« on: May 24, 2017, 02:45:15 pm »
Recently, I picked up a KE 230 for USD$200-ish. The unit was advertised AS-IS.
After a multi-day warm-up the unit is being measured with a KE DMM7510. The 230 is set to output 10V.
The measured statistics are
Average: 10.000911
Std Dev: 248.5 uV
Readings: 396,000
NPLC 5
So, the stability is about 25ppm RMS while my 731Bs are typically 0.X ppm stable in the same rack over the same temperature range.
I am wondering what can be done if anything to make the PWM circuit more stable. Suggestions are welcomed. I have never delved into a PWM voltage generator.