I was intrigued by the poor performance of these probes at high frequencies. So I took a closer look.

I measured the response using an RF signal generator and an oscilloscope.


First the gain at 1MHz:

This is correct it is showing a 500x reduction in signal amplitude.
At 44 MHz I found a resonance:

The gain is much higher than it should be, 23.96mV out with 320mV input. That is divide by 13.3 instead of divide by 500.
I noticed that gain was sensitive to the arrangement of the leads.
A quick LTspice model shows that the 7pF capacitance of the probes is resonating with a few uH of inductance.


I am sure that this could be engineered better to dampen this resonance.