohm resistor in parallel to one branch of the pot for the 10V boost circuit.
The voltage reference was adjusted to the corrected readings (reads 9,9999945V instead of 10,000000V with respect to the calibration report) on our 3458A (still within 90day specs). It will now take the big trip overseas in its double thermally insulated package to reach its new owner. Hopefully it does a good job there.
-branadic-